According to MDR Annex III, a PMS plan contains
- a PMCF plan as referred to in Part B of Annex XIV, or a justification as to why a PMCF is not applicable.
A PMCF plan is always required. It might not be applicable for devices that have no direct influence on the patient, e.g. disinfectants. Or when the device is used in another device, e.g. dental implant colouring agents.
Always remember: PMCF ≥ PMCF study. You can often justify not to perform PMCF studies, but you can rarely justify not to perform any PMCF at all.
 
#3
According to MDR Annex III, a PMS plan contains

A PMCF plan is always required. It might not be applicable for devices that have no direct influence on the patient, e.g. disinfectants. Or when the device is used in another device, e.g. dental implant colouring agents.
Always remember: PMCF ≥ PMCF study. You can often justify not to perform PMCF studies, but you can rarely justify not to perform any PMCF at all.
Thank You for your response. How can I justify not performing PMCF studies?
 
Thread starter Similar threads Forum Replies Date
A Post-market clinical followup studies (PMCF) EU Medical Device Regulations 16
M PMCF (Post Market Clinical Followup)/ PSUR template EU Medical Device Regulations 9
S PMCF (Post Market Clinical Followup): Rationale for Exemption - Software Medical Devices EU Medical Device Regulations 6
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
W Interesting Discussion PMCF (Post-Market Clinical Followup) vs PMCF studies EU Medical Device Regulations 39
C PMCF (Post Market Clinical Followup) Enrollment Sizes (Spinal Implants) EU Medical Device Regulations 5
A Post-market Clinical Studies of Orthopaedic (Revision) Implants EU Medical Device Regulations 1
S Post Market Clinical Follow up (PCMF) Clarification EU Medical Device Regulations 6
L Lost in the Post-Market Surveillance Clinical Follow-up EU Medical Device Regulations 1
K Post-Market Clinical Follow-up as part of Post Market Surveillance? EU Medical Device Regulations 15
L Surveys for post-market clinical follow-up EU Medical Device Regulations 1
D 'Post market surveillance' versus 'Post Market Clinical Follow Up' EU Medical Device Regulations 18
D Is post-market clinical follow-up needed for IVD-MDs? MDD (Annex X, 1.1c) EU Medical Device Regulations 5
S Post Market Surveillance Report (PMSR) for Medical Devices EU Medical Device Regulations 6
A SaMD: Post Market Surveillance after Decomissioning EU Medical Device Regulations 2
MaHoDie Summative Evaluation with Post-Market Data? Human Factors and Ergonomics in Engineering 2
I Post Market Surveillance Process EU Medical Device Regulations 3
I Post-market surveillance and Post-market performance follow-up for in vitro diagnostic medical devices EU Medical Device Regulations 0
M Post Market Safety Update Report for devices that will be up-classified under MDR but are currently under MDD EU Medical Device Regulations 6
thisby_ Post Market/Production Risk Assessment ISO 14971 - Medical Device Risk Management 0
N Post Market Surveillance Package Example Wanted ISO 13485:2016 - Medical Device Quality Management Systems 1
M Differences in post market safety reporting for Combination Product Applicants Medical Device and FDA Regulations and Standards News 1
K Product (Device) Life cycle - Product Life AFTER the Post Market Phase Medical Device and FDA Regulations and Standards News 3
M Post Market Surveillance for Class II Medical Devices Medical Device and FDA Regulations and Standards News 3
H Post-market surveillance strategy EU Medical Device Regulations 2
B Procedures for Complaint Handling and Post Market Surveillance EU Medical Device Regulations 10
M Informational Creating a post market surveillance (PMS) system for medical devices – Part 1 Medical Device and FDA Regulations and Standards News 7
E Post Market Surveillance Standard Various Other Specifications, Standards, and related Requirements 4
T Post-Market Surveillance Plan - Requirements outlined in Annex III (1.1b) EU Medical Device Regulations 3
Y Post Market Surveillance Indicators and Analysis method EU Medical Device Regulations 1
M Post Market Surveillance Plan Template EU Medical Device Regulations 12
F Vietnamese Post-Market Surveillance Requirements for a medical device in Japan Other Medical Device Regulations World-Wide 0
S Post Market Surveillance and Annual Product Review in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 2
M Medical Device News Health Canada Notice of intent: Strengthening the post-market surveillance and risk management Canada Medical Device Regulations 1
K EU MDR - PMS (Post Market Surveillance) SOP - definition of "register(s)" EU Medical Device Regulations 2
S An auditor asked for a Post Market Surveillance Report ISO 14971 - Medical Device Risk Management 4
T Proactive / Preventive Post Market Surveillance question ISO 13485:2016 - Medical Device Quality Management Systems 4
M ISO TR 20416 Medical Devices - Post-Market Surveillance for Manufacturers Other Medical Device Related Standards 8
S Merging Post Market Surveillance and Complaint Handling ISO 13485:2016 - Medical Device Quality Management Systems 3
S Post Market Surveillance procedure/ SOP Medical devices Imported Legacy Blogs 0
A Industry best practice about Post-Market Surveillance and Risk Analysis ISO 14971 - Medical Device Risk Management 6
S Korean Post Market Surveillance and Recall Regulations Other Medical Device Regulations World-Wide 3
D What is your Post-Market Surveillance Strategy 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
Z When to stop Post Market Surveillance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Verification and Validation of Post Market Design Change Design and Development of Products and Processes 2
C How to write a Post Market Surveillance REPORT (not procedure) EU Medical Device Regulations 4
N MEDDEV guideline on Post Market Surveillance (PMS) EU Medical Device Regulations 8
Q Post Market Surveillance SOP - ISO 13485 Audit Nonconformance ISO 13485:2016 - Medical Device Quality Management Systems 3
T Post Market Surveillance Sample Procedure for Medical Devices EU Medical Device Regulations 2

Similar threads

Top Bottom