Hi!
Based on MEDDEV 2.12/2 recommendations:
PMCF studies can follow several methodologies, for example:
a. the extended follow-up of patients enrolled in premarket investigations;
b. new clinical investigation;
c. review of data derived from a device registry; or
d. review of relevant retrospective data from patients previously exposed to the device.
1. Is it obligatory to perform the PMCF of patients? Please, give us a detailed explanation of sec. c and d.
2. Can we study only patients from EU countries or we can use patients from other world regions?
3. Is there any other alternative of PMCF? Can we skip it?
 
Elsmar Forum Sponsor
#2
Hi Aleksandra,

1. Short answer: it depends.
Longer answer: it depends on the risk class and general safety and performance of the device, and how it interacts with patients/users. For example, an implantable device will most likely require that PMCF studies are conducted, although I can't say for certain what types would be required (I don't have any experience with implantables, so maybe someone else can shine some light on that). Devices intended for high-risk populations, high-risk anatomical locations, or for severe disease/treatment challenges generally require them.

A good way to determine if PMCF studies are needed is by using the MEDDEV 2.12/2 Section 5 criteria. There's a really great list that can be formulated into a checklist. I'd suggest bringing this into your QMS as a PMCF Determination Form. Decide yes or no for each line, with a justification - this should put you on the right path to determining if they are needed, as well as providing evidence that your decision was formulated based on the guidance and not just a guess.

2. This depends on local regulations, I would think. Where are you marketing the device? How does the device interact with the patient/user? Some countries require that clinical data include persons of that nationality. Others accept foreign clinical trial data. Again, it depends.

3. Yes, there is, but again, this depends on the device in question. Another example: suppose you are a manufacturer of an x-ray tube. This is a higher class device that requires exposing the patient to radiation to get images for diagnosis. You'd think that something so potentially dangerous would require PMCF, right? Looking closer, we see that the tube itself does not interface with the patient, and that the technology is already well-known and established as being safe. From the MEDDEV guidance, "PMCF studies may not be required when the medium/long-term safety and clinical performance are already known from previous use of the device or where other appropriate post-market surveillance activities would provide sufficient data to address the risks". How long has your device been marketed? Have there been any significant changes to the design that may impact patient safety? Is the technology long-standing and regarded as safe? Depending on your answers to these questions, complaint handling, PSURs, database searches, tracking and trending, etc., may be enough to determine if there are any emergent clinical risks, or a need to update the RBA.

Hope this answers some of your questions! Maybe someone with a little more experience can chime in.
 

SKM.Sunil

Involved In Discussions
#3
Hi!
Based on MEDDEV 2.12/2 recommendations:
PMCF studies can follow several methodologies, for example:
a. the extended follow-up of patients enrolled in premarket investigations;
b. new clinical investigation;
c. review of data derived from a device registry; or
d. review of relevant retrospective data from patients previously exposed to the device.
1. Is it obligatory to perform the PMCF of patients? Please, give us a detailed explanation of sec. c and d.
YES
2. Can we study only patients from EU countries or we can use patients from other world regions?
YES- You can perform from all over the world
3. Is there any other alternative of PMCF? Can we skip it?
NO - PMCF is mandatory as long as your device is intended for human use. (excuse PMCF Studies)

c. review of data derived from a device registry; or
There are major device registries around the globe, major to mention- MAUDE, MHRA registry, TGA registry, etc...
when there is a reported event for a device similar/equivalent to your device, you can always collect the data/analysis report on the relevant event and evaluate it with your device, how it may or may not affect your device while use in forcible or nonforcible intend.
<This method require very explanatory demonstration how to do it ......I would suggest you should run a research on your own.....if you want I can help.>

d. review of relevant retrospective data from patients previously exposed to the device.
As per PMCF your device is already in the market and used on patient. You can adopt a cleaver way to trace the patients/humans used your device till date and collect the device data from them like how it performed and safety complications.
you require to take help from doctor and ethics committee approval to perform certain studies like this kind.
once you set you can target the regions and collect the data and evaluate and frame a report.
<This method require very explanatory demonstration how to do it ......I would suggest you should run a research on your own.....if you want I can help.>
 

Watchcat

Trusted Information Resource
#5
There are major device registries around the globe, major to mention- MAUDE
MAUDE is not a registry, in that no devices or patients are registered there. MAUDE contains data taken from the reports of issues with medical devices that submitted to FDA by device manufacturers, healthcare providers, and patients. FDA has a device registry, which is its registration and listing database. This database does not include any data on the devices that are listed in it. They are just listed.

MAUDE is a legacy database. It is accurately described as containing data "taken from reports," not data about devices identified in the reports. The data are junk and shouldn't be used for much of anything. If you don't believe me, believe FDA. It says so in a bulleted list of caveats posted on the MAUDE search page, where anyone who is interested in the quality of the data in the MAUDE database can't miss them. I expect most of the other databases companies search for the sake of searching have similar problems, although perhaps the owners of some of these other databases are not so transparent about their limitations.

If you want useful postmarket data from a registry, you must design and populate it yourself, which some companies do. Other companies have been scrambling around for the past several years, trying to figure out how to avoid setting up their own registries, or otherwise collect postmarket data on use of their devices. One approach to this avoidance is to use data for their purposes that they find in databases that were set up by other entities (e.g. regulators, health organizations) to serve the purposes of those who set them up.

I don't think this is going to work out well for these companies over the long haul. How long a haul it is likely to be...in the EU, that depends on the NBs that are designated under the MDR, and maybe also whatever replaces the current MEDDEVs, post-MDR. Otherwise, we wait for another "PIP" to inspire further scrutiny of the industry's postmarket surveillance practices.
 

SKM.Sunil

Involved In Discussions
#7
MAUDE is not a registry, in that no devices or patients are registered there. MAUDE contains data taken from the reports of issues with medical devices that submitted to FDA by device manufacturers, healthcare providers, and patients. FDA has a device registry, which is its registration and listing database. This database does not include any data on the devices that are listed in it. They are just listed.

MAUDE is a legacy database. It is accurately described as containing data "taken from reports," not data about devices identified in the reports. The data are junk and shouldn't be used for much of anything. If you don't believe me, believe FDA. It says so in a bulleted list of caveats posted on the MAUDE search page, where anyone who is interested in the quality of the data in the MAUDE database can't miss them. I expect most of the other databases companies search for the sake of searching have similar problems, although perhaps the owners of some of these other databases are not so transparent about their limitations.

If you want useful postmarket data from a registry, you must design and populate it yourself, which some companies do. Other companies have been scrambling around for the past several years, trying to figure out how to avoid setting up their own registries, or otherwise collect postmarket data on use of their devices. One approach to this avoidance is to use data for their purposes that they find in databases that were set up by other entities (e.g. regulators, health organizations) to serve the purposes of those who set them up.

I don't think this is going to work out well for these companies over the long haul. How long a haul it is likely to be...in the EU, that depends on the NBs that are designated under the MDR, and maybe also whatever replaces the current MEDDEVs, post-MDR. Otherwise, we wait for another "PIP" to inspire further scrutiny of the industry's postmarket surveillance practices.

Thanks for Clarifying more...
 

Wolf.K

Involved In Discussions
#8
MAUDE is a legacy database. It is accurately described as containing data "taken from reports," not data about devices identified in the reports. The data are junk and shouldn't be used for much of anything.
Well, I like to use it, as it is the largest database I can freely access and includes most of our competing products. Our NB likes our analysis of the reports found there, as this data directly feeds into our risk management. What other registers could you suggest?
 

Watchcat

Trusted Information Resource
#9
Once the MDR designation thing settles down, I'm going to do my best to educate the NBs on the worthlessness of the data in MAUDE. I might critique some other registries, too, if I ever get the time.

The registry of implant suppliers referenced by dgrainger above looks promising.

I'm not a fan of registries, especially the ones "maintained" by regulatory agencies. I tend to think they are there mostly so they can say they have one, like a lot of companies have Quality departments mostly for show.

I would recommend a registry based on a complete understanding of what the data in it tell you, which is usually determined by who reports, what is reported, when, and why.
 

Ed Panek

QA RA Small Med Dev Company
Trusted Information Resource
#10
We use MAUDE data for our device type FLL for trends with devices like ours. I pull that data out and pivot table the Sh&^ out of it and present it. Are you suggesting this is not valuable? Maybe I misread your comment.
 
Thread starter Similar threads Forum Replies Date
M PMCF (Post Market Clinical Followup)/ PSUR template EU Medical Device Regulations 8
S PMCF (Post Market Clinical Followup): Rationale for Exemption - Software Medical Devices EU Medical Device Regulations 6
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
W Interesting Discussion PMCF (Post-Market Clinical Followup) vs PMCF studies EU Medical Device Regulations 36
C PMCF (Post Market Clinical Followup) Enrollment Sizes (Spinal Implants) EU Medical Device Regulations 5
A Post-market Clinical Studies of Orthopaedic (Revision) Implants EU Medical Device Regulations 1
S Post Market Clinical Follow up (PCMF) Clarification EU Medical Device Regulations 6
L Lost in the Post-Market Surveillance Clinical Follow-up EU Medical Device Regulations 1
K Post-Market Clinical Follow-up as part of Post Market Surveillance? EU Medical Device Regulations 15
L Surveys for post-market clinical follow-up EU Medical Device Regulations 1
D 'Post market surveillance' versus 'Post Market Clinical Follow Up' EU Medical Device Regulations 18
D Is post-market clinical follow-up needed for IVD-MDs? MDD (Annex X, 1.1c) EU Medical Device Regulations 5
N Post Market Surveillance Package Example Wanted ISO 13485:2016 - Medical Device Quality Management Systems 1
M Differences in post market safety reporting for Combination Product Applicants Medical Device and FDA Regulations and Standards News 1
K Product (Device) Life cycle - Product Life AFTER the Post Market Phase Medical Device and FDA Regulations and Standards News 3
M Post Market Surveillance for Class II Medical Devices Medical Device and FDA Regulations and Standards News 2
H Post-market surveillance strategy EU Medical Device Regulations 1
B Procedures for Complaint Handling and Post Market Surveillance EU Medical Device Regulations 7
M Informational Creating a post market surveillance (PMS) system for medical devices – Part 1 Medical Device and FDA Regulations and Standards News 7
E Post Market Surveillance Standard Various Other Specifications, Standards, and related Requirements 4
T Post-Market Surveillance Plan - Requirements outlined in Annex III (1.1b) EU Medical Device Regulations 2
Y Post Market Surveillance Indicators and Analysis method EU Medical Device Regulations 1
M Post Market Surveillance Plan Template EU Medical Device Regulations 6
F Vietnamese Post-Market Surveillance Requirements for a medical device in Japan Other Medical Device Regulations World-Wide 0
S Post Market Surveillance and Annual Product Review in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 2
M Medical Device News Health Canada Notice of intent: Strengthening the post-market surveillance and risk management Canada Medical Device Regulations 1
K EU MDR - PMS (Post Market Surveillance) SOP - definition of "register(s)" EU Medical Device Regulations 2
S An auditor asked for a Post Market Surveillance Report ISO 14971 - Medical Device Risk Management 4
T Proactive / Preventive Post Market Surveillance question ISO 13485:2016 - Medical Device Quality Management Systems 4
M ISO TR 20416 Medical Devices - Post-Market Surveillance for Manufacturers Other Medical Device Related Standards 6
S Merging Post Market Surveillance and Complaint Handling ISO 13485:2016 - Medical Device Quality Management Systems 3
S Post Market Surveillance procedure/ SOP Medical devices Imported Legacy Blogs 0
A Industry best practice about Post-Market Surveillance and Risk Analysis ISO 14971 - Medical Device Risk Management 6
S Korean Post Market Surveillance and Recall Regulations Other Medical Device Regulations World-Wide 3
D What is your Post-Market Surveillance Strategy 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
Z When to stop Post Market Surveillance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S Verification and Validation of Post Market Design Change Design and Development of Products and Processes 2
C How to write a Post Market Surveillance REPORT (not procedure) EU Medical Device Regulations 4
N MEDDEV guideline on Post Market Surveillance (PMS) EU Medical Device Regulations 8
Q Post Market Surveillance SOP - ISO 13485 Audit Nonconformance ISO 13485:2016 - Medical Device Quality Management Systems 3
T Post Market Surveillance Sample Procedure for Medical Devices EU Medical Device Regulations 2
A Applicable Regulations for a PMS (Post Market Surveillance) System ISO 14971 - Medical Device Risk Management 5
J ISO 13485 - Procedure for a Feedback System (Post Market Surveillance) ISO 13485:2016 - Medical Device Quality Management Systems 10
D Post-Market Compliance with Investigational AEs 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
T Post Market Surveillance Requirements in Israel Other Medical Device Regulations World-Wide 17
E Post Market Surveillance Guidelines or examples of procedures US Food and Drug Administration (FDA) 7
SteveK Post Market Surveillance SOP - Word Version EU Medical Device Regulations 42
J Post-Market Surveillance required by the EU's MDD vs. ISO 13485 Requirements ISO 13485:2016 - Medical Device Quality Management Systems 55
C Audit NonConformance - Evidence of Post Market Vigilance and Product Recall. ISO 13485:2016 - Medical Device Quality Management Systems 6

Similar threads

Top Bottom