Post-Market Surveillance required by the EU's MDD vs. ISO 13485 Requirements

Z

zkoulou

#21
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

Hi Al.
what template? it has been awhile
 
Elsmar Forum Sponsor
T

tec11

#22
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

Thank you sooo much. this doc was immensely useful to me.
 

RCW

Quite Involved in Discussions
#23
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

As a contract manaufacturer of medical devices, would I need to have a PMS system in place? (Assume ISO 13485 requirements are present.) As hard as it is to get feedback out of the user of the medical device, when you are a level below the company marketing the device, it's even more difficult.

Note: My customers are addressing the advisory notices and are interfacing with the medical device users.
 
M

MIREGMGR

#24
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

As a contract manaufacturer of medical devices, would I need to have a PMS system in place? (Assume ISO 13485 requirements are present.)
No as a regulatory requirement.

Maybe yes as a contractural requirement, if one or more of your customer agreements calls for it.
 
S

SteveK

#25
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

I suppose if there was a problem with a competitors product it would be relevant, but this would be captured in a Device Alert (a MDA, Safety Notice or Recall in the UK - available on the MHRA site) of some form (i.e. a publication again).
Just as a follow up to my comment about using alerts, FSNs, recalls, safety notices etc (e.g. issued in the UK by MHRA) as a form of PMS i.e. monitoring for medical devices similar to your own, these sites listed may be useful. Obviously it can be a pain to trawl through a lot information to find anything relevant, but at least I can show an auditor a degree of vigilance. I do this by recording the total number of alerts etc for each month and authority on a spreadsheet. Then I highlight any link to the type of devices similar to our own. You may encounter a problem you were not aware of, so then you can update your Risk Assessment file accordingly. For foreign language sites you can use the Google translator. Note for the FDA site since so many alerts are thrown up, I filter these by worst case i.e. causing ‘death’ (it helps first off to get the 3 letter code for your type of device for the search).

MHRA
http://www.mhra.gov.uk/Safetyinformation/Safetywarningsalertsandrecalls/index.htm

FDA
http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfMAUDE/search.CFM

HC (Canada)
http://www.hc-sc.gc.ca/dhp-mps/compli-conform/recall-retrait/_list/index-eng.php

MDB (Malaysia)
http://www.mdb.gov.my/mdb/index.php?option=com_content&task=view&id=69&Itemid=119

HAS (Hong Kong)
http://www.mdco.gov.hk/english/recalls/recalls.html

IMB (Ireland)
http://www.imb.ie/EN/Medical-Devices/Safety-Notices.aspx

Saudi FDA
http://212.100.220.58/services/md_Recalls/recallsDispaly.aspx

Afssaps (France)
http://www.afssaps.fr/Infos-de-securite/Autres-mesures-de-securite#dm

bfarm (Germany)
http://www.bfarm.de/cln_103/EN/medDev/riskinfo/fca/fca-node.html

Salute (Italy)
http://www.salute.gov.it/dispositivi/archivioAvvisiDispo.jsp?menu=avvisi&lingua=italiano

Swissmedic
http://www.swissmedic.ch/rueckrufe_medizinprodukte/archiv/index.html?lang=de&RlArchiv=2010-07
Not all countries publish details of their alert, but if anyone else knows of any other sites that do, please let me know.

Steve
 
M

MegSinha

#26
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

Hi Steve,

My Company has the PMCF NC from our NB, I am currently working on putting a procedure together but the post related to Post Market Clinical follow up named
"Post Market Surveillance Guidelines or examples of procedures"
(I just joined so can't put the exact link )
In this post Sam says that it is not deemed necessary if duly justified? Do you think its just better do it or it is better to justify why we are not doing it since the product classification is IIa?

Please advise.
 
S

SteveK

#27
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

The link is:

Post Market Surveillance Guidelines or examples of procedures

Anyway I believe a "clinical follow-up" relates to implantable/class III products. There is a current case (originally initiated by the FDA) that DePuy (J&J) must conduct a "clinical follow-up" on their metal on metal hip replacements (i.e. on individual patients) - which have now been withdrawn. I think this is the context of the requirement, so I do not see how it could easily relate to a class IIa device. As Sam says, PMCFU is a sub-set of PMS. I would be interested in what your NC actually says. It may be down to how your NB defines PMCFU i.e. for your class IIa device instead that is the problem.

Steve
 
M

MegSinha

#28
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

This is what exactly the NC says:
... "The procedure on Post-Market Surveillance System was provided. Although it refers to the need to establish “a post market surveillance plan for each product”, it does not introduces the notion of Post Marketing Clinical Follow-up as required by the Medical Device Directive in it Annex X - section 1.1.quater"

... What I figured is that they needed a PMCFU to be a part of our PMS, but as I said in Annex X though is does not state explicitly, it does state that this is required for Class III and implantables, so should I just add a procedure for clinical evaluation or should I talk to them and say that our product is Class IIa, does not require clinical evaluation?
 
S

SteveK

#29
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

This is what exactly the NC says:
... "The procedure on Post-Market Surveillance System was provided. Although it refers to the need to establish “a post market surveillance plan for each product”, it does not introduces the notion of Post Marketing Clinical Follow-up as required by the Medical Device Directive in it Annex X - section 1.1.quater"

... What I figured is that they needed a PMCFU to be a part of our PMS, but as I said in Annex X though is does not state explicitly, it does state that this is required for Class III and implantables, so should I just add a procedure for clinical evaluation or should I talk to them and say that our product is Class IIa, does not require clinical evaluation?
You will require a clinical evaluation, but this can be done by the literature route (I have have an example document if you do a search on 'humidifier'). I do not think this is the same as a PMCFU for a Class IIa device for the reasons already stated.

Steve
 
M

MegSinha

#30
Re: Post-Market Surveillance required by the EU’s MDD vs. ISO 13485 requirements

Thanks a lot Steve. That helps!
 
Thread starter Similar threads Forum Replies Date
N Post Market Surveillance Package Example Wanted ISO 13485:2016 - Medical Device Quality Management Systems 1
M Post Market Surveillance for Class II Medical Devices Medical Device and FDA Regulations and Standards News 2
H Post-market surveillance strategy EU Medical Device Regulations 2
B Procedures for Complaint Handling and Post Market Surveillance EU Medical Device Regulations 7
M Informational Creating a post market surveillance (PMS) system for medical devices – Part 1 Medical Device and FDA Regulations and Standards News 7
E Post Market Surveillance Standard Various Other Specifications, Standards, and related Requirements 4
T Post-Market Surveillance Plan - Requirements outlined in Annex III (1.1b) EU Medical Device Regulations 2
Y Post Market Surveillance Indicators and Analysis method EU Medical Device Regulations 1
M Post Market Surveillance Plan Template EU Medical Device Regulations 10
F Vietnamese Post-Market Surveillance Requirements for a medical device in Japan Other Medical Device Regulations World-Wide 0
S Post Market Surveillance and Annual Product Review in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 2
M Medical Device News Health Canada Notice of intent: Strengthening the post-market surveillance and risk management Canada Medical Device Regulations 1
K EU MDR - PMS (Post Market Surveillance) SOP - definition of "register(s)" EU Medical Device Regulations 2
S An auditor asked for a Post Market Surveillance Report ISO 14971 - Medical Device Risk Management 4
T Proactive / Preventive Post Market Surveillance question ISO 13485:2016 - Medical Device Quality Management Systems 4
M ISO TR 20416 Medical Devices - Post-Market Surveillance for Manufacturers Other Medical Device Related Standards 6
S Merging Post Market Surveillance and Complaint Handling ISO 13485:2016 - Medical Device Quality Management Systems 3
S Post Market Surveillance procedure/ SOP Medical devices Imported Legacy Blogs 0
A Industry best practice about Post-Market Surveillance and Risk Analysis ISO 14971 - Medical Device Risk Management 6
S Korean Post Market Surveillance and Recall Regulations Other Medical Device Regulations World-Wide 3
D What is your Post-Market Surveillance Strategy 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
Z When to stop Post Market Surveillance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
C How to write a Post Market Surveillance REPORT (not procedure) EU Medical Device Regulations 4
N MEDDEV guideline on Post Market Surveillance (PMS) EU Medical Device Regulations 8
Q Post Market Surveillance SOP - ISO 13485 Audit Nonconformance ISO 13485:2016 - Medical Device Quality Management Systems 3
T Post Market Surveillance Sample Procedure for Medical Devices EU Medical Device Regulations 2
A Applicable Regulations for a PMS (Post Market Surveillance) System ISO 14971 - Medical Device Risk Management 5
J ISO 13485 - Procedure for a Feedback System (Post Market Surveillance) ISO 13485:2016 - Medical Device Quality Management Systems 10
T Post Market Surveillance Requirements in Israel Other Medical Device Regulations World-Wide 17
L Lost in the Post-Market Surveillance Clinical Follow-up EU Medical Device Regulations 1
E Post Market Surveillance Guidelines or examples of procedures US Food and Drug Administration (FDA) 8
S Post Market Surveillance SOP - Word Version EU Medical Device Regulations 42
K Post-Market Clinical Follow-up as part of Post Market Surveillance? EU Medical Device Regulations 15
D 'Post market surveillance' versus 'Post Market Clinical Follow Up' EU Medical Device Regulations 18
B ISO13485 Sterilisation Validation, Risk Management, Post Market Surveillance ISO 13485:2016 - Medical Device Quality Management Systems 13
thisby_ Post Market/Production Risk Assessment ISO 14971 - Medical Device Risk Management 0
M Differences in post market safety reporting for Combination Product Applicants Medical Device and FDA Regulations and Standards News 1
K Product (Device) Life cycle - Product Life AFTER the Post Market Phase Medical Device and FDA Regulations and Standards News 3
A Post-market clinical followup studies (PMCF) EU Medical Device Regulations 16
M PMCF (Post Market Clinical Followup)/ PSUR template EU Medical Device Regulations 8
S PMCF (Post Market Clinical Followup): Rationale for Exemption - Software Medical Devices EU Medical Device Regulations 6
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
W Interesting Discussion PMCF (Post-Market Clinical Followup) vs PMCF studies EU Medical Device Regulations 36
C PMCF (Post Market Clinical Followup) Enrollment Sizes (Spinal Implants) EU Medical Device Regulations 5
A Post-market Clinical Studies of Orthopaedic (Revision) Implants EU Medical Device Regulations 1
S Verification and Validation of Post Market Design Change Design and Development of Products and Processes 2
S Post Market Clinical Follow up (PCMF) Clarification EU Medical Device Regulations 6
D Post-Market Compliance with Investigational AEs 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
L Surveys for post-market clinical follow-up EU Medical Device Regulations 1