Power Input Marking requirements in IEC 60601 and IEC 60950
Roland chung
Hi all,
The equipment is designed to be connected to an AC MAINS having any voltage between 100 V and 240 V.
According to Cl. 7.2.7 of IEC 60601-1, the rated input shall be given as a range (e.g. 2 A - 1 A).
In the case of ME EQUIPMENT for one or several RATED voltage ranges, the RATED input shall always be given for the upper and lower limits of the range or ranges, if the range(s) is/are greater than ± 10 % of the mean value of the given range.
However, equipment marked with only maximum rated input (e.g. max 2 A) is allowed according to IEC 60950.
equipment with a RATED VOLTAGE RANGE shall be marked with either the maximum RATED CURRENT or with the current range
In the case of ME equipment employs a stand-alone 60950 compliant SMPS which marked with maximum rated input only, does this constitute a failure?
Thanks,
Roland
Peter Selvey
If the equipment incorporates a stand alone power supply, it may be best to handle as a system under Clause 16. That way, little differences between the two standards like this don't cause a problem. It is allowed in 3rd edition that the medical device gets it's power from a 950 supply.
Roland chung
Yes, it is a good approach to consider the combination as a system.
I have seen many approved ME equipments with voltage range, but only maximum volt-ampere is given. For example: 100 V - 240 V, 240 VA.
It is not uncommon that, the volt-amperes for upper limit of the range (in this case, 240 V) and for lower limit of the range (100 V) are close (the current could be very different). For example:
100 V --> 2.2 A --> 220 VA
240 V --> 1.0 A --> 240 VA
Is it reasonable to give the maximum volt-ampere only?
Peter Selvey
Because the related test (power input 4.11) only has an upper limit (+10%), having one rating for 100-240V only can still comply. The requirements really only makes sense if the test in 4.11 has both upper and lower limits, for example +10%, -30%.
P Medical Device Marking - Power Input Ports IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60335-1 power input deviation - Table 1 Values Other ISO and International Standards and European Regulations 0
K How to determine rated input power for the markings? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Judy Abbott The impact of laser power on biocompatibility of the products Manufacturing and Related Processes 6
C Power Supply EN60601-1 vs EN62368-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S DC-AC Power inverter with two MOPPs IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Surge Protectors and other power protection Manufacturing and Related Processes 7
Sidney Vianna Informational APQP4Wind - Advanced Product Quality Planning for the Wind Power Supply Chain APQP and PPAP 3
U Power supply - Change From 160 to 300 Watts - Significant change? EU Medical Device Regulations 2
M MS Power Automate/Sharepoint Files Document Control Systems, Procedures, Forms and Templates 1
O Are power distribution units Multiple Socket Outlets (MSOs)? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L Medical-grade power adapter labeling Other US Medical Device Regulations 2
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Proud Liberal Power Query to combine multiple tables of unequal number of columns Statistical Analysis Tools, Techniques and SPC 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
L IEC 62368-1 / Electrical Energy Source vs. Power Source? CE Marking (Conformité Européene) / CB Scheme 2
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
I Containment procedure for parts after a power outage. IATF 16949 - Automotive Quality Systems Standard 1
P UDI labeling for power module IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
S Power rating for medical devices in the US - Radiotherapy devices US Food and Drug Administration (FDA) 0
T Is anyone working with N299.1 (Supply/service to nuclear power plants)? Various Other Specifications, Standards, and related Requirements 0
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Power point training slides on DOE (Design of Experiments) Wanted Training - Internal, External, Online and Distance Learning 2
Marc Renewables, led by wind, provided more power than coal in Germany in 2018 Sustainability, Green Initiatives and Ecology 2
S Responsibilities between government owner and private subcontractor of a power plant seeking ISO 9001 certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
C Electrical Power Plant Supply failure in PFMEAs FMEA and Control Plans 10
H IEC 60601-1 Main Power Cord Requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Infusion Pump Power Failure Alarm IEC 60601-2-24 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Power Cord Length Issue - Medical Device Electrical Safety IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ISO 9001:2015 - Non Applicable Design and Development (Power Plant Operator) Design and Development of Products and Processes 2
H Enclosed vs. Open-Frame Power Supplies - Overdoing it? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 2
G Medical Device Power Inlet/Outlet Voltage Labeling IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
x-files [Thermal Power plant KOSTOLAC, SERBIA] Sucessfully certified to ISO 50001 Sustainability, Green Initiatives and Ecology 1
M Is IEC 60601 compliance possible when only generic power-supply is specified? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
U IEC 60601-1, Class 1 or Class 2 on emergency power system IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Power Traceability certificate for values for voltage, current and phase General Measurement Device and Calibration Topics 1
D NQA-1 Standard Applied to Services rather than Items for Nuclear Power Various Other Specifications, Standards, and related Requirements 3
R USB power between Medical Device Data Collection Device Component IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F Power and sample size for factorial design Using Minitab Software 4
I Reaction plan in case of power outage Misc. Quality Assurance and Business Systems Related Topics 7
B Detachable Power Supply Cable Connection ESD Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R Selling of Medical Power Supply in Australia/New Zealand and Japan Other Medical Device Regulations World-Wide 5
M Multiple versions of device due to power connectors? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2