Roland chung

Trusted Information Resource
#12
Sorry Peter, it is difficult to get your points in this topic.

It is also worth to note that most EMC capacitors have a negative temperature co-efficient. In some cases the leakage can be up to 20% higher in cold condition due to this effect. It has nothing to do with the equipment absorbing moisture. So even if you don't run the humidity treatment, make sure to measure the leakage in cold condition.
I agree with the leakage in cold condition is higher than leakage in warm-up state. However, the same situation also exist in leakage after humidity treatment. We are talking about the leakage before and after humidity BUT not different states (i.e. cold AND warm-up) in same period (before humidity OR after humidity).

Again, the increase in leakage current after humidity conditioning is most likely due to the negative temperature co-efficient (NTC) in EMC capacitors.
I really want to know what is the relationship between humidity treatment and cold condition/warm-up you mentioned.

Again, the temperature in humidity chamber is same as ambient temperature (cold condition) in order to avoid condensation. Therefore, the temperature in humidity chamber never influence EMC capacitor unless you switch on the equipment when perform the humidity test so cause the equipment to warm up.

The fact is that the leakage current after humidity is 10% higher than leakage current before humidity. I believe it is because of moisture.
 

Peter Selvey

Staff member
Super Moderator
#13
Just for reference, the standard requires two measurements, one at normal operating temperature/normal humidity, and one cold condition/93% RH. The IECEE CB scheme protocol uses the terms "before" and "after" to distinguish these two measurements. So, when looking at the numbers in a test report (or talking to an IEC test engineer), "before" and "after" means "hot" and "cold". If the leakage is primarily due to EMC capacitors, the "after" readings will normally be higher than "before" due to the NTC parameter.

But you may be referring to "before" to mean a special measurement, not required by the standard, in cold condition before placing the device in the humidity chamber (i.e. one taken for reference purpose only).

In this case I would be suspicious if there is any increase due to the basic physics. Water itself is not a very good conductor, rather it aids conduction by being a good supporter of ions in solution (and the ions do the conducting). Most insulating materials only absorb trace amounts of water and do not release any ions into solution.

However, as mentioned before, some insulating materials such paper (used in some larger transformers) and powder (used in high power heaters) do become highly conductive in the presence of water. These are protected by sealing, and the humidity test is important to verify the effectiveness of sealing. If the sealing is not effective, the presence of water will normally cause catastrophic breakdown, not a gentle increase in leakage.

If leakage currents do really vary by 10% due to moisture, then more investigation seems reasonable, as it suggests some material is not only hygroscopic but also ions are becoming free, which is significant.

Consider repeating the test at 40C (as there is more than 2 times available water than at 23C) and test only neutral open condition which would also ensure repeatability as there is no power to the device (no heating) and have the highest reading to minimize set up errors. You could also test each isolating component separately and see which is causing the increase.
 
Thread starter Similar threads Forum Replies Date
P Humidity Pre-Conditioning before leakage current and dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R Is pre-conditioning supplier component before incoming inspection tests allowed? Quality Manager and Management Related Issues 16
E Humidity Pre-Conditioning IEC 60601-1 (2005) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
B In house NIOSH pre Testing accepted by NIOSH? US Food and Drug Administration (FDA) 1
M AS9100D Registrar pre-audit requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
S Pre-Submission for Breakthrough Device Designation - What's the threshold? US Food and Drug Administration (FDA) 8
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
V Supplier wants to perform pre-qualifications prior to every run instead of locking in parameters. Supplier Quality Assurance and other Supplier Issues 7
F Internal Audit before Pre-Assessment ISO 17025 related Discussions 2
H Queries: RFD & Pre-RFD Program Medical Device and FDA Regulations and Standards News 0
J Design file for pre-existing products - Inputs and Outputs ISO 13485:2016 - Medical Device Quality Management Systems 5
R Gauge R&R on pre-set torque wrench Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D Pre-Production Capability Assesment Reliability Analysis - Predictions, Testing and Standards 5
M Is there any pre-defined Control plan format/template acc. VDA? VDA Standards - Germany's Automotive Standards 0
D Pre-Submission after 510(K) Submission? Other US Medical Device Regulations 4
J Pre-clinical Studies for RUO Product Other Medical Device Regulations World-Wide 1
M Informational Health Canada guidance document – Pre-market Requirements for Medical Device Cybersecurity Medical Device and FDA Regulations and Standards News 0
J Custom-made implants (Class IIb) and pre-commercialization clinical trials in Spain EU Medical Device Regulations 4
C Shipping Medical Devices to US before pre market clearance US Food and Drug Administration (FDA) 1
A UDI requirements for pre-filled combination devices - Box containing a pre-filled device US Food and Drug Administration (FDA) 2
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
D Redacted pre-submission bundle that you would be willing to share US Food and Drug Administration (FDA) 3
I Bench Testing & Pre-verification vs. Formal verification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D FDA Pre-Submission Meeting Advice US Food and Drug Administration (FDA) 0
M Medical Device News Health Canada – Consultation: Pre-market Requirements for Medical Device Cybersecurity Medical Device and FDA Regulations and Standards News 0
L 510 (K) Pre-Sub meeting - Does it worth? US Food and Drug Administration (FDA) 4
A Examples of Pre-Sub, SRD, PMA Shells and Templates Other US Medical Device Regulations 3
B Manufacturing site relocation of pre amendment Class II medical device Manufacturing and Related Processes 4
A Users won't return devices from pre-CE performance evaluation study EU Medical Device Regulations 3
B GM GP-12 Pre-Launch - They asking for a I chart Statistical Analysis Tools, Techniques and SPC 2
GoSpeedRacer Pre-Work and Study for ISO 13485:2016 Lead Auditor Class ISO 13485:2016 - Medical Device Quality Management Systems 1
S Pre-release version of the upcoming AIAG FMEA Revision 5 FMEA and Control Plans 1
V Is 'Compliance' a pre-requisite or is it ok before Batch Release to Market? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
V Writing a Pre-Submission to the FDA 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 47
P EASA STC (Supplemental Type Certificate) on a Pre-mod FAA STC Configuration EASA and JAA Aviation Standards and Requirements 1
H Pre compliance test for Glucometer Other Medical Device and Orthopedic Related Topics 1
V Pre-dispensed API material received in predetermined quantities & given to production Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
somashekar Pre-Announced FDA 4 Day Inspection, Class 2 Devices, Contract Manufacturer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
Q FDA Consensus Standard for PMA (Pre-Market Approval) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
A Does anyone have experience providing a summary of Pre-IDE discussions in the 510k US Food and Drug Administration (FDA) 1
JoCam Format of Health Canada pre-market licence application Canada Medical Device Regulations 3
V 510(K) Cyber Security Documentation for Pre-market Submission (Templates) Other US Medical Device Regulations 6
A Infusion Pump - Pre-Clearance Inspections Learnings US Food and Drug Administration (FDA) 4
Crimpshrine13 ISO/TS 16949 Pre-Audit Requirement IATF 16949 - Automotive Quality Systems Standard 18
Marc USA to insist on pre-flight mobe power probe Travel - Hotels, Motels, Planes and Trains 3
G 510(k) Pre-market approval Application Requirements - Effectiveness of the product Other US Medical Device Regulations 1
K Harmonisation of Global Medical Device Pre-Market Regulations Other Medical Device Regulations World-Wide 5
D Pre-Design Verification - Using Continuous / Variables Data Statistical Analysis Tools, Techniques and SPC 2
A Pre-clinical Animal Studies for long term implantable medical devices EU Medical Device Regulations 5

Similar threads

Top Bottom