Predicate device recalled - Still a valid predicate device?

tehuff

Involved In Discussions
#1
Hi Covers,
My company has selected a predicate device that was just recently recalled (Class 2 recall per FDA) because it doesn't meet its specifications. Unfortunately, much testing has already been done. Can this device still be used as a viable predicate for a 510(k) submission? There is a performance standard for these devices so the plan is to compare the new device to the performance standard and cite the specifications from the 510(k) of the predicate device to demonstrate substantial equivalence. Is this reasonable?
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
I don't think there is a definitive answer for you, or path forward. You may need to request a meeting with FDA to determine how to proceed. They may request a Pre-Sub to define the questions that the meeting is to address.
 

somashekar

Staff member
Super Moderator
#3
Hi Covers,
My company has selected a predicate device that was just recently recalled (Class 2 recall per FDA) because it doesn't meet its specifications. Unfortunately, much testing has already been done. Can this device still be used as a viable predicate for a 510(k) submission? There is a performance standard for these devices so the plan is to compare the new device to the performance standard and cite the specifications from the 510(k) of the predicate device to demonstrate substantial equivalence. Is this reasonable?
For me, the recall is to a specific lot / batch of that device.
This recall does not invalidate the device you predicate.
Its more an issue to that manufacturer than to his device which has a 510k.
A final word with FDA is the way to go, now that you know of the recall.
 
#4
As long as the predicate hasn't been removed from the market by the Commissioner or hasn't been determined to be misbranded or adulterated by a decision of the courts, then the predicate will still be valid.
 
M

MIREGMGR

#5
My take would be that the question is not theoretical...it instead goes to what actually happens when the reviewer group becomes aware of the recall and begins asking questions about the testing process, the device substantial-equivalence comparison, and to what extent the performance discrepancy of the predicate device is potentially applicable to the subject device.

If I were the submitter, I very much would want to discuss those matters with FDA as early as possible, rather than having them be the initiator of the interactions after considerable submission effort had been made.
 

mihzago

Trusted Information Resource
#6
Devices are frequently recalled because they fail to meet the specified performance or safety.
In the 510(k) you compare your new device to a predicate that was submitted to the FDA, not the one that's currently on the market. So unless, as J0anne said, the device hasn't been removed you can still use it; but, be prepared to address questions related to the recall. Better yet, include the cause of the recall of the predicate device in your risk assessment and show how you've addressed it.
 
J

Julie O

#7
If you really mean "recalled," rather than "withdrawn," then, yes, it is still a valid predicate. You will want to pay attention to the reason for the recall, and consider whether your device might have the same problem, or whether it was just due to a goof by the manufacturer of the predicate. For example, the fact that they put the wrong label on a lot of their predicate, doesn't mean anything to you, except as a reminder for you (and all of us), that it's important to assure that you are putting the right labels on your product.
 
#8
This is not a comment on the validity of the predicate - several comments above have covered that.
If there has been any changes as to how the FDA views the predicate or the specific class of device in light of the recall, in terms of the classification of the device itself or the software contained in it, that will affect the new submission.
 
Thread starter Similar threads Forum Replies Date
P Predicate Device Indications for Use US Food and Drug Administration (FDA) 1
S Discontinuation of the Predicate Medical Device Medical Device and FDA Regulations and Standards News 1
Watchcat Does "Similar Device" = "Predicate"? EU Medical Device Regulations 7
S 510k: What to include if your own device is the predicate US Food and Drug Administration (FDA) 5
P Predicate Device Information for 510(k) Submission Medical Device and FDA Regulations and Standards News 4
M Predicate device - Search for a Class IIa device US Food and Drug Administration (FDA) 2
shrutisancheti Predicate device accelerated aging - 1 year of shelf life is already over Other Medical Device Related Standards 1
K Substantial Equivalence Predicate Device Testing Other US Medical Device Regulations 1
R OTC with Rx Predicate - 510(k) for a Class II medical device Other US Medical Device Regulations 6
E 'Special' 510k as a Predicate Medical Device? Other US Medical Device Regulations 3
R Working on a 510(k) that is very similar to the predicate device Other US Medical Device Regulations 4
L Predicate Device for 510(k) - Disposable vs Reusable 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Registering a Class II Medical Device with no Predicate Device - FDA 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Can a Predicate device be a higher class? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
S Predicate Device Requirements - Firmware vs. Software Updates 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S Predicate Device Test Results - ASTM F1717 Static and Dynamic Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
H 510(k) Filing and Predicate Device Clinical Data 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R Which information about predicate device should be included in 510k submission? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
A Characteristics of Predicate Device for 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
M Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Le Chiffre Selecting a predicate device for FDA 510(k) approval Other US Medical Device Regulations 14
S Predicate Material 510(k) US Food and Drug Administration (FDA) 1
S Predicate Material 510(k) Medical Device and FDA Regulations and Standards News 3
B Submit a 510(k) with a Predicate cleared but not Listed US Food and Drug Administration (FDA) 9
M What if information in predicate 510k is obviously wrong? Other US Medical Device Regulations 5
R Identification of Russian Predicate Devices as part of the Registration Process Other Medical Device Regulations World-Wide 1
J ISO 13485 Requirements - Design Exclusion - Predicate Products Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 8
R Reimbursement of an IVD vs. Predicate used in the 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
G Additional Intended Use Predicate 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
B List all 510K numbers on predicate for 510K? Other US Medical Device Regulations 4
M Can a prescription type material be used as a predicate for an OTC material? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
H Existing cloud based medical device - questions regarding improving the processes IEC 62304 - Medical Device Software Life Cycle Processes 0
P Certification process for registered device China Medical Device Regulations 3
Nihls MSA after the device move Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
I How to determine "generic/me too" device? Japan Medical Device Regulations 0
R Debug mode in software/device validation IEC 62304 - Medical Device Software Life Cycle Processes 2
K Unused Service Parts in Newly Manufactured Medical Device? Other US Medical Device Regulations 1
S Determining a device category according to the MDR EU Medical Device Regulations 3
B A.I diagnostic software is considered as medical device in FDA? US Food and Drug Administration (FDA) 5
J Medical device repairs (to upholstery) Manufacturing and Related Processes 4
Ed Panek Apple Provides New Medical Device Cautions 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
shimonv Can you submit a traditional application for a class I device? Other US Medical Device Regulations 7
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 1
DanBOS Cloud Connected Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
dgrainger Informational DRAFT: The Medical Devices (Coronavirus Test Device Approvals) (Amendment) Regulations 2021 UK Medical Device Regulations 0
M Changing Investigational Device from Non-Sterile to Sterile - what are the implications? Other Medical Device and Orthopedic Related Topics 3
B FDA Breakthrough Device can be overlapped with a designated device? US Food and Drug Administration (FDA) 6
D Add new device to current license Canada Medical Device Regulations 0
S Obsolescence of the Medical Device in Various Countries Other Medical Device Related Standards 1
A MDR - Legacy Device Review Timeframe and Requirements EU Medical Device Regulations 3

Similar threads

Top Bottom