Pressure Gauge Calibration
Can I use a hydraulic dead weight tester to calibrate a pressure gauge used to measure air flow. Will there be any performance related issues if we do so?
dwperron
It depends on things like possible contamination of the pressure gauge by hydraulic fluid.
The pressure range being calibrated will also matter, since dead weight testers are usually for high pressures and pressure gauges used in flow measurements tend to be low pressure. The ranges involved may not be compatible.
RW451
Dead weight testers can be used with distilled water, otherwise I agree with dwperron, dead weight testers are typically for high pressure gauges.
Google calibration of manometers.
Example: Calibration of Manometers
Caveat: Never use a dead weight tester that previously used oil. Never, ever use hydrocarbon based oils with oxygen system pressure gauges.
BradM
First, welcome to The Cove!
Second, total agreeance with the other two responses.
I guess I'm intrigued with the phrase "performance related issues". An accuracy verification (hooking up a standard and verifying accuracy) should not impeded how that device works.
When you state pressure gauge, is this the Bourdon tube type with an open port on the bottom?
What range/ accuracy is your pressure gauge?
What range/accuracy is your dead weight tester?
S Pressure Gauge Calibration Uncertainty Calibration and Metrology Software and Hardware 5
V Analog Pressure Gauge Calibration - smallest resolution practice General Measurement Device and Calibration Topics 12
J Pressure Gauge Tolerances after Calibration General Measurement Device and Calibration Topics 5
W Pressure Gauge Calibration Procedure needed General Measurement Device and Calibration Topics 3
H Analog Pressure Gauge with no EN837 accuracy class information General Measurement Device and Calibration Topics 5
N Correct reading for analog pressure gauge ISO 17025 related Discussions 2
N Procedure/calculation for determining the foot pressure of dial thickness gauge General Measurement Device and Calibration Topics 11
J Calibrating an 80,000 psi Fluid Pressure Gauge General Measurement Device and Calibration Topics 6
C How Long it takes to Condition Pressure Gauge and Test Gauge General Measurement Device and Calibration Topics 4
V High Pressure Gauge Transducer (30,000 PSI) / Transmitter Procedures General Measurement Device and Calibration Topics 5
ScottK Gauge R&R on a valve pressure setting system Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P Is it possible to do a Gage R&R Study for a Pressure Gauge? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
C Need help verifying certification for 3000 psi pressure gauge General Measurement Device and Calibration Topics 3
A Gauges on pressure tank - What is the best way to calibrate a psi gauge internally? General Measurement Device and Calibration Topics 6
D How to testing of Pressure relief valve as per API 6D ? Oil and Gas Industry Standards and Regulations 1
J Cleanroom Differential Pressure Monitoring, frequent manual observation US Food and Drug Administration (FDA) 6
Q Autoclave and Sample Container - Pressure-steam cooking's effect Reliability Analysis - Predictions, Testing and Standards 1
O Selection of a calibrator (pressure generator, temperature) for my field devices General Measurement Device and Calibration Topics 2
M Proper ISO standard for Pressure Testing and Leak Rates Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
M Sampling Plan for Alumin High Pressure Die Castings Manufacturing and Related Processes 0
Rincewind Pressure energy - Essential Requirements (MDD/MDR) EU Medical Device Regulations 2
C Pressure requirement of home based Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
R Opening pressure of one-way valve in anaesthetic workstation IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Colin Pressure Vessels for Human Occupation (PVHO) Various Other Specifications, Standards, and related Requirements 4
renenatasha MDD Classification for Pressure Monitoring Set CE Marking (Conformité Européene) / CB Scheme 1
Ajit Basrur MS Excel to study Correlation between Humidity & Differential Pressure Excel .xls Spreadsheet Templates and Tools 7
S What is the minimum pressure differential required for ISO Class 8 Clean Room? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Pressure Equipment Directive (PED) - Fastener questions CE Marking (Conformité Européene) / CB Scheme 4
R Pressure Equipment Directive (PED) Manufacturing and Related Processes 3
M Control Plan MSA's - Pressure Gages IATF 16949 - Automotive Quality Systems Standard 2
R Looking for calibration specification for Fluke 7250 pressure controller General Measurement Device and Calibration Topics 5
A FDA Blood Pressure Monitoring Equipment Measurement Requirements US Food and Drug Administration (FDA) 1
O CE TCF (Technical Construction File) Template for LVD/EMC/Pressure Directive CE Marking (Conformité Européene) / CB Scheme 5
A German MTK (MPBetreibV ? 11) Blood Pressure Meter Requirements EU Medical Device Regulations 2
S Pressure from Directors - confused! ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
O Pressure Equipment Directive PED (97/23/EC) CE Marking (Conformité Européene) / CB Scheme 3
C What would be the Device Classification for a Blood Pressure Monitor in EU ? EU Medical Device Regulations 3
R Applicable Medical Device Standards for Home Use Blood Pressure Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
GURU SDI Measurement - Pressure Drop vs. Leak Rate Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
F Pressure Regulator Testing and Selection Other Medical Device Related Standards 2
R Pressure Test Dial Gage Selection Quality Tools, Improvement and Analysis 1
T Who pays for certification to the PED (Pressure Equipment Directive)? Registrars and Notified Bodies 9
D Calibration and Tolerance Ranges of Pressure Gages General Measurement Device and Calibration Topics 8
M Calibrating Digital Pressure Gages In-House General Measurement Device and Calibration Topics 10
K ASME BPVC (Boiler & Pressure Vessel Code) Calibration Requirements General Measurement Device and Calibration Topics 6
M Pinout of Keller PAA-15 Pressure Transducer General Measurement Device and Calibration Topics 1
P EN 1618 Clarification Re: Catheter Pressure Testing Other ISO and International Standards and European Regulations 4