We are a just in time manufacturer. We satisfy 4.11.2 (f) by accessing current production with a known good gage and infering past production OK. My question is how does this requirement apply to gages used to calibrate gages,etc.? For example, if I have a shadow graph that is calibrated by an outside service, found to be out of calibration, how do I perform an assessment of the validity of previous inspections when I could have used that gage to check many other gages used in production, incoming inspection parts, fixtures etc...how do I handle nonconforming specialized gages like this in terms of requirement 4.11.2 (f)?
Thanks
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
You note the amount of the out of tolerance reading on the gage you sent to the outside vendor. You then use that to determine the amount of induced error in the gages it was used to calibrate. Determine from that if there were any erroneous in-tolerance or erroneous out-of-tolerance readings on those gages. If the accuracy of your standard gage was enough better than those gages it was used to calibrate, you MAY find that there was no implication. But you must determine if there was or was not erroneous results as described above. Hopefully the standard gage was much higher accuracy than those gages it was used to calibrate (at least 4:1 ratio of uncertainty). Final step is to evaluate those gages that were erroneously found in-tolerance or erroneously found out-of-tolerance. Compare the amount of this error with your process or product specifications (hopefully in this case, you had a suitable ratio between the accuracy and stability of the gages that were calibrated by your standard, and the specification limits of your process.

If the erroneous findings on the gages caused erroneous measurement results on your product, you then need to make a formal determination as to if any containment or other corrective actions need to be taken to assure no defective product reaches the customer. And, of course, it all needs to be well documented.

I have tried to walk through the stages of an out-of-tolerance evaluation. This situation underscores the importance of using adequate standards (I don't know whether yours are or not). But the point is to make sure that a standard is of adequate test uncertainty ratio between the standard and the units to be calibrated. This gives you breathing room, so that when a standard is out of tolerance, quite often there will be no impact to product due to the guard band. And the same is the case with measurement tools used for product. It is always wise on any measurement tool that measures an important or critical parameter on your product or process, to assure it is preferrably a little too accurate for the process (using good MSA methodologies as a determinant, and whatever other statistical techniques are appropriate). With these two guard bands in place you minimize the possibilities of things (for example) like having to do a tire recall. I won't throw any stones at that one. But when ever these things happen, they remind those of us in metrology that we have a responsibility that can sometimes effect peoples lives. Sorry... got on my soapbox.

Anyway, hope this is of some help.

------------------
 
D

dhillsburg

#3
Thank you for your answer. I will implement as you descibe. I appreciate your response and it was most helpful.
Diane
 
Thread starter Similar threads Forum Replies Date
E Validity of previous inspections when equipment is found out of calibration Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Gap analysis on ISO 14971:2019 with previous revision ISO 14971 - Medical Device Risk Management 3
W Internal Auditing carried out by a 3rd party - Review of previous audits AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M Informational Commission Regulation (EU) 2019/319 of 6 February 2019 amending previous regulations as regards health certification at import into the Union concerni Medical Device and FDA Regulations and Standards News 0
Sidney Vianna IAQG News Future AS9147- Management of Unsalvageable Parts - Previous scrap parts AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
S MDD in 2018 - Company with no previous EU approval EU Medical Device Regulations 1
S Our products borrow DHF documents from a previous DHF for another product ISO 13485:2016 - Medical Device Quality Management Systems 2
T Technical File - Previous Device Generations EU Medical Device Regulations 2
C Leveraging previous results in IQOQs - Environmentally Controlled Room 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
K Internal Auditing a previous Nonconformance? Internal Auditing 19
N How I should process previous non-conformities on documents or records? ISO 13485:2016 - Medical Device Quality Management Systems 4
N Incompleted tasks from previous management review meeting Management Review Meetings and related Processes 1
Ajit Basrur Medical device recalls in Q2 2013 were 30% higher than previous quarter ! Other US Medical Device Regulations 2
R KPI Review Meetings never discusses previous meeting action plan status Quality Manager and Management Related Issues 5
D Assessing the Validity of Previous Measuring Results? General Measurement Device and Calibration Topics 8
M CQI-11 & CQI-12 Version 2 - Changes from the previous version IATF 16949 - Automotive Quality Systems Standard 3
M ISO 9001 Quality Manual - Is it OK for me to use content from my previous company? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
G A2LA's ISO/IEC 17025 and Accreditation Training - Opinions from previous attendees ISO 17025 related Discussions 2
jasonb067 Applying Previous Failures to new PFMEA APQP and PPAP 2
A Inspection Check and Validation of work from the previous operation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
B Electronic Copies of previous versions of Controlled Documents in editable format Document Control Systems, Procedures, Forms and Templates 3
H Working from a previous design history. How can I best leverage the exisitng DHF? Other US Medical Device Regulations 4
D Beta > 1 for new design? Durability test for a new system and no previous test data Reliability Analysis - Predictions, Testing and Standards 1
J Drawings and previous revisions, control ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Q Looking for a certain article in previous issues of Quality Digest World News 6
Q Registrar Comparison - Searching for previous post related to Registrar Comparison IATF 16949 - Automotive Quality Systems Standard 1
P Disposal of obsolete documents - Mandatory in ISO9001:2000 to keep previous versions? Document Control Systems, Procedures, Forms and Templates 7
R Control of Documents - Change history question - Changes from previous revision OK? Document Control Systems, Procedures, Forms and Templates 11
M 7.6 - Validity of the previous measuring results - Is this what validity means? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
T Review the validity of previous monitoring and measuring results ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
G 7.6 Validity Of Previous Measuring Results General Measurement Device and Calibration Topics 2
Q Previous Sanctioned Interpretations QS-9000 - American Automotive Manufacturers Standard 1
C Calibration Revisited, software too - Previous approach was to calibrate everything General Measurement Device and Calibration Topics 5
P FDA inspections and electronic signatures Other US Medical Device Regulations 2
E South Korea, MFDS - GMP inspections Other Medical Device Regulations World-Wide 4
S Gage R & R for large amount of inspections IATF 16949 - Automotive Quality Systems Standard 10
M Informational USFDA draft guidance – Nonbinding Feedback After Certain FDA Inspections of Device Establishments Medical Device and FDA Regulations and Standards News 0
M Medical Device News TGA Consultation: Good Clinical Practice Inspections Program Medical Device and FDA Regulations and Standards News 0
S FDA Inspections and Pest Control - Pallets too close to walls 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
C How to Control Temporary Inspections Manufacturing and Related Processes 5
G What to Check - Set Up vs. In Process Inspections Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
S Setting up the process - First piece inspections Statistical Analysis Tools, Techniques and SPC 3
T Difference betwee General vs. Follow-up Inspections US Food and Drug Administration (FDA) 3
Q South Africa Clinical Trial Inspections by MCC Quality Manager and Management Related Issues 1
A Infusion Pump - Pre-Clearance Inspections Learnings US Food and Drug Administration (FDA) 4
Q Please suggest how to improve Cosmetic Visual Inspections Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
T Raw Parts Inspections for NPI (New Product Introduction) Phase Supplier Quality Assurance and other Supplier Issues 1
M Validating a Product through Tool Control and Tool Inspections Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
A Can Internal Audit Observations be kept confidential from State FDA inspections ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
G Do Inspections need to be performed by QA/QC Personnel 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5

Similar threads

Top Bottom