Background:
I am a QMS Manager for a small design & build engineering corp. We are dual registered 13485:2003 & 9001:2008. We have a deeply committed management team, and staff who see the benefits to our program and have navigated two complete recertification cycles plus one surveillance audit with only four audit findings issued in that time (one of which was for registrar logo font size). Becuase our program's fundamental structures have not changed, we entered our second surveillance audit for our 3rd certification cycle last week with confidence. Apparently, that was a mistake. Our new auditor (a Ph.D. with four months experience as a registrar auditor) advised me the morning of day two (of a 2 day audit) that we already had 3 minor nonconformances and one major for record keeping. After some argument, he reduced the major to three minors, advised through the day there were two more, then closed the audit with a final issuance of four minors.
His first finding is as follows -
Statement of nonconformity:
Mandatory records are not being established and maintained properly.
Objective evidence:
The internal audit schedule is not treated as a record. [Internal record citations] & [Ref. ISO 13485: 2003 8.2.2; ISO 9001: 2008 8.2.2].
Is there a section of the standard I am missing that requires the audit schedule be treated as a mandatory record?
We revise our audit schedule annually (to review processes), leveraging the schedule as a log (with revision control & write protections). The schedule receives added information noting when an audit is completed for easy tracking of the audit program progress throughout the year. Our documented procedure outlines all of this in detail. I dont mean to be disrespectful of our auditor, but I do not see where he is coming from on this finding? Could he be misreading the requirement or laying on the 13485:2016 requirements for records?
Does anyone have any insight? Should I complain or appeal?
Our audit ran long and he didn't have time to discuss the findings during his closing which is why I am posting this now.
:confused::confused::confused:
Marcelo
Re: Audit help - please
Do you want the conceptual answer or the practical one?
Here's both, anyway.
- 8.2.2 Internal audit requires that "The organization shall conduct internal audits at planned intervals"... and "The responsibilities and requirements for planning and conducting audits, and for reporting results and maintaining records (see 4.2.4) shall be defined in a documented procedure." So, in principle it would depend on what your procedure say.
- However, the audit schedule is obviously part of the requirement for planned audits, and a schedule would not in principle be a document, it would better fit as a record (as you probably use a template for scheduling each audit), so it does make sense that it needs to be controlled.
Anyway, the Statement of nonconformity is a little misleading because there's no specific mandatory record requirement in 8.2.2. It would be better if he called 4.2.4 Control of records.
Randy
If he waited until the 2nd day to inform you of a major and the minors he found on the 1st day he's wrong, totally and absolutely.
As an instructor of over 200 Lead Auditor courses, a qualifying reviewer for a CB and over 15 years of 3rd party auditing I'll stand my ground on this one, he's wrong and he's resting on his PhD to avoid being challenged....File a complaint and demand that he never revisit. If he did this on a team of mine I'd rip his knickers for poor audit conduct.
W
WilBryan
Marcelo & Randy,
thanks to both of you for the insight!
Randy - I kind of felt the same way about it and am just figuring out the next steps...
Marcelo -
You cite that the schedule would be a better fit as a record. With respect to your experience and opinion, based on the standard's text in 8.2.2 isn't this our decision to make? If we define clearly what we do regarding our audit schedule (and other audit requirements) and we adhere to those documented procedures, shouldn't we be compliant? As I read it, there is no mandatory audit record required in section 8.2.2 until 13485:2016.
Marcelo
There's what I said is the conceptual answer.
However, there's other requirements that make it more difficult to justify.
For example, the requirement on record mention that:
"Records shall be established and maintained to provide evidence of conformity to requirements and of the effective operation of the quality management system."
8.2.2, as I mentioned before, says:
The question then would be, how do you who's evidence that you are conducting internal audits at planned intervals?
If you have another way of showing the evidence (and again, this would need to be a record by the same record definition), then you would comply. If not, it's difficult to say that you are compliant.
Coury Ferguson
Moderator here to help
Marcelo & Randy,
Marcelo -
In my opinion, the schedule is not a record. It is more of a living document that changes based upon importance. However, the results of that audit are records.
I agree with Randy in regards to the auditor trying to use their degree to make an outcome of the results without challenge. Go back to the Registrar challenge their findings, and ask for another Assessor for the next go around.
Even though the Registrar is a 3rd party that verifies that the system meets requirements at the time of the audit, you are still a client of theirs. In the next future audit, ask to see the resumes/CVs of the auditor(s) they feel would make a good fit to your organization.
Sidney Vianna
Post Responsibly
As others already opined, the two standards in question do not require the internal audit schedule to be treated as records and subjected to the requirements of 4.2.4. So, that specific NC should be appealed, in my view.
Since the new auditor admitted being new to this, I would give him the benefit of the doubt and not necessarily require a re-assignment of auditors, if you feel he was objective, balanced and fair. Looking back at my early days of auditing, I now admit having made bad calls too. Who doesn't?
somashekar
Marcelo & Randy,
Marcelo -
What is your larger purpose .. To confront OR To continually improve.
If its to improve, then Marcelo has shown the way. If I were you, I would never go by the NC count. Your good performance is never tainted with few NC. I would see this as an opportunity to refine the documented procedure and records requirement. Good Luck ~~~
Mike S.
Happy to be Alive
The auditor was wrong to write a NC on this – there is no shall being violated.
That being said, we keep our audit schedules as records because it serves us well to do so. It makes it easy to provide objective evidence to all auditors of many aspects of 8.2.2 and it costs virtually nothing to do so.
Nadaabo
The auditor should plan accordingly to give time to review all the findings with you.
I have never had any auditor challenge me on the audit schedule being a record. We have a documented procedure that lists all the records we keep and how long.
Your Registar should have a process for appeal.
But did you find that any of his findings were going to help you improve your system? if so, then an NC is an opportunity for improvement, if not, then I would talk to the sales Rep of your registar and explain to them that your auditor was not value added.
H Problems with RABQSA - Applying for my Provisional Auditor (PA) status ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 3
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Problems with process equipment ISO 13485:2016 - Medical Device Quality Management Systems 4
A Feedback of typical maintenance problems with LMI digital probes Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
S New problems exporting to Germany - February 2019 ISO 13485:2016 - Medical Device Quality Management Systems 7
C Resource allocation problems in Software DFMEA FMEA and Control Plans 4
R Tapping problems - 3/8" hot roll pickled & oiled steel - seeking expertise Manufacturing and Related Processes 15
Marc Problems - Contacting Us about Problems (How To - Etc.) Elsmar Xenforo Forum Software Instructions and Help 0
Marc Youtube - Viewing problems Elsmar Xenforo Forum Software Instructions and Help 4
O How do reduce the risk of suppliers having similar problems? Supplier Quality Assurance and other Supplier Issues 4
L Problems while documenting the SOUPs used for the software we are developing IEC 62304 - Medical Device Software Life Cycle Processes 4
L Relying on Suppliers - Most Quality and Warranty problems are caused by Suppliers Supply Chain Security Management Systems 5
M The biggest gaps and problems in QMS or Quality in Automotive Misc. Quality Assurance and Business Systems Related Topics 3
I Request an FDA GUDID Account - Problems Other US Medical Device Regulations 3
S Document Review Problems ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
L How to avoid Nonconformance Reoccurrence problems IATF 16949 - Automotive Quality Systems Standard 20
C Software Process Implementation Problems ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
K Bottle Labeling Machine Problems - Non conformance in FCMG industry Nonconformance and Corrective Action 8
T Contacting the Supplier's CB vs a Supplier Audit - Nonconformance Problems AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Peter Fraser Problems and Deficiencies with the ISO 9001:2015 Draft ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
B Quality problems that cost Millions of dollars (or euros, or pounds, etc.) Quality Manager and Management Related Issues 7
AnaMariaVR2 How Open Innovation is Solving Some of NASA?s Trickiest Problems World News 0
T A survey on problems during ISMS implementation - need help IEC 27001 - Information Security Management Systems (ISMS) 2
S Common Electric Fan Motor Problems Reliability Analysis - Predictions, Testing and Standards 3
M Problems with Competence/Skills Matrix (6.2 Human Resources) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
ScottK Is anyone else having problems uploading Proprietary Names into the DRLM? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
M FDA Warning Letter - Product Association with Search Engine Results - Problems Ahead? US Food and Drug Administration (FDA) 10
M Problems with Unrealistic Customer Expectations Customer Complaints 12
AnaMariaVR2 Chronic Regulatory Compliance Issues Causing Drug Shortages & Safety Problems Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
Q Method on how to cause people to solve problems Benchmarking 5
O Aerospace Registrar Problems - Starting Over? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
M 8.2.4 Problems with Monitoring and Measurement of Product AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R Problems with Supplier - Always with 100% Sorting - No Corrective Actions Supplier Quality Assurance and other Supplier Issues 22
G Product Returns are above average for Labeling Problems Quality Manager and Management Related Issues 4
V Supplier's First Run had Plating Problems - Can they Strip, Re-Plate and PPAP them? APQP and PPAP 14
1 A New Job - Replaced a QM who passed away - BIG Problems Quality Manager and Management Related Issues 25
N English as a Second Language - English problems with Legal Terms Contract Review Process 2
Uriel Alejandro Problems with the Employee Skills Matrix - Document Control System Quality Manager and Management Related Issues 8
L Avoiding Recurrence of the Supplier?s Quality Problems Supplier Quality Assurance and other Supplier Issues 5
M Problems implementing ISO 13485 for Software-Only Medical Device Manufacturers? ISO 13485:2016 - Medical Device Quality Management Systems 4
B How can I resolve Chronic Nonconformance Problems? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
S Problems with the application of EN 62366:2008 IEC 62366 - Medical Device Usability Engineering 13
F Can you tell me if this table is correct? Gage R&R Problems Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
D Quality Metrics to use to solve Problems Quality Tools, Improvement and Analysis 5
P Self Motivation - How to keep it when overwhelmed with problems? Career and Occupation Discussions 13
S Systemic Problems Identified in Audit and How to Fix Them IATF 16949 - Automotive Quality Systems Standard 10
T Lollipop Production Line Problems - Defectives and Root Cause Analysis Reliability Analysis - Predictions, Testing and Standards 6
T Big problems with Inspection - Manual Operation - Needs experts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 27
C Efficient Supplier Development and Problems with Buyers Supplier Quality Assurance and other Supplier Issues 5
V Common Coating Problems in Refinery Manufacturing and Related Processes 1