Procedure/calculation for determining the foot pressure of dial thickness gauge

N

nurhakim

#1
Hi,

I am looking on procedure/calculation for determining the foot pressure of dial thickness gauge as well as the accepted tolerance.

Anyone can help? Thanks in advance.

Regards,
Nur
 
Elsmar Forum Sponsor

Jim Wynne

Staff member
Admin
#2
Re: Foot pressure of dial thickness gauge

Hi,

I am looking on procedure/calculation for determining the foot pressure of dial thickness gauge as well as the accepted tolerance.

Anyone can help? Thanks in advance.

Regards,
Nur
I've moved your post to a more appropriate forum.

A picture of what's commonly known as a dial thickness gage is attached. Is this the type of device you're asking about? If not, could you please explain further? If it is what you're asking about, I'm not sure how foot pressure comes into play.
 

Attachments

N

nurhakim

#3
Re: Foot pressure of dial thickness gauge

Dear Jim,

many thanks for your quick response. The picture that you attached is what I am looking for. Actually I receive an NCR from previous audit regarding this. In the workplace, I measure the thickness of film using this device. The auditor was concern about the correct mesurement and to take into consideration the foot pressure when conducting the measurement for more accurate result.
 

Jim Wynne

Staff member
Admin
#4
Re: Foot pressure of dial thickness gauge

Dear Jim,

many thanks for your quick response. The picture that you attached is what I am looking for. Actually I receive an NCR from previous audit regarding this. In the workplace, I measure the thickness of film using this device. The auditor was concern about the correct mesurement and to take into consideration the foot pressure when conducting the measurement for more accurate result.
The word "foot" is what's puzzling me. Are you referring to the anvils--the surfaces of the gage that contact the part? The pressure isn't usually adjustable, and if the gage has been properly calibrated and is being used properly, pressure shouldn't be an issue.

Please correct me if I'm wrong in my interpretation.
 
T

True Position

#6
Re: Foot pressure of dial thickness gauge

The word "foot" is what's puzzling me. Are you referring to the anvils--the surfaces of the gage that contact the part? The pressure isn't usually adjustable, and if the gage has been properly calibrated and is being used properly, pressure shouldn't be an issue.

Please correct me if I'm wrong in my interpretation.
I'm fairly sure he's talking about the anvils. You're making a number of assumptions about calibration and proper use. I wouldn't be surprised at all to hear the gage is set using blocks or a hard pin and then used to measure a soft film. In that case, the force applied would be a concern.

The real problem is using the wrong gage in the first place.
 

harry

Super Moderator
#7
Re: Foot pressure of dial thickness gauge

......................... The real problem is using the wrong gage in the first place.
I agree. It is important to understand your purpose, your material and any standards that you need to meet in choosing the right gage. Example:

for textile, you need to meet ASTM D 1777 and for

elastomer or rubber, ASTM D3767

Details are given in the various tables of the standards.
 
N

nurhakim

#8
Re: Foot pressure of dial thickness gauge

Dear Jim,

Yes I am referring to the anvil. The concern is that how much pressure applied during measurement that would affect the accuracy. Due to some reason that one may apply more pressure.
 

Jim Wynne

Staff member
Admin
#9
Re: Foot pressure of dial thickness gauge

Dear Jim,

Yes I am referring to the anvil. The concern is that how much pressure applied during measurement that would affect the accuracy. Due to some reason that one may apply more pressure.
MSA, and in particular GR&R would be helpful in providing evidence of measurement consistency.

If the gage is being properly used there should be no pressure being applied by the user; the device's spring force should be sufficient. it's not like a micrometer or caliper where closing pressure can affect results. It seems an odd thing for an auditor to be concerned about unless there is evidence of misuse of the gage.
 
I

iamtroll

#10
I would consult the manufacturer's published load claim first in order to determine what range will be needed on the device with which to evaluate that load on your tool. In all probability, a Chatillon force gage of the correct range would be a reasonable device to validate that your device operates within the published force. From there, I would do a combined Gage R&R on both the force gage and the dial gage to prove that you can check the force of that gage and also that you are measuring the thickness of the film repeatably. You will then need to evaluate the compressive strength of the film to assure that you are not compressing it with the force that you now know that you are applying with the original tool. A little convoluted, but I hope that helps.
 
Thread starter Similar threads Forum Replies Date
B Procedure packs with non-medical devices EU Medical Device Regulations 1
K Need procedure for D&D inputs? ISO 13485:2016 - Medical Device Quality Management Systems 4
J Example of a defined procedure for carrying out Material Review Board (MRB) Manufacturing and Related Processes 0
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
A SPC procedure / flow chart Statistical Analysis Tools, Techniques and SPC 2
J Action Not defined in procedure - NC - Incoming inspection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Procedure ( SOP) for Device Master Record ( DMR ) and for Device History Record (DHR)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
shimonv Document Control Procedure Header Content Document Control Systems, Procedures, Forms and Templates 3
U Procedure pack components EU Medical Device Regulations 3
G Control Plan & PFMEA Review Procedure? FMEA and Control Plans 1
J Strategy for MDR Regulatory Compliance Procedure ISO 13485:2016 - Medical Device Quality Management Systems 4
A IVD in MDR procedure pack EU Medical Device Regulations 2
G RoHS Procedure REACH and RoHS Conversations 2
John Predmore Configuration Management as a process instead of a procedure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
L How to add exemption or statement to control of document procedure? ISO 13485:2016 - Medical Device Quality Management Systems 5
A Design and development procedure for API Spec Q2 Oil and Gas Industry Standards and Regulations 6
W Using tailoring guidelines to tailor a QMS procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Change management procedure when 7.3 is not applicable ISO 13485:2016 - Medical Device Quality Management Systems 5
L Sample Procedure from Contract Manufacturer Prospective ISO 14971 - Medical Device Risk Management 1
T Internal Nonconformance procedure thoughts (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Anonymous16-2 SOP (Standard Operating Procedure) Numbering Document Control Systems, Procedures, Forms and Templates 7
M UDI procedure/ report 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
DuncanGibbons Resources for aiding in procedure, work instruction and manufacturing plan development and management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Article 22 Procedure Pack Labelling EU Medical Device Regulations 4
N EU MDR - Applicability Article 22 Systems and Procedure Packs CE Marking (Conformité Européene) / CB Scheme 4
M Example of statement for procedure pack MDR Article 22? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
R Procedure, Frequency and Acceptance Criteria for Replicate, Recalibration, Before-After and Intermediate Checks ISO 17025 related Discussions 8
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 0
M MDR - Is a formal GSPR Procedure required? EU Medical Device Regulations 20
Moncia Production Planning procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Q Tracking Procedure Revisions (Document Control) Document Control Systems, Procedures, Forms and Templates 9
L Purchasing procedure and workload ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Roles and accountabilities procedure? ISO 13485:2016 - Medical Device Quality Management Systems 9
K Procedure equals a process? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
M EU MDR - RA procedure for EU EU Medical Device Regulations 5
F Procedure for Device End of Life/Obsolescence Other Medical Device and Orthopedic Related Topics 3
A MAH license transfer (medical device registrant) procedure in Indonesia Other US Medical Device Regulations 0
F ISO 13485 8.2.3 Reporting to regulatory authorities: Question regarding a procedure for this clause. ISO 13485:2016 - Medical Device Quality Management Systems 4
J Manufacturing procedure / work instruction depth ISO 13485:2016 - Medical Device Quality Management Systems 12
F Hi friends, can anyone show me an example of a procedure for ISO 13485 6.4.1 Work Environment? ISO 13485:2016 - Medical Device Quality Management Systems 2
U Product Lifecycle Procedure Guidance Design and Development of Products and Processes 7
V Datron 4707 or 4708 33K, NAVAIR or TB cal procedure General Measurement Device and Calibration Topics 1
R MDR Article 22: Systems and procedure packs Other Medical Device Regulations World-Wide 12
I Containment procedure for parts after a power outage. IATF 16949 - Automotive Quality Systems Standard 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
H Statistical Techniques Procedure - What should be included Document Control Systems, Procedures, Forms and Templates 4
E AS9100 installtion procedure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Procedure on Privacy Policy in the ISO 13485 quality management system ISO 13485:2016 - Medical Device Quality Management Systems 3

Similar threads

Top Bottom