SBS - The Best Value in QMS software
Procedure, Frequency and Acceptance Criteria for Replicate, Recalibration, Before-After and Intermediate Checks
#1
Hello Everyone,
In my calibration laboratory, we are required to carry out replicate calibration, recalibration of retained items, intermediate checks and in-out / before-after / inservice checks as per Clause 7.7 of ISO/IEC 17025:2017. Could someone out there guide me regarding its detailed procedure, frequency and acceptance criteria for these checks? Or maybe point me to a standard?
Thanks in advance.
Peace.
Richie Rich
Elsmar Forum Sponsor
dwperron
Trusted Information Resource
#2
The criteria for the frequency will be based on risk to your process - what is the risk of a particular standard being out of tolerance to your process? If dealing with a high use standard like a multifunction calibrator there is a lot more risk than there is with a gauge block, so the frequency would be higher.
The detailed procedure is one you must come up with, as you alone can determine what is important for your lab to be checking based on risk to your processes. You go after the items that can cause the most trouble for you if they are out of tolerance, and you figure what level of error is "acceptable" when you set your acceptance criteria.
Subhasis Mallik
Starting to get Involved
#3
Enclosing a reference document, which may be useful for you to decide on these aspects.
Thanks.
Moderator Note: A copyright-protected document was removed. It carried this warning in its copyright notice:
"...you must not modify, copy, reproduce, republish, frame, upload to a third party, store in a retrieval system, post, transmit or distribute this content in any way or any form or by any means without express written authority from NATA. "
#4
Hello @Subhasis Mallik The moderator has removed the reference document due to copyright issue. Could you please name the reference document, so I could look it up?
Thanks in advance.
Peace.
Richie Rich
Subhasis Mallik
#5
Actually this document has no copyright issue since it is available in Australian Accreditation Agency called NATA (National Association of Testing Authorities, Australia) as free downloads for anybody.
Since I can not put any link so far, you may help the members here by putting the link.
Search for NATA Australia in Google search and go the home page of National Association of Testing Authorities, Australia.
Then go to the "Accreditation Information" in main menu.
Then look at the sub-menu list at left and go to "Accreditation Criteria and Guidance" block and then further to "OECD Principles of Good Practices" and then you will see list of lots of downloadable materials.
In that list look for BOLD HEADER as "GENERAL ACCREDITATION GUIDANCE" and the document name below that as "General Equipment Table".
You will be able to download the document which is to be used as a guide only by laboratory, with necessary validation or justification.
Thanks for your interest.
Best regards.
Subhasis
#7
Best regards.
Subhasis
Hello @ Subhasis Mallik Thanks for the clear instructions. I found the document and it was useful for me.
Jim Wynne
Staff member
Admin
#8
Please read again the bit I quoted earlier from the document:
While the document might be available for free downloading for personal use, it doesn't mean that it can be posted in a public forum.
Thread starter Similar threads Forum Replies Date
B Process / Procedure - Radiographic (X-Ray) Non-Film Document Control Systems, Procedures, Forms and Templates 0
M Procedure for vigilance system according to new MDR EU Medical Device Regulations 0
M Procedure for clinical evaluation according to new MDR EU Medical Device Regulations 1
S Production Work Order Procedure 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
E System&Procedure pack or accessory EU Medical Device Regulations 0
K Contract Manufacturer Do they need a complaint procedure? Medical Device and FDA Regulations and Standards News 8
E Design and Development file Procedure ISO 13485:2016 - Medical Device Quality Management Systems 0
B Procedure facility license in karnataka Manufacturing and Related Processes 3
S MDR - System and procedure pack article 22 and all sub processes that apply ISO 13485:2016 - Medical Device Quality Management Systems 0
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 6
B Procedure packs with non-medical devices EU Medical Device Regulations 1
K Need procedure for D&D inputs? ISO 13485:2016 - Medical Device Quality Management Systems 4
J Example of a defined procedure for carrying out Material Review Board (MRB) Manufacturing and Related Processes 0
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
A SPC procedure / flow chart Statistical Analysis Tools, Techniques and SPC 2
J Action Not defined in procedure - NC - Incoming inspection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Procedure ( SOP) for Device Master Record ( DMR ) and for Device History Record (DHR)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
shimonv Document Control Procedure Header Content Document Control Systems, Procedures, Forms and Templates 4
U Procedure pack components EU Medical Device Regulations 3
G Control Plan & PFMEA Review Procedure? FMEA and Control Plans 8
J Strategy for MDR Regulatory Compliance Procedure ISO 13485:2016 - Medical Device Quality Management Systems 4
A IVD in MDR procedure pack EU Medical Device Regulations 2
G RoHS Procedure REACH and RoHS Conversations 2
John Predmore Configuration Management as a process instead of a procedure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
L How to add exemption or statement to control of document procedure? ISO 13485:2016 - Medical Device Quality Management Systems 5
A Design and development procedure for API Spec Q2 Oil and Gas Industry Standards and Regulations 6
W Using tailoring guidelines to tailor a QMS procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Change management procedure when 7.3 is not applicable ISO 13485:2016 - Medical Device Quality Management Systems 5
L Sample Procedure from Contract Manufacturer Prospective ISO 14971 - Medical Device Risk Management 1
T Internal Nonconformance procedure thoughts (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Anonymous16-2 SOP (Standard Operating Procedure) Numbering Document Control Systems, Procedures, Forms and Templates 7
M UDI procedure/ report 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
DuncanGibbons Resources for aiding in procedure, work instruction and manufacturing plan development and management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Article 22 Procedure Pack Labelling EU Medical Device Regulations 4
N EU MDR - Applicability Article 22 Systems and Procedure Packs CE Marking (Conformité Européene) / CB Scheme 4
M Example of statement for procedure pack MDR Article 22? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 0
M MDR - Is a formal GSPR Procedure required? EU Medical Device Regulations 21
Moncia Production Planning procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Q Tracking Procedure Revisions (Document Control) Document Control Systems, Procedures, Forms and Templates 9
L Purchasing procedure and workload ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Roles and accountabilities procedure? ISO 13485:2016 - Medical Device Quality Management Systems 9
K Procedure equals a process? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
M EU MDR - RA procedure for EU EU Medical Device Regulations 5
F Procedure for Device End of Life/Obsolescence Other Medical Device and Orthopedic Related Topics 3
A MAH license transfer (medical device registrant) procedure in Indonesia Other US Medical Device Regulations 0
F ISO 13485 8.2.3 Reporting to regulatory authorities: Question regarding a procedure for this clause. ISO 13485:2016 - Medical Device Quality Management Systems 4
J Manufacturing procedure / work instruction depth ISO 13485:2016 - Medical Device Quality Management Systems 12
Similar threads
Top Bottom