Process Qualification - Heat Treatment Process Furnace - What is Thermal Mapping?

D

dr madhavan

#1
What is Thermal Mapping? Are there any standards or codes of practices for conducting this with reference to the Heat Treatment Process Furnaces? How frequently is it to be conducted?
Is it to be comparable to 'calibration' requirements?
 
Elsmar Forum Sponsor

Geoff Cotton

Quite Involved in Discussions
#2
Are you taking about establishing the temperature profile / heat zones within a belt furnace?

We check our belt furnace using a thermocouple a couple of times a year, compare the results to the temperature charts and previous readings.


Hope that helps
 
D

dr madhavan

#3
Process Qualification

I am referring to 'Measurement of uniformity of Temperature across the cross-section as well as in the depath, particularly in 'PIT TYPE' Heat Treatment Furnaces -say, a Gas Carburising Furnace or even a Salt bath furnace.
 
L

lday38

#4
Qualification - Heat Treatment Process Furnace - What is Thermal Mapping?

The thermocouples then are not used every time but are used to verify that the heat is consistent throughout the furnace. Is that correct?
When you say you compare them to the temperture, I believe your saying the gage outside of the furnace.
Have you ever heard of someone using the thermocouples for each lot during processing?
Where can I find out more about process controls for the heattreating process?
I understand that some suppliers use graphs? I would suppose the graph is a picture over time. Is that correct?

Thanks
Linda
 

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#5
Re: Qualification - Heat Treatment Process Furnace - What is Thermal Mapping?

The thermocouples then are not used every time but are used to verify that the heat is consistent throughout the furnace. Is that correct?
It depends upon what your purpose is. I have seen processes where thermal profile devices go through with every lot, but that's not typical.

Thermal Mapping is taking an oven or furnace and profiling the temperature at different locations. If it's a long furnace line, you would choose multiple 'essential' points and monitor the temperature at each point. If it's a temperature chamber (such as a closed oven) you measure different points within the oven.

Some threads here: Thermal Profiling

Some other Thermal Profile resources.
 
L

lday38

#6
Thank you , Marc
Yes, they are putting the thermocuple (3 of them) through each lot. The product is a shaft for a blender blade. Having never seen a thermocouple, I would think that it is recording the temperture in the differant locations. Right?
I am hearing that we will get charged more for charts , whch makes some sence. However, I would think they are recording data off these thermocouples that should be available for past lots if there is an issue.
Linda
 

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#7
Are the 'thermocouples' traveling on the belt through the furnace, or are they measuring 3 points in an oven (i.e.: three places along the oven, as I assume this is a long belt oven)?

A thermocouple is two pieces of special wire which are welded together. There are different types each of which is for a specific temperature range. The ends of these wires are attached to a measurement device which measures electrical potential. Some people call the wire and the measurement device a 'thermocouple', but I'm from the old days where a thermocouple was the welded wire junction.

Wikipedia reference-linkThermocouple

There are devices which are a compact, single 'box' which can be run through an oven with a lot which internally records, or transmits to an external analyzer, the temperature as the device proceeds through an oven with a lot.

I need more specific information on what you're referring to - Spot measurements at places in an oven, or a device which actually goes through the oven with the lot.

** I assume this is a belt oven (long oven). Or, is this one big stationary oven?
 

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#8
From an email: Hi there
Big statinary oven , atmospherically controlled.

Linda


In those you decide where you want thermocouples, but again it depends upon the oven design. Is it gas or electric? Is there an air mover which keeps air circulating or is it a 'static' oven?

I'm not sure what you mean by atmospherically controlled.
 
Thread starter Similar threads Forum Replies Date
V Process, component or full product re-qualification: leaded to unleaded solder Qualification and Validation (including 21 CFR Part 11) 8
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
B VDA 6.3 Qualification as Process Auditor training course and exam VDA Standards - Germany's Automotive Standards 0
K Is Mold qualification / process validation always required? Manufacturing and Related Processes 7
J Simple supplier evaluation qualification process form that includes Risk Assessment Document Control Systems, Procedures, Forms and Templates 2
S Qualification Builds & Process Changes (AS9102 - First Article Inspection) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
C Process Qualification Sample size Document Control Systems, Procedures, Forms and Templates 6
S Must I document qualification(s) or qualification process for Internal Auditors? IATF 16949 - Automotive Quality Systems Standard 10
P Approved Suppliers - ISO 13485 Normal Supplier Qualification Process Requirements ISO 13485:2016 - Medical Device Quality Management Systems 9
P Process or Performance Qualification Requirements Qualification and Validation (including 21 CFR Part 11) 5
M Process Qualification - What is the recommended Sample Size for PQ? Qualification and Validation (including 21 CFR Part 11) 10
R Process Qualification, Equipment Qualification, Operator Qualification - Sequence? Manufacturing and Related Processes 10
M Process Validation and Qualification of Legacy Equipment ISO 13485:2016 - Medical Device Quality Management Systems 10
R Line Leader Assessment form - Qualification process for production line leader Training - Internal, External, Online and Distance Learning 1
M Process Validation/Equipment Qualification - Class 1A (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 12
S Operational Qualification for injection molding process Qualification and Validation (including 21 CFR Part 11) 2
M PQ and PPQ - Process Qualification for medical device manufacturing ISO 13485:2016 - Medical Device Quality Management Systems 1
S Example Flow Chart of Qualification of Process wanted APQP and PPAP 3
Y NPI (New Product Introduction) Process Qualification System Required Document Control Systems, Procedures, Forms and Templates 8
R Gage Tolerance vs. Process Parameter Tolerance - Process qualification Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Q Determining the Appropriate Sample Size for Process Validation/Qualification Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
J Product/Process Qualification Plan Manufacturing and Related Processes 6
B Toyota PPAP Process - Three Questions APQP and PPAP 3
R Changes vs CMO - How can we simplify this process? Supplier Quality Assurance and other Supplier Issues 3
A Ethics Committee Review Process for IVD Products EU Medical Device Regulations 2
V Laser Welding Process - Impact on Electrical Properties Reliability Analysis - Predictions, Testing and Standards 4
Q Process: Knowledge Section 7.1.6 of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
L Documented Information in Internal Audits Process (9.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A Sampling plan for in-process QC (medical devices) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
R MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
C In-process inspection - Tooling and assembly lines for automotive companies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Efficacy of an IT process after a cyber attack ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
N Riveting - special process Manufacturing and Related Processes 11
M Material incoming to the production process reflected in PFMEA FMEA and Control Plans 9
A API Spec Q1 Purchasing Process - Supplier Reevaluation based on Supplier Risks 5.6.1.4 Oil and Gas Industry Standards and Regulations 15
B Handling lower detection limits for SPC and process performance Statistical Analysis Tools, Techniques and SPC 1
D Measurables for Plastic Injection molding process Manufacturing and Related Processes 1
S Cleaning process center change ISO 13485:2016 - Medical Device Quality Management Systems 4
Z Rapid audit template for plastic parts manufacturing process Manufacturing and Related Processes 12
R Inspection and Work order process Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
A How to reduce the process SPC monitoring Capability, Accuracy and Stability - Processes, Machines, etc. 3
John Predmore Configuration Management as a process instead of a procedure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
U Internal Auditor not trained but done Audit for some process Nonconformance and Corrective Action 5
B Two excellent examples of process capability analysis from Quality Magazine Capability, Accuracy and Stability - Processes, Machines, etc. 5
D ECO (Engineering Change Order) process questions ISO 13485:2016 - Medical Device Quality Management Systems 7
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3

Similar threads

Top Bottom