Process Qualification - Heat Treatment Process Furnace - What is Thermal Mapping?

D

dr madhavan

#1
What is Thermal Mapping? Are there any standards or codes of practices for conducting this with reference to the Heat Treatment Process Furnaces? How frequently is it to be conducted?
Is it to be comparable to 'calibration' requirements?
 
Elsmar Forum Sponsor

Geoff Cotton

Quite Involved in Discussions
#2
Are you taking about establishing the temperature profile / heat zones within a belt furnace?

We check our belt furnace using a thermocouple a couple of times a year, compare the results to the temperature charts and previous readings.


Hope that helps
 
D

dr madhavan

#3
Process Qualification

I am referring to 'Measurement of uniformity of Temperature across the cross-section as well as in the depath, particularly in 'PIT TYPE' Heat Treatment Furnaces -say, a Gas Carburising Furnace or even a Salt bath furnace.
 
L

lday38

#4
Qualification - Heat Treatment Process Furnace - What is Thermal Mapping?

The thermocouples then are not used every time but are used to verify that the heat is consistent throughout the furnace. Is that correct?
When you say you compare them to the temperture, I believe your saying the gage outside of the furnace.
Have you ever heard of someone using the thermocouples for each lot during processing?
Where can I find out more about process controls for the heattreating process?
I understand that some suppliers use graphs? I would suppose the graph is a picture over time. Is that correct?

Thanks
Linda
 

Marc

Fully vaccinated are you?
Leader
#5
Re: Qualification - Heat Treatment Process Furnace - What is Thermal Mapping?

The thermocouples then are not used every time but are used to verify that the heat is consistent throughout the furnace. Is that correct?
It depends upon what your purpose is. I have seen processes where thermal profile devices go through with every lot, but that's not typical.

Thermal Mapping is taking an oven or furnace and profiling the temperature at different locations. If it's a long furnace line, you would choose multiple 'essential' points and monitor the temperature at each point. If it's a temperature chamber (such as a closed oven) you measure different points within the oven.

Some threads here: Thermal Profiling

Some other Thermal Profile resources.
 
L

lday38

#6
Thank you , Marc
Yes, they are putting the thermocuple (3 of them) through each lot. The product is a shaft for a blender blade. Having never seen a thermocouple, I would think that it is recording the temperture in the differant locations. Right?
I am hearing that we will get charged more for charts , whch makes some sence. However, I would think they are recording data off these thermocouples that should be available for past lots if there is an issue.
Linda
 

Marc

Fully vaccinated are you?
Leader
#7
Are the 'thermocouples' traveling on the belt through the furnace, or are they measuring 3 points in an oven (i.e.: three places along the oven, as I assume this is a long belt oven)?

A thermocouple is two pieces of special wire which are welded together. There are different types each of which is for a specific temperature range. The ends of these wires are attached to a measurement device which measures electrical potential. Some people call the wire and the measurement device a 'thermocouple', but I'm from the old days where a thermocouple was the welded wire junction.

Wikipedia reference-linkThermocouple

There are devices which are a compact, single 'box' which can be run through an oven with a lot which internally records, or transmits to an external analyzer, the temperature as the device proceeds through an oven with a lot.

I need more specific information on what you're referring to - Spot measurements at places in an oven, or a device which actually goes through the oven with the lot.

** I assume this is a belt oven (long oven). Or, is this one big stationary oven?
 

Marc

Fully vaccinated are you?
Leader
#8
From an email: Hi there
Big statinary oven , atmospherically controlled.

Linda


In those you decide where you want thermocouples, but again it depends upon the oven design. Is it gas or electric? Is there an air mover which keeps air circulating or is it a 'static' oven?

I'm not sure what you mean by atmospherically controlled.
 
Thread starter Similar threads Forum Replies Date
S Performance Qualification and Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
V Process, component or full product re-qualification: leaded to unleaded solder Qualification and Validation (including 21 CFR Part 11) 8
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
B VDA 6.3 Qualification as Process Auditor training course and exam VDA Standards - Germany's Automotive Standards 0
K Is Mold qualification / process validation always required? Manufacturing and Related Processes 7
J Simple supplier evaluation qualification process form that includes Risk Assessment Document Control Systems, Procedures, Forms and Templates 2
S Qualification Builds & Process Changes (AS9102 - First Article Inspection) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
C Process Qualification Sample size Document Control Systems, Procedures, Forms and Templates 6
S Must I document qualification(s) or qualification process for Internal Auditors? IATF 16949 - Automotive Quality Systems Standard 10
P Approved Suppliers - ISO 13485 Normal Supplier Qualification Process Requirements ISO 13485:2016 - Medical Device Quality Management Systems 9
P Process or Performance Qualification Requirements Qualification and Validation (including 21 CFR Part 11) 5
M Process Qualification - What is the recommended Sample Size for PQ? Qualification and Validation (including 21 CFR Part 11) 10
R Process Qualification, Equipment Qualification, Operator Qualification - Sequence? Manufacturing and Related Processes 10
M Process Validation and Qualification of Legacy Equipment ISO 13485:2016 - Medical Device Quality Management Systems 10
R Line Leader Assessment form - Qualification process for production line leader Training - Internal, External, Online and Distance Learning 1
M Process Validation/Equipment Qualification - Class 1A (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 12
S Operational Qualification for injection molding process Qualification and Validation (including 21 CFR Part 11) 2
M PQ and PPQ - Process Qualification for medical device manufacturing ISO 13485:2016 - Medical Device Quality Management Systems 1
S Example Flow Chart of Qualification of Process wanted APQP and PPAP 3
Y NPI (New Product Introduction) Process Qualification System Required Document Control Systems, Procedures, Forms and Templates 8
R Gage Tolerance vs. Process Parameter Tolerance - Process qualification Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Q Determining the Appropriate Sample Size for Process Validation/Qualification Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
J Product/Process Qualification Plan Manufacturing and Related Processes 6
M Passivation Process Manufacturing and Related Processes 2
G Mfr. Process Validation BEFORE Design Transfer? Other Medical Device and Orthopedic Related Topics 1
M Clinical evaluation interface with the risk management process EU Medical Device Regulations 9
D Integrated Management System (IMS) - Process Vs Clause Based Audits ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
D IATF NC Interaction of process diagram - Missing contingency plan Process Maps, Process Mapping and Turtle Diagrams 9
J Cleaning Process for Delicate Parts Manufacturing and Related Processes 4
E Audit Finding - Measurement of Process - Continuous Improvement - Trend Analysis Oil and Gas Industry Standards and Regulations 22
J ISO 10993-1:2018 Format to Perform Risk Management Process US Food and Drug Administration (FDA) 1
I Are suppliers required to hand over process validation reports? ISO 13485:2016 - Medical Device Quality Management Systems 20
S Process Monitoring using SPC software Quality Assurance and Compliance Software Tools and Solutions 6
PQ Systems Introducing Process Improvement to New Employees Using SQCpack Software 0
Quality Runner Continuous monitoring of validated process – sample sizes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
H What kind of chart to detect WHEN a in-process NC is about to occur? Reliability Analysis - Predictions, Testing and Standards 2
FuzzyD ISO 13485 change in process ISO 13485:2016 - Medical Device Quality Management Systems 3
S Combo Product Assembly Process OQ Qualification and Validation (including 21 CFR Part 11) 0
V Process revalidation periodic criteria Manufacturing and Related Processes 5
G Fully refurbishing process EU Medical Device Regulations 3
A Information on Process-based Internal Auditing Needed Internal Auditing 6
Sortinghat Auditing the Manufacturing Process IATF 16949 - Automotive Quality Systems Standard 3
V Process Owner Quality Management System (QMS) Manuals 1
S Measurment of DPU/DPMO in Assembly process Capability, Accuracy and Stability - Processes, Machines, etc. 1
J Can you have one Core Process (PEAR) with multiple plants? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
D "FAI" Requirements/Process for New Supplier Materials Design and Development of Products and Processes 3
W Nonconforming-Material Process Nonconformance and Corrective Action 1
K Sr Process Engineer (Hillsboro, OR) - SPC Program Focused - Ultra Clean Technology Job Openings, Consulting and Employment Opportunities 0
S RFQ Process Automotive Supplier Quality Assurance and other Supplier Issues 1
M Vendor selection process AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5

Similar threads

Top Bottom