Hi All, I have a prototype medical device that's a single-use medium-risk handheld tool. It has buttons on the outside and a battery on the inside. I send parts to the vendor who assembles it, does packaging, and does sterilization (via Ethylene Oxide) so it can be sold to customers.
The vendor is already able to do full process validation for the packaging and sterilization, but they only test the product afterwards to ensure that the sterile container isn't compromised (maintains sterility). It's up to my team to determine how to test for functionality. We need to confirm that the product remains functional after shipping testing, and also after sterilization & accelerated aging (for shelf life). Our 'functionality' is basically to prove the buttons work and that the battery lasts for 3 hours of continuous operation.
The FDA regulation 21 CFR § 820.75 "Process validation" requires that "Where the results of a process cannot be fully verified by subsequent inspection and test, the process shall be validated". There are two problems here...
The vendor is already able to do full process validation for the packaging and sterilization, but they only test the product afterwards to ensure that the sterile container isn't compromised (maintains sterility). It's up to my team to determine how to test for functionality. We need to confirm that the product remains functional after shipping testing, and also after sterilization & accelerated aging (for shelf life). Our 'functionality' is basically to prove the buttons work and that the battery lasts for 3 hours of continuous operation.
The FDA regulation 21 CFR § 820.75 "Process validation" requires that "Where the results of a process cannot be fully verified by subsequent inspection and test, the process shall be validated". There are two problems here...
- I cannot test the functionality of products by 100% inspection because the 'battery test' would be to literally use up the battery by running it for 3 hours.
- I also cannot test the functionality of products by 100% inspection after the entire process is over, because it would be 'destructive' testing to tear open the sterile container.
- If our design input requirement states that the battery must last for 3 hours then do I need to literally have a huge batch of devices running for 3 hours? Can I instead just test that the battery is working at all (pass/fail) and also state that the battery OEM has validated the battery to have a certain capacity, which would last 3 hours for the power drawn by my device design?
- Am I required to perform a 100% inspection ('functional test') of the product immediately after assembly, before packaging/sterilization?
- Or can I instead just 'validate' the functionality of the product after it's been packaged/sterilized/aged?
- If I only 'validate' the functionality after packaging/sterilization/aging, then what should my "control plan" look like? Would I have to re-test a small sample size from every future batch manufactured, or could I instead provide some justification that re-testing for functionality isn't necessary until a design change / etc. occurs?