Proficiency Test and Interlaboratory Variability Testing - Required?

I

isoboy - 2008

#11
A form to summarize study results for management review

Here is a form that can be used to summarize the results of these studies for management review. Enjoy!
 

Attachments

Elsmar Forum Sponsor
S

Susan1

#12
Re: Required? Proficiency Test and Interlaboratory variability testing

Actually, the "Shall" comes from the MRA groups like ILAC, APLAC, EA, and NACLA....an accrediting body that does not require PT/ILC - WHERE IT EXISTS - can be denied recognition by the MRA (in other words, they will not be allowed to become a signatory), or if already in the MRA, the accrediting body can be suspended.

Hope this helps.

Hershal
ISO Guide 43 has now been replaced by ISO/IEC 17043...1st issue 2010-02-01!
 
Thread starter Similar threads Forum Replies Date
C Difference between proficiency test and interlaboratory test? General Measurement Device and Calibration Topics 3
G Where to find Proficiency Test provider for photometrics (spectral %R, %T, photo density) General Measurement Device and Calibration Topics 9
T Response to ISO 17025 Proficiency Test z score ISO 17025 related Discussions 7
A Proficiency Test Service Provider for Water Testing needed General Measurement Device and Calibration Topics 5
E A2LA Proficiency Test Requirement help needed General Measurement Device and Calibration Topics 5
T Providing A2LA w/ Proficiency Test results General Measurement Device and Calibration Topics 5
Hershal PT/ILC - ISO 17025 section 5.9 - Proficiency Test/Inter-Laboratory Comparisons ISO 17025 related Discussions 27
T Interlaboratory comparison or proficiency testing in destructive testing of welded joints ISO 17025 related Discussions 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
D What foreign language proficiency would give the biggest edge to medical device regulatory affairs professionals in the world today or in future? Career and Occupation Discussions 6
C What is "Participation in interlaboratory comparison or proficiency-testing programs" General Measurement Device and Calibration Topics 8
E Defining Sub-Disciplines for Chemical Testing Laboratory Employee Proficiency Testing General Measurement Device and Calibration Topics 1
E Inter-Laboratory Proficiency Testing Result Analysis - z Score or Normalized Error General Measurement Device and Calibration Topics 3
R NAPT's 10V Reference Electricals 732B Proficiency Testing enrollment General Measurement Device and Calibration Topics 8
P Do proficiency testing samples require a 510(k)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N CMM Measurement Uncertainty - Proficiency Testing for 3D on CMM Measurement Uncertainty (MU) 3
Hershal IAS offers Proficiency Testing General Measurement Device and Calibration Topics 1
D Proficiency Testing - Interpersonnel comparison among the calibration technicians General Measurement Device and Calibration Topics 1
L Technician Proficiency Testing - Small Scale and Balances Laboratory General Measurement Device and Calibration Topics 7
C New to 17025 and need a little help in proficiency testing ISO 17025 related Discussions 7
K Proficiency Testing - ISO 17025 and Laboratory Scope ISO 17025 related Discussions 5
J Proficiency Testing - Guide 25 and ISO 17025 ISO 17025 related Discussions 1
D Looking for information on calibration proficiency testing (dimensional) Training - Internal, External, Online and Distance Learning 1
Marc In-tolerance Probability - Level of Confidence - Uncertainty and Proficiency Measurement Uncertainty (MU) 0
Marc Uncertainty and Proficiency Testing Measurement Uncertainty (MU) 3
Marc APLAC - A2LA - ISO/IEC Guide 43-1 (1997) - 17025 - Laboratory Proficiency Testing ISO 17025 related Discussions 0
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
E Tensile test machine, tinius olsen 602 Manufacturing and Related Processes 1
M Test failure Root cause not found Customer Complaints 10
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5
Ed Panek Disaster Recovery Test IT (Information Technology) Service Management 3
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1

Similar threads