Proper FDA Classification for a DICOM Viewer

c.mitch

Quite Involved in Discussions
#11
Hello Sagai,

Once again sorry to answer late. Better than never.
100% aggree with you that the sampling based on my own experience doesn't enforce the classification of software. I just wanted to show that class C is not systematic for DICOM viewer.

bye.
 
Elsmar Forum Sponsor

cmeby

Involved In Discussions
#12
Hello
Just came across this thread, and maybe can provide some extra help since it was recent.
Our product is a DICOM viewer. It is classified as ClassII for FDA.
Some of the other DICOM viewers are classified as Class I for FDA.
The reason and difference is the Intended Use and the quality of the images you are showing. (evaluate Lossy and Lossless as part of this)

If your device is intended for 'Diagnostic Use'...it is Class II.
If your device is intended for 'Reference and Referral Only' (where the diagnosis is still done using the PACS workstation)...it is Class I.

Another item of issue is Measurement function. If you do use measurement it will change your classification in most countries. In EU and TGA this is a special case. 'Class I software with measuring'
 

c.mitch

Quite Involved in Discussions
#13
Thank you cmeby.

Just another information about measurement: computing roi, voi or other geometric or statistical data is not a measurement function.
Justification: the measurement is done by the imager when it acquires the modality. The dicom viewer doesn't measure but processes data (with simple or complex maths) to generate new data.
So in EU, a simple dicom viewer is class I, not Im.
 

sagai

Quite Involved in Discussions
#14
I should write :eek:

Diagnostic purpose is information only to determine that you have a medical device based on the Food Drug and Cosmetic Act.
It has nothing to do the with classification, sorry. This information neither determine the class nor any product code, classification CFR, etc.

I kindly suggest the following public information sources to take into consideration, when sharing such information on a public domain like this forum.

CDRH Learn

On the page, please consider the Overview of Regulatory Requirements: Medical Devices training and the others relating to this subject.

I think all of us have a responsibility to declare, define, share and communicate only the information we absolutely sure in order to prevent rumors, fictions and almost "religious" beliefs rolling over planet earth.

Cheers!
 
Last edited:

cmeby

Involved In Discussions
#15
My information on classification came straight from the FDA themselves during discussions about the medical imaging device and its intended use.
So I did not think sharing this information that could not be found in the online documents would be wrong. I was trying to be helpful and save someone in the industry some time. (the answer cannot be found in CDRH Learn)

Respectfully sir, in the scenario of medical image viewing, 'diagnostic usage' is referring to the radiologist using the images to make their diagnosis. Therefore the images must be of a certain quality in order to qualify as such.

A viewer that is not intended for 'Diagnostic use' can be classified according to the guidance as Class I. It is just sharing images - compressed or not and not of a quality that can assure the radiologist they can use them for diagnostic decisions.
They are considered for Reference and Referral Only and have to be marketed as such.
 

cmeby

Involved In Discussions
#16
Thank you VERY much c.mitch!!!
This is indeed the case...and has caused confusion for us as we try to classify our device for CE.
Your help is appreciated.
And so is your delivery of correction about the measuring function in such a humble and polite way :)
 

sagai

Quite Involved in Discussions
#17
My caveat was for this content you have shared.
If your device is intended for 'Diagnostic Use'...it is Class II.
and its still exist.
Diagnostic use does not mandate that the device is Class II, it mandates the device is a medical device.
Cheers
 

cmeby

Involved In Discussions
#18
You took the statement out of context and read as you want to read.

The original poster Sorcerer was asking why some viewers are Class I and some viewers are Class II when he was looking for predicate devices.

When it comes to medical image viewer, the intended use makes the difference between these classes.

The intended usage including the ability to diagnose based on the images:
known in the industry as radiological "Diagnostic Use" medical viewer. These are Class II viewers.
The intended usage for 'Reference and Referral' only are Class I viewers.

This is simple information provided to me by the FDA, and coincidentally is the same for Health Canada.

Why you decided to 'skewer me' and suggest I am spreading mis-information? Or that I go back to the basic training - I cannot take a guess. But as a newer user on this forum it makes me more hesitant to share information in the future when there are people like you out there.

And don't sign your message 'cheers' unless it is an apology. It is patronizing.
 

sagai

Quite Involved in Discussions
#19
You guy kidding with me :tg: it is not personal stuff.
I have added the value may help the kind audience.
Cheers!
 
Last edited:

Ronen E

Problem Solver
Staff member
Moderator
#20
You took the statement out of context and read as you want to read.

The original poster Sorcerer was asking why some viewers are Class I and some viewers are Class II when he was looking for predicate devices.

When it comes to medical image viewer, the intended use makes the difference between these classes.

The intended usage including the ability to diagnose based on the images:
known in the industry as radiological "Diagnostic Use" medical viewer. These are Class II viewers.
The intended usage for 'Reference and Referral' only are Class I viewers.

This is simple information provided to me by the FDA, and coincidentally is the same for Health Canada.

Why you decided to 'skewer me' and suggest I am spreading mis-information? Or that I go back to the basic training - I cannot take a guess. But as a newer user on this forum it makes me more hesitant to share information in the future when there are people like you out there.

And don't sign your message 'cheers' unless it is an apology. It is patronizing.
Hi,

Don't stress too much. If you have useful content that you're confident about and willing to share - please continue to share. I'm sure that there will always be someone who will appreciate it / be benefited by it. I appreciate it. There are many participants around and they all have their personal views and styles, but until further notice this is a private website and only the owner (Marc) or anyone he delegates authority to, has the right to rule out anything we write here. Each and every user will do their own math when they use whatever advice they pick up here. I certainly won't rely on a single unofficial source when making critical decisions. Heck, I don't even trust myself!...:lol:

Take it easy and thanks again for sharing.

Regards...
Ronen.
 
Thread starter Similar threads Forum Replies Date
M What is the proper way to document measurements Measurement Uncertainty (MU) 1
M Proper way to develop LCL when only given a lower spec limit Reliability Analysis - Predictions, Testing and Standards 7
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L How to write a proper defect description in a NCR Nonconformance and Corrective Action 4
M Proper method to test flow of a regulator Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
M Proper terms for describing Hardness General Measurement Device and Calibration Topics 5
M Proper ISO standard for Pressure Testing and Leak Rates Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
C Proper Thickness Gauge for Foam/Felt and Soft Materials General Measurement Device and Calibration Topics 5
D Proper Tool to Measure Wall Thickness (ISO 3611 - Micrometers) Manufacturing and Related Processes 7
x-files Proper Document Control Requests Document Control Systems, Procedures, Forms and Templates 1
M Proper Gauging - GD&T - Attribute gage to check true positioning on the shop floor Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
S Proper Record/Document Manipulation Document Control Systems, Procedures, Forms and Templates 2
I Proper gauging of tight tolerance (reamed) holes General Measurement Device and Calibration Topics 4
kedarg6500 Selection of proper modifying agent for Aluminium Silicon Alloy Manufacturing and Related Processes 2
J Proper Amount of Data Points for Cpk Capability, Accuracy and Stability - Processes, Machines, etc. 12
K Proper YTD PPM Supplier Calculation Six Sigma 3
E Measuring Parts without Proper Tools - Measurement Device Resolution Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
J Proper Method of Measuring Hex Socket Heads Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
B Getting the proper information from the supplier - Returning Nonconforming Material Nonconformance and Corrective Action 4
G What is the proper Thanks Giving turkey sampling size? Coffee Break and Water Cooler Discussions 9
B How to Determine the Proper Solder Class Manufacturing and Related Processes 5
S What is the proper flow to feedback the defective found during QA sampling check? Quality Tools, Improvement and Analysis 8
M Proper Format for a Manufacturing Route Sheet Manufacturing and Related Processes 6
K Putting a regulated device on the market without the proper ISO Certs and CE Marking EU Medical Device Regulations 9
L Proper Personnel Ratio - Quality to Manufacturing Manufacturing and Related Processes 6
M ISO14001 nonconformance - no proper treatment of wastewater prior to discharge ISO 14001:2015 Specific Discussions 6
A What is the proper definition for Class 1 & 2 design changes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C Proper GD&T for spindle & housing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
A Proper term for 'Easy Identification'? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
M How to implement a proper Calibration System? General Measurement Device and Calibration Topics 5
F Is there a proper protocol to follow for Self Certification? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
B True Position without tertiary datum? Establishing a proper rotation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 12
J Proper way to calibrate a check fixture General Measurement Device and Calibration Topics 7
G Proper Poundage for Checking XX Tolerance Plain Plug Gages? General Measurement Device and Calibration Topics 2
G Labeling Responsibility - Manufacturer vs. Repacker? Proper Component Use ISO 13485:2016 - Medical Device Quality Management Systems 7
L Procedure for proper use and protection of Registration and Accreditation Marks Document Control Systems, Procedures, Forms and Templates 21
C What is the proper method for cleaning dial calipers? General Measurement Device and Calibration Topics 4
N Proper use of the Blueing process when inspecting the taper of bushings General Measurement Device and Calibration Topics 6
J Proper use of gauge pins - Inspecting thru holes of plastic parts General Measurement Device and Calibration Topics 19
I What is the proper date to use as effective date? Old 'procedures', New documents Document Control Systems, Procedures, Forms and Templates 10
D Rockwell vs. Vickers vs. Knoop - Does Thickness Determine the Proper Test? General Measurement Device and Calibration Topics 1
M What is the proper way to use a GO NO/GO guage? General Measurement Device and Calibration Topics 3
B Preventive and Corrective Action Mapping - Proper Sequencing ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Registered or Certified - What is the proper terminology? QS-9000 - American Automotive Manufacturers Standard 17
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
D FDA Information - Revising the Instructions for Use US Food and Drug Administration (FDA) 0
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
S Manufacturing Process FDA FOIA Medical Device and FDA Regulations and Standards News 3