Hi All,
Below is the High level depiction of our Class I Medical electrical equipment, powered with an external Medical grade power supply.
The power supply for this medical device is medical graded and has a protective earth at its mains power chord.
The medical graded power supply is power Via an UPS.
Inside the box named Medical Device is operated from the 18V DC sourced by the medical power supply
Does the block below named as Medical Device needs a protective earth line connected to it.
Some more insight:
The Box is made of Sheet metal and powder coated.
The digital ground of the circuit is connected to the metal box via the mechanical mounting hole
The device has applied parts both of type CF and BF.
The PC interface to this device is via Fibre optic cable.
The PC and the monitor is not placed inside the patient room. Only the medical device is placed in the patient room.
Please let me know if you have any additional queries...
Benjamin Weber
First of all: You may have the power supply classified as class I and the medical device as class II.
You as the manufacturer decide wether the medical device is class I or II. It depends on your insulation concept.
If you want the medical device to be class I: Is the PE conductor from the power supply input available at the output? I am not talking about the ground line I am almost sure, it is not. If I am right, you have to make the medical device class II compliant, i.e. double or reinforced insulation!
Thank you very much for your quick response.
For your first question... The answer is... as you mentioned. the PE is not available at the output of the power supply. Output is only the DC +V & GND.
The existing design is a old design and has the digital ground connected to the metal enclosure.
All the Leakage currents and isolation satisfies the standard requirement
Is it not that when we classify a ME equipment... Power supply with the Medical device together is considered as an ME equipment? and in that case this device in question falls under Class I right? This is my understanding. Please comment...
michael Cejnar
If your power supply (PSU) is a specified component in UIM and preferably on device decal and has a unique connector to prevent foreseeable misuse of customer connecting a different non-medical PSU, then I believe it is all one medical device and is Class I, with functionally earthed SELV secondary circuit with accessible metal parts, and you already have double / reinforced insulation from mains in the medical PSU.
However, operating from SELF (e.g. 18V) alone is not protection enough - see Cl. 6.2 of IEC60601-1, for why it does not recognize SELV operated Class III.
Once you have accessible metal parts, that are only functionally earthed, then you have problems.
Key issue is what other equipment may be connected to your medical device, and whether faults in other equipment could make your metal case live - failing touch current per Figure 17 or Fig 18 of IEC 60601-1:2005+AMD1:2012 (Ed3.1).
Consider also if your ME is in Patient Environment - if user can touch patient, especially if any open tissue or saline flowing into veins, and touch your metal chassis - then your allowable touch leakage current becomes much lower - at CF level.
We have had this design for over a decade, but we are an installed system, run the whole medical system from a medical isolation transformer and require in UIM that any connected equipment be suitably medically isolated to IEC60601-1 and system leakage tested by hospital biomeds prior to use.
In our latest redesign, we have protectively earthed chassis in one version and have no accessible metal parts in another, including a plastic ECG socket and recessed HDMI an USB sockets.
Hope that helps.
J Secondary circuits and protective earth IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
U Protective vs. Functional Earth Differences IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Is protective earth connection regarded as a part of the applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Medical Device Protective Earth (PE) issues IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R What is the meaning of "Highest RATED Current" in Protective Earth Impedance Test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
S Protective Earth Test in IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Can the shield of a cable be used as a protective bonding conductor? Design and Development of Products and Processes 2
N Accelerated Aging Test for Disposable Protective Garment for Incontinence Other US Medical Device Regulations 1
N Protective Earthing of Medical Device Moving Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Minimum Testing Requirements for a Clear Protective Equipment Barrier (Class II 510K) Other US Medical Device Regulations 1
R How much Protective Current Rating of Building Branch Circuit should be considered? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R How to Design a Protective System for Infusion Pump IEC 60601 - Medical Electrical Equipment Safety Standards Series 48
S How to overload a linear transformer with multiple protective devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
J Protective Earthing for Medical Device with two wire power cord in metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A Essential Requirements Checklist for PPE (Personal Protective Equipment) CE Marking (Conformité Européene) / CB Scheme 6
G ASTM b137 - Is protective value equal to protective capability? Various Other Specifications, Standards, and related Requirements 3
Z Packaging and shipping of multiple PWBA/PCBAs in one ESD protective bag Supplier Quality Assurance and other Supplier Issues 5
B Allergic reactions complaints for a protective apparel (impervious gown) Customer Complaints 4
A Maintaining traceability through our manufacturing process - Protective suits - 110 Misc. Quality Assurance and Business Systems Related Topics 1
S Potential Equalization Conductor vs Functional Earth Terminal IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
J IEC 60601-1 Electrical Safety testing - Class 1 ME externally powered - Earth resistance testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Sidney Vianna The Future of Spaceship Earth - The UN Sustainable Development Goals Sustainability, Green Initiatives and Ecology 1
Marc Google Earth Timelapse update shows Earth from 1984-2016 After Work and Weekend Discussion Topics 1
A Insulation between Secondary Accessible Parts and Earth IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M ISO 13485 - Are Earth Leakage Tests Required? ISO 13485:2016 - Medical Device Quality Management Systems 2
R Separation requirement for Class II equipment with functional earth IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
Richard Regalado Our Beautiful World courtesy of NASA - Photos of Earth Coffee Break and Water Cooler Discussions 3
marmotte Adding a Funct. Earth. to a class II / BF device...retest needed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C USB, Clearances and Earth Leakage - Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P Is it true? Mars closer to Earth in August Coffee Break and Water Cooler Discussions 8
T Earth Hour - Sat March 28, 2009 Miscellaneous Environmental Standards and EMS Related Discussions 27
Marc Google Earth in 4D with Historical 'Overlay' Maps After Work and Weekend Discussion Topics 0
I Google Earth Hacks Web Site Coffee Break and Water Cooler Discussions 2
Marc Google Earth v4 Released: Mac and Linux Support at Last After Work and Weekend Discussion Topics 3
Wes Bucey Google maps and Google Earth satellite photos - WoW! After Work and Weekend Discussion Topics 28
D QMS - Could someone give me a down to earth explanation of QMS? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
D Control Number for Class III and IV devices Canada Medical Device Regulations 0
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
K "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14