PVC Tubing Extrusion - Class I medical device
G
gereard_kgb
Hi Everyone,
We are just launching a factory of medical devices. One of the products is medical tubing (no contact with patient)- class I medical device according to MDD.
Can anyone share their experiance with this product?
Does it have to be manufactured in the clean-room?
How about the extrusion process itself? This is made of PVC, what temperature should be used for extrusion?
Thanks in advance for any comments.
Bye,
Gereard!
:thanks:
Rob Udo
Environmental requirments regarding the manufacturing location depends on the intended use of the device and the associated risks. You have to comply with the essential requirements from the MDD! If you can comply with these requirements without a clean-room environment, it is fine.
Regards,
Rob
G
gereard_kgb
Thank you Rob.
I fully agree with your comments. EN550 also points out that intended use is very important as far as sterilization and manufacuring conditions are concerned.
Yet there is one more issue raised by EN550 - being bio-burden. Its control plays some role in sterilization.
The point is we talked to a few auditors and each had different point of view.
The intended use for this product is such that it could be in full manufactured and assambled out of the clean-room.
My main concern now is sterilization. I think it should be validated for a certain bio-burden (no matter low or big). In case a product is manufactured out of the controlled cleanness zones there is nothing you can say about the bio-burned, right?
In this case what sterilization process is appropriate?
Is there a EO sterilization process long enough that it can be assumed all the products were sterilized no matter what the bio-burden was?
Is there an EN norm for this? (I don't know any)
Any hint is welcome.
G
Gert Sorensen
No. If it is non-invasive, and it does not get in contact with invasive parts, then it does not need to be.
I will highly recommend that you test this yourselves. Calibration, maintenance, the extruder you are using, the humidity of your compound etc. will have an impact. Depending on the tolerances you may have a "walk in the park" production or a nightmare when you extrude PVC. Only you will be able securely establish your process parameters.
Bio-burden. Its control plays some role in sterilization.
Yes it does, but basically you only need to consider bio-burden for invasive devices - in my experience.
Well, small wonder! :notme:
Now, you lost me here. If the product can be assembled and manufactured in a non-controlled zone, then why does it need to be sterilized. Could you elaborate a bit?
Usually, when sterilizing you use an incubation of 2 million. This should be - more or less - easily sterilized in an EtO sterilization (depending on productdesign).
It is just like foods. You can "nuke" your product from here to kingdom come, it still may not be safe :confused: Some bacteria create poisonous substances when they die. This won't be solved by sterilization. And this is one of the reasons that bio-burden is interesting. However, if the product is non-invasive I believe that the usual interpretation is that you have nothing to worry about.
:bigwave:
G
gereard_kgb
Gert,
Thank you very much for your detailed answer.
As far as the extrusion itself is concerned we are making some progress and exactly as you said - the experimental method is the one :)
The product we are to manufacture are connecting tubes - product that has no contact with the patient. Yet for some customers it needs to be delivered sterile (and most of the manufacturers have the tubes available as both sterile and non-sterile) so we need to be able to have it certified by a notified body (the goods need to be CE marked of course).
I fully agree that bio-burden itself needs to be carefully controlled for all the invasive products as their pyrogenity is of serious concern.
However were said that that bio-burden needs to be controlled so that it could be said that the sterilization process can be properly validated.
That's why I was wondering if there is any process long enough that could give acceptable certainty that no matter what bacteria were there they are all dead (of course not gone - but for this particular product it doesn't matter).
Is there any standard to support the discussed issues?
:rolleyes:
G
Gert Sorensen
I think you are looking for:
ISO 14937:2000 - Sterilization of health care products - General requirements for characterization of a sterilizing agent and the development, validation and routine control of a sterilization process for medical devices
chris1price
Hi
I don't think there is a sterilisation process long enough to guarantee sterilising all materials, irrespective of the bioburden. For Gamma, VDMax25 methods are suitable for bioburden upto 1000CFU, beyond this, more intensive validation methods are required. For EtO, knowledge of the bioburden is critical to establishing the exposure, if the bioburden can vary, then sterility cannot be guaranteed. Extremely long EtO cycles may look attractive, however they also result in long aeration times and difficulties in removing the Eto Residuals.
One further comment, for this type of application, cleanrooms are used to control contamination by dust and hair on the product more than bioburden.
Q
Quality Priest
We have several product using silicone tubing.
The tubing is supplied to us in a `clean` state, But the tubing is very rarely in a state that we would class as acceptable for sterilization.
As you can well appreciate form your extrusion process the product has a tactile nature, we routinely send away samples for Bioburden prior to manufacture to find the counts (1000cfu +) to high to guarantee sterility at Gamma Irradiation dose of 25kGy.
I have done allot of work with the stability of the post irradiated silicone tubing including elasticity, memory also dosing at 50+ kGy (not recommended). Silicone tubing has probably been the hardest product to justify to our regulatory authority.
As Chris has mentioned about the cleanroom if you can justify that the product will be contaminant free with out the use of a cleanroom (thinking also visually to you customer) then you are only looking at reducing your bioburden to limits of below 1000cfu
chris1price
I'm surprised your bioburden on silicone tube is that high, I would be looking at ways of reducing it. When extruded it should be essentially sterile, from then on it's getting contaminated. If its kept clean and not handled too much, the bioburden should stay low. I would also suggest a wash with 50/50 IPA/water before its pouched will remove most bioburden.
G
gereard_kgb
Chris1price, what is IPA?
I agree to what Quality Priest said about the use of clean-room. If it could be justified that the product can be manufactured outside of it with low contamination. Yet how can it be justified that the product is contamination and dust free when it is manufactured out of the clean-room? What are the other zones of controlled cleanness? what are the applicable norms?
I share the opinion of Chris1price that the product coming out of the extruding machine is almost sterile. There is only an issue what happens next and how the product is dealt with.
What do you think about washing the product with alcohol? This should reduce bioburden quite much.
chris1price Naturally Occurring DEHP in PVC Tubing - Statements about tubing being DEHP-Free EU Medical Device Regulations 6
R How to bond PVC (Tygon) Tubing and Validate the Process ISO 13485:2016 - Medical Device Quality Management Systems 13
rob73 Advice/recommendations for equipment to measure a soft pvc tube - Optical Measuring? General Measurement Device and Calibration Topics 4
T What Test Certificate should accompany supply of Cable XLPE/SWA/PVC 4CX150MM2 Document Control Systems, Procedures, Forms and Templates 2
P IEC 60601 Clause 57.3 / 8.11.3.2 - PVC Flexible Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K How to define PVC Pellet Quality Manufacturing and Related Processes 8
R Standard Relevance & Updating - ASME B&PVC Various Other Specifications, Standards, and related Requirements 4
S Benchmarking PVC Profile Extrusion Metrics Benchmarking 2
somashekar DEHP vs. DINP as Plasticizer in PVC (Emerging Health hazard) Other Medical Device and Orthopedic Related Topics 3
somashekar Material Compatibility - PVC (halogenated polymers) and Santoprene (TPE) Manufacturing and Related Processes 4
M PVC balls fail out laquered process drying Manufacturing and Related Processes 3
F Conical gauges (ISO 5356) how to use on PVC (soft part) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
S Use of Phthalates (PVC) - MDD 2007/47/EC - Has anyone started Labelling Devices EU Medical Device Regulations 71
K Any adhesive needed for PVC and PC natural bond? ISO 13485:2016 - Medical Device Quality Management Systems 1
ScottK ASME B&PVC Section VIII Div 1 question regarding "Seal" Various Other Specifications, Standards, and related Requirements 0
ScottK The term "Quality Control Manual" - ASME B&PVC question Various Other Specifications, Standards, and related Requirements 2
D PVC Packaging on Pass Through Parts - An environmental issue? Miscellaneous Environmental Standards and EMS Related Discussions 3
R PVC Profile Extrusion - Custom Profile Extrusion company Manufacturing and Related Processes 4
D Measuring fine tubing (< 0.100") "Wall by weight" General Measurement Device and Calibration Topics 11
A Tubing Clog Tester - Convert from Air Bubble to Flow Rate Reliability Analysis - Predictions, Testing and Standards 1
M Butt welding of 13Cr tubing Manufacturing and Related Processes 0
M Regulations for oxygen/air-tubing and connectors inside of ME equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Insufflation Tubing & Filter Set Product Classification question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
J Micro Tubing Concentricity Checking using Crosshairs Manufacturing and Related Processes 1
J Mechanical Tubing Mill Tolerance Reference Manufacturing and Related Processes 5
R Fuses that in series with each supply lead are only housed by Heat Shrink Tubing IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
T API 5CT PSL1 & PSL2 Levels - Specification for Casing and Tubing Oil and Gas Industry Standards and Regulations 5
B Profilometer Damages Parts? Extruded Plastic Tubing General Measurement Device and Calibration Topics 5
B SAE AS81914 - Convoluted Tubing Tensile Stress Various Other Specifications, Standards, and related Requirements 5
Ron Rompen Very Unique Prototype Requirement - Cutting Holes in Tubing Manufacturing and Related Processes 2
G Standard to use as a guide in measuring surface "roughness" of aluminum tubing General Measurement Device and Calibration Topics 1
N Grilamid L25 vs. L20 for Nylon 12 Balloon Tubing Manufacturing and Related Processes 1
J Inspection Equipment for Checking Serrated Steel Tubing Depths Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Health Canada Device Classification - Close-off clip for use with a tubing set Canada Medical Device Regulations 9
D Calculating Cpk for Tubing Wall Thickness - Extrusion Process - How to approach? Statistical Analysis Tools, Techniques and SPC 16
P Outsourced Process or Supplier? Purchased Formed Metal Tubing ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Looking for MIL-P-15147 - Ink marking copper-nickel tubing Other ISO and International Standards and European Regulations 4
J ASTM D709 - Does D709 give a wall thickness tolerance for tubing? Various Other Specifications, Standards, and related Requirements 5
apestate Tolerances for cold rolled, hot rolled, DOM tubing Other ISO and International Standards and European Regulations 10
Nicole Desouza Plastic Extrusion of Medical Device Components - No ISO or AS 9100 Certifications Other Medical Device and Orthopedic Related Topics 7
Nicole Desouza AS9100D - Calibration of Dies & Tooling - Plastic Extrusion AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
F AS9100 D - Verification of purchased product - Sheet metal and aluminum extrusion stock AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
L Injection & Extrusion Process FMEA FMEA and Control Plans 1
A 1st Time Purchase of Aluminum Extrusion - First Article? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
M Improving Quality in an Extrusion Blow Molding Company Quality Manager and Management Related Issues 2
K Is there any specific Standard on visual defects for extrusion and machined parts ? Quality Tools, Improvement and Analysis 6
D Chemical Training in use of Sodium Hydroxide to clean Extrusion Dies Training - Internal, External, Online and Distance Learning 2
H Is there a Standard on the Reuse of Resin in Extrusion Process? ISO 13485:2016 - Medical Device Quality Management Systems 1
QMMike Extrusion Dies - Perishable Tooling? APQP and PPAP 4
T How clean should the water be needed during extrusion as a cooling agent - Catheters Other Medical Device and Orthopedic Related Topics 5