We are a company in Turkey with all QA system managed in Turkish. Our ISO 13485 and CE mark are registered by a Turkish NB. Based on their interpretation of clause 6.4 of MEDDEV 2.7.1 Rev 04, our NB insists that the evaluator must be a "medical doctor" per definition of "specialist" on dictionary for Medical English. More of it, they say the Ministry of Health urges them to make this adaptation.
I tried to reason with them by explaining/translating the clause: when it states "e.g.
specialist clinical expertise in the relevant medical specialty", it means a clinician is a possible person, not a requisite.
Please correct me if I am wrong or help me convince them if I am correct.
Many thanks!
Marc
Fully vaccinated are you?
Leader
My Thanks in advance to anyone who can help with this - "Bump"!
Mark Meer
Trusted Information Resource
In my opinion, you are correct.
First, MEDDEV documents are intended as guidance, and so the NB shouldn't be citing you non-conformances based on MEDDEV contents. It is the regulations (or Directive) that is the authoritative requirements.
With regards to clinical evaluations, I'm not aware of anything in either the old Directive (MDD 93/42/EEC), or new regulation (EU 2017/745) that calls for persons doing clinical evaluation reports to have specific qualifications. That being said, there should be some documentation of their qualification within the context of your quality system.
Clinical investigations, on the other hand, are different... For example, the MDD 93/42/EEC states: "The investigations must be performed under the responsibility of a medical practitioner or another authorized qualified person in an appropriate environment." (Annex X, 2.3.6). ...but even then, to interpret this as it must be a "medical doctor" is a bit of a stretch...
FoGia
Involved In Discussions
I share the opinion that the guidance is not an authoritative requirement but I have the interpretation that it sets a level of expectation for the competent authorities/notified bodies on which they can base their evaluation of the clinical evaluation.
Even the MEDDEV doesn't call for a medical doctor, but it says that you must have in your team of clinical evaluators someone with sufficient clinical expertise. Depending on the type of medical device and its intended purpose I would imagine that a notified body would be in their right to impose the participation of a doctor during a clinical evaluation.
Thread starter Similar threads Forum Replies Date
R Qualification of Evaluator - Clinical Evaluation Report MEDDEV 2.7.1 Rev 04 Other Medical Device and Orthopedic Related Topics 1
D Supplier Qualification Qualification and Validation (including 21 CFR Part 11) 2
S Sample size considerations in medical process qualification Reliability Analysis - Predictions, Testing and Standards 11
A Guidance on Supplier Qualification - Phase 1 pharma manufacturing Supplier Quality Assurance and other Supplier Issues 4
K Analytical Method Qualification Vs Validation expectations ISO 13485:2016 - Medical Device Quality Management Systems 1
B ISO 13485 Qualification of R&D Equipment Qualification and Validation (including 21 CFR Part 11) 1
F Supplier : Product Qualification Supplier Quality Assurance and other Supplier Issues 0
V Is it required to recalibrate reference- standard instruments after equipment qualification General Measurement Device and Calibration Topics 0
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Revamping Supplier Qualification, Re-evaluation, and Monitoring Requirements ISO 13485:2016 - Medical Device Quality Management Systems 8
S Performance Qualification and Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
Stoic Details of Operational Qualification (OQ) test design for plastic extrusion processes, effect of material property noise, and GHTF/SG3/N99-10 US Medical Device Regulations 2
B Clean room shut down and re-validation or re-qualification? Other Medical Device Related Standards 6
Y Creating Equipment qualification protocol Qualification and Validation (including 21 CFR Part 11) 2
W Medical Device - fixture qualification and MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
G RO/DI water system qualification Qualification and Validation (including 21 CFR Part 11) 3
A Qualification of clinical assessment scales EU Medical Device Regulations 1
K Supplier Qualification for Engineering Consultant? Supplier Quality Assurance and other Supplier Issues 3
F Component maintenance qualification - Regulation (EU) No 1321/2014 Article 5 EASA and JAA Aviation Standards and Requirements 6
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
R Qualification list of businesses approved to Off-Shore Aerospace work. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
E Qualification for non gmp service providers Supplier Quality Assurance and other Supplier Issues 1
V Process, component or full product re-qualification: leaded to unleaded solder Qualification and Validation (including 21 CFR Part 11) 8
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Qualification and Validation (including 21 CFR Part 11) 5
Y Examples of TRB Reports for MIL-PRF-31032 Qualification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
Jane's Like-for-like critical raw material change qualification - type of testing/ number of lots required ISO 13485:2016 - Medical Device Quality Management Systems 4
K PQ validation qualification - Asked to write a PQ protocol ISO 13485:2016 - Medical Device Quality Management Systems 6
dinaroxentool Question about qualification as a medical device or accessory in Europe EU Medical Device Regulations 2
T Qualification testing of Lead acid batteries Reliability Analysis - Predictions, Testing and Standards 0
Z Qualification of AzureDevOps in Gxp Qualification and Validation (including 21 CFR Part 11) 12
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
D Determining of sample size for 'Operational Qualification' AQL - Acceptable Quality Level 5
B VDA 6.3 Qualification as Process Auditor training course and exam VDA Standards - Germany's Automotive Standards 0
D Performance Qualification per GHTF Guidance Other Medical Device Related Standards 12
M Informational EU – Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 2
dgrainger Informational MDCG 2019-11: Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 0
H How to prepare Performance Qualification (PQ) for autoclave General Measurement Device and Calibration Topics 2
D ISO 13485 - Equipment validation , qualification Qualification and Validation (including 21 CFR Part 11) 6
DuncanGibbons Qualification vs Certification in an Aerospace context AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
Z Operational Qualification for Injection Molding Supplier Quality Assurance and other Supplier Issues 6
S AIAG CQI Auditor Qualification and 3rd Party Certification Requirements General Auditing Discussions 2
Dobby1979 FDA Biomarker Qualification Program - Has anyone prepared an application? Medical Device and FDA Regulations and Standards News 0
R Qualification of Distributors such as McMaster-Carr and Digikey to Medical Device Companies ISO 13485:2016 - Medical Device Quality Management Systems 3
xcanals_tecno-med.es Qualification of sterilization indicators as medical devices with the EU MDR 2017/745 EU Medical Device Regulations 8
Z Training Qualification Matrix for ISO 9001:2015 Audit General Auditing Discussions 8
S Internal Auditing for API Spec Q1 - auditor qualification requirements Oil and Gas Industry Standards and Regulations 6
J Qualification of a Software as a Medical Device (SaMD) guidance under MDR EU Medical Device Regulations 9
I Compressed air system qualification and 1 micron particle analysis Other Medical Device Regulations World-Wide 0
S Medical Device Supplier Component Qualification Service Industry Specific Topics 0
Similar threads
Top Bottom