Qualifying Suppliers of Calibration Services

  • Thread starter Thread starter Janie
  • Start date Start date
"By adding exceptions to our procedure" can put us in compliance with everything in QS we are not doing or do not wish to do.
Are you saying we do not need to have testing done by certified to NIST labs if we state they are not certified in the procedure?
 
Elsmar Forum Sponsor
This thread started with your saying you were getting services from an unqualified supplier - unqualified by the terms of your procedure requirements. That you can change the procedure for - change your requirements.

BUT - if their standards are not traceable to NIST and/or they are not in compliance with generally accepted calibration laboratory standards (have cal procedures, etc.) you have a problem. These are two different issues.

By the way - as far as I know, there is no such thing as a Laboratory being NIST *certified*. I may be missing something but as far as I know there is no NIST certification for a cal lab - only for standards. Nor is there a QS registration (or certification) for calibration labs. Labs within QS registered firms are considered *qualified* but not for outside testing or calibrating other companies equipment.

An FYI:

-----snippo-----


Subject: RE: Q: Certificate of Calibration/Guerra/MacKenzie
Sender: "J. H. MACKENZIE"

> Guerra Asks:
"Some times, I must send the equipment calibrate in manufacturer, since I
can't find accredited laboratories with capacity to calibrate the
equipment. In these cases, I ask the calibration certificate to indicate
standards used and traceability, measured values and calibration
uncertainties.

Do you think this information in calibration certificates is enough, too
> much or too less?"
>- - - - - - - -

This information should be sufficient. There is no requirement to use accredited labs for calibration as long as measurements are traceable to national or international standards.In fact you can do the calibrations yourself if you have the facilities and equipment.

Regards,

Jim Mackenzie
Consultant, Medical Devices Lead Auditor.
QUALIMEDD LTD.
 
One more FYI:

-----snippo-----

Miguel -

I have seen many cases where the manufacturer's certificate that comes with an instrument is not compliant with the calibration requirements of the standard. It is really only a certificate of conformance, rather than a certificate of calibration, and certifies only that the instrument meets the manufacturer's published specifications. When you back track, this is based on a statistical sampling technique, and a true traceable calibration cannot be shown. This may be what your auditor was seeing.

On the other hand, an accredited lab is usually extreme overkill, for shop instruments especially, as well as being economically unjustifiable. You generally don't need the level of uncertainty available from an accredited lab for shop floor or general inspection instruments. You are better off putting in the time and calculations up front when setting up your system (or afterwards, as you make new purchases), to determine which instruments need what level of calibration, and basing your calibration services - both internal & external - on those requirements.

My experience in setting up and running calibration labs is that a large percentage of registrar auditors really do not understand calibration and metrology, and therefore self-impose requirements which have nothing to do with the standard or its practical implementation. Many of these requirements have everything to do with making the auditor's job easier & absolutely nothing to do with ensuring the quality and consistency of the product, which after all, is what this whole thing is about.

Hope this helps.

Nancy Harris
 