Hello Everyone, i have a question.
We have this handheld device (class II) that is powered by an alkaline battery for a 510k. It also has a USB port that is strictly for offline downloading of data by connecting to a computer (and in no way can be used to charge the device) and at that time the main device screen (that shows vital) is disabled.
With regards to 60601-1-2 (4th ed) & 510k, what should be the possible test modes:
1. An internally powered handheld device - we already considered and it passed 60601-1-2.
2. Do we have to EMC test the data downloading through the USB as a separate mode?
OR
Can we argue (for not doing this is) that the port is strictly for offline data transfer and put cautions/warnings in the instructions instructing users (along with risk analysis,etc)? And if the user does it anyways and connects to a USB charger/computer it shows a screen to disconnect USB (and disables showing the current vital values). Will that be sufficient?
Thank you for reading ! Any insights are most appreciated!
Elsmar Forum Sponsor
somashekar
Staff member
Super Moderator
#2
Re: Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic p
Due to Good Friday and Easter Sunday holidays, you may find some delay in response. Please wait, till some 60601 experts here will come up with a response....
Thanks for joining Elsmar and WELCOME to the Cove.
shimonv
Trusted Information Resource
#3
OR
Hi titanius,
I am not an expert on EMC test but this is my two cents worth:
Your argument plus risk analysis plus mitigation makes sense providing that the offline download does not occur in the patient environment. I don't know what the workflow is but if you are near the patient while downloading you should - in concept - check for immunity and interference.
-Shimon
M
MedMartin
#4
Hi Titanius,
best is to talk to your test house for their opinion.
For the designs I was involved we tried to test all modes of the devices.
One design was very similar to yours, a handheld and no measurement while USB connected. We tested also in the USB communication mode as we considered the data transfer critical for immunity. And emission testing is also necessary.
The 4th edition of the 60601-1-2 also recommends testing e.g. in standby modes which would also support testing USB communication mode.
Best regards
Martin
VinceTech
Involved In Discussions
#5
Hi,
If the USB is not restricted to be used by operator or patient, it's SIP/SOP. Then EMC test shall include it. If USB is restricted only to be used for engineering 'for example, for development or factory use', it maybe exempt from EMC test.
Regards,
T
titanius
#6
Thank you VinceTech, MedMartin, shimonv, somashekar!
:thanx: so much! this helps a lot.
Your arguments totally make sense. The test house didn't recommend data download mode due to the 'development nature' of the connection.Maybe I might need to rethink about this.
Thread starter Similar threads Forum Replies Date
8 Drawing Feature Question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
D Question: How to analyze numerical and attribute data Reliability Analysis - Predictions, Testing and Standards 11
D Question on electronic signatures and initials on batch records ISO 13485:2016 - Medical Device Quality Management Systems 3
A Special Characteristic question from Automotive CSR Customer and Company Specific Requirements 9
Ron Rompen GDT Question - is this even correct? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
A PPAP question for audit APQP and PPAP 16
C ISO Question - Do you say "I-S-O" or "I-Soh"? Misc. Quality Assurance and Business Systems Related Topics 14
D Question regarding where "validations" fit according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Question on Supplier Quality (SCAR) ISO 13485:2016 - Medical Device Quality Management Systems 3
A +0/-.001 Tolerance question - Feature size is 1.249 +0/-.001 - Actually measures 1.2493 (.0003 OHL Inspection, Prints (Drawings), Testing, Sampling and Related Topics 12
D Question and advice for a supplier self audit questionnaire ISO 13485:2016 - Medical Device Quality Management Systems 6
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
A Design Change/ECO Related Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Configuration Item definition question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
C Quality Management System Question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
Q ISO 9001/IATF 16949 Audit Finding Question - Document Retention IATF 16949 - Automotive Quality Systems Standard 11
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
I IMDS Error Message Question RoHS, REACH, ELV, IMDS and Restricted Substances 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
C Gauge R&R Question Using Minitab Software 1
J IATF 16949 Internal Audit question - Auditor's responsibility Internal Auditing 6
K Question on MDR classification EU Medical Device Regulations 4
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
D Work Instruction Question ISO 13485:2016 - Medical Device Quality Management Systems 5
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
John C. Abnet VDA 6.3 - Question 7.3 - "blocking of parts" VDA Standards - Germany's Automotive Standards 6
D Approved supplier list - Distributors question ISO 13485:2016 - Medical Device Quality Management Systems 6
D Equipment Register and PM question ISO 13485:2016 - Medical Device Quality Management Systems 2
D Question regarding "storage and distribution" ISO 13485:2016 - Medical Device Quality Management Systems 1
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
D Question regarding customer feedback process ISO 13485:2016 - Medical Device Quality Management Systems 3
D Equipment Register related question ISO 13485:2016 - Medical Device Quality Management Systems 1
S Study sign off question / responsibilities ISO 13485:2016 - Medical Device Quality Management Systems 3
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
M Question for Auditors - "Off the Record" Conversation? General Auditing Discussions 14
D Question regarding ECO process, specifically for Life Science products and defining form fit and function ISO 13485:2016 - Medical Device Quality Management Systems 1
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
A Question on Authorized Representative in Malaysia Other Medical Device Regulations World-Wide 3
D Limited Scope for second site Question? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
Anonymous16-2 Labeling Question (Dietary Supplements/Food) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
T Question for: Cg & Cgk calculation General Measurement Device and Calibration Topics 3
N ASL Question for GitHub ISO 13485:2016 - Medical Device Quality Management Systems 7
hogheavenfarm GDT Flatness measurement question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
A Question on ISO 14001:2015 - Are annual audits required? ISO 14001:2015 Specific Discussions 11
dinaroxentool Question about FDA Classification of a Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
J Another DFAR question 252.225-7009 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
F Conflict Mineral Smelter Question RoHS, REACH, ELV, IMDS and Restricted Substances 8
R NRTL - Scope Question - Off-the-Shelf Plug In IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Similar threads
Top Bottom