Question on IEC 60601 clause 15.5.1.1 (transformer overheating)
Roland chung
Hi all,
Clause 15.5.1.1 (transformer overheating) in A1:2012 states that protective devices should be included as part of the short circuit test and overload test provided that:
- the component is one with High-Integrity characteristics, and
- 2 MOOP are provided between the output contacts of the transformer up to the component with High-Integrity characteristics.
For the first dash, it is not easy to determine a component is High-Integrity or not. From the rationale of clause 4.9, there are some steps and additional considerations. There are several different types of protective devices, such as fuses, PTCs, breakers and ICs. It seems that if we identify fuses as High-Integrity Component, most people would say OK. But when it comes to the PTCs, it would be totally other case I think.
For the second dash, it would require at least 2 mm of air clearance (2 MOOP) from Table 15 even for 12 VAC output (rated mains voltage 230V, pollution 2). Under the same conditions, Table 12 requires 1.6 mm (2 MOPP). But my instinct is that both 2 mm and 1.6 mm are too large. According to IEC 60664, 1.0 mm for air clearance is enough (transient value 1500V for secondary circuit). The clearance 1.0 mm derived from IEC 60664 is exactly the same as the creepage derived from Table 16 (2 MOOP).
Could you please share your point of view? Thanks!
Regards,
Roland
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J IEC 60601-1, ed 3.1 - Blue icon question IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R IEC 60601-2-18 Endoscopic Medical Electronic Equipment question IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A IEC 60601-2-24 Question: Free Flow Clamp on Set? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Question about IEC 60601-2-10 - Which waveform should I measure? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 60601-1 - RS232 Connector Requirements - Female Connectors question IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
M Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K IEC 61010-2-100 Labeling question - IVD equipment Other ISO and International Standards and European Regulations 1
S IEC/EN 60335-2-30 cl.30 Plastic Enclosure on Fan Heater question Various Other Specifications, Standards, and related Requirements 2
3 Scope question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
L Question regarding "Customer Property" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
T Noob question regarding sample size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
K 510(k) Manufacturer question (private label and manufacture under another persons 510(k)) US Food and Drug Administration (FDA) 1
M Question About Rolled Throughput Yield Manufacturing and Related Processes 5
M Quick Gage R&R Question Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
S Question: Best alternatives to audits for ensuring compliance Quality Tools, Improvement and Analysis 5
Ed Panek Question from 13485 Auditor ISO 13485:2016 - Medical Device Quality Management Systems 2
C Question on KAFO (Knee Ankle Foot Orthosis device) classification CE Marking (Conformité Européene) / CB Scheme 0
T BMW GS 93010-2 Question Various Other Specifications, Standards, and related Requirements 1
S FDA 510(K) submission question US Food and Drug Administration (FDA) 4
Ron Rompen Question regarding CMM probe damage General Measurement Device and Calibration Topics 4
T FDA UDI Question - Class II Medical Device Other US Medical Device Regulations 2
C Complaint Return Sample Size Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
8 Drawing Feature Question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
D Question: How to analyze numerical and attribute data Reliability Analysis - Predictions, Testing and Standards 11
D Question on electronic signatures and initials on batch records ISO 13485:2016 - Medical Device Quality Management Systems 4
A Special Characteristic question from Automotive CSR Customer and Company Specific Requirements 9
Ron Rompen GDT Question - is this even correct? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
A PPAP question for audit APQP and PPAP 17
C ISO Question - Do you say "I-S-O" or "I-Soh"? Misc. Quality Assurance and Business Systems Related Topics 14
D Question regarding where "validations" fit according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Question on Supplier Quality (SCAR) ISO 13485:2016 - Medical Device Quality Management Systems 6
A +0/-.001 Tolerance question - Feature size is 1.249 +0/-.001 - Actually measures 1.2493 (.0003 OHL Inspection, Prints (Drawings), Testing, Sampling and Related Topics 12
D Question and advice for a supplier self audit questionnaire ISO 13485:2016 - Medical Device Quality Management Systems 6
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
A Design Change/ECO Related Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Configuration Item definition question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
C Quality Management System Question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
Q ISO 9001/IATF 16949 Audit Finding Question - Document Retention IATF 16949 - Automotive Quality Systems Standard 11
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
I IMDS Error Message Question RoHS, REACH, ELV, IMDS and Restricted Substances 5
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
C Gauge R&R Question Using Minitab Software 1
K Question on MDR classification EU Medical Device Regulations 5
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
D Work Instruction Question ISO 13485:2016 - Medical Device Quality Management Systems 5
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
John C. Abnet VDA 6.3 - Question 7.3 - "blocking of parts" VDA Standards - Germany's Automotive Standards 6