Radiated Immunity Device Test Orientations - IEC 61000-4-3

rothlis

Involved In Discussions
#1
It appears to be standard practice to test four 90° "side" device orientations during radiated immunity (61000-4-3) and then also the "top" and "bottom" if the device can be used as such. This is reflected in several guidances I've encountered, and is presumably the inference behind clause 8.2 of 61000-4-3, where it says that "all sides shall be exposed ... In other cases, as determined for example by the type and size of EUT or the frequencies of test, more than four azimuths may need to be exposed".

From an electrical perspective, it would make sense to test orientations for all six sides of a cube to ensure all possibly orthogonal fields are covered by at least a 45° angle of incidence, on the assumption that any one wire or circuit in the device could be oriented in any direction and could be vulnerable to interference. So why isn't it standard practice to test all six orientations for every device? Is it because large devices aren't easily tested in the "top" and "bottom" orientations in a typical EMC lab setup?

As I see it, to exclude the "top" and "bottom" orientations is to presume that the real-world interference is never going to originate from a vector above or below the device and, unless I'm missing something, that doesn't seem like a justifiable assumption.
 
Elsmar Forum Sponsor

phase90

Starting to get Involved
#2
What I understand is it relates to the feasibility of testing. If the DUT is a smaller hand-held type of device that can operate in any orientation, it can be tested on all six sides. If it is a unit that cannot be tipped onto its side and operate, then four sides is all you can do. I guess until someone configures a chamber to be 10 meters high and have the antenna above the DUT, that is the reality of what can be done.
 

rothlis

Involved In Discussions
#3
What I understand is it relates to the feasibility of testing. If the DUT is a smaller hand-held type of device that can operate in any orientation, it can be tested on all six sides. If it is a unit that cannot be tipped onto its side and operate, then four sides is all you can do. I guess until someone configures a chamber to be 10 meters high and have the antenna above the DUT, that is the reality of what can be done.
Thanks for the input. That's what I suspected, but I also perceive tradition has limited testing to four sides whenever top and bottom orientations aren't obviously possible during normal operation, regardless of whether testing is feasible with the device on its side. And it seems to me that 61000-4-3 contributes to this oversight by inferring that four sides is the expected set of orientations, when really that is only appropriate to a minority of devices.
 

phase90

Starting to get Involved
#4
I just pulled up edition 3.1 of 61000-4-3. In clause 8.2, it says the "test shall normally be performed with the generating antenna facing each side of the EUT. When the equipment can be used in different orientations (i.e. vertical or horizontal) all sides shall be exposed t the field during the test. When technically justified, some EUT's can be tested by exposing fewer faces to the generating antenna. In other cases, as determined for example by the type and size of EUT or the frequencies of test, more than four azimuths may need to be exposed."

So really, they are saying if it can operate in any orientation, it needs to be tested on all side, otherwise only the four sides as that is what is feasible.

Yes, technically, floor standing equipment may be exposed from above or below out in the field, but the typical exposure will come from a source horizontal to the equipment. A EMC test house engineer kind of told me years ago that was why they only test four sides. Nowadays, I guess one would have to look at it from a risk analysis perspective to determine if other faces are needed.
 

rothlis

Involved In Discussions
#5
I read that excerpt as follows:
The "four azimuths" in the last sentence are equivalent to "each side" in the first sentence, so that "normally be performed" = left, right, front, back. The second sentence is then saying that if the device can be used without the top and bottom facing the ceiling and floor then the top and bottom also need to be exposed, and the last sentence is essentially saying the same thing in more general terms that include other conditions. So exposure of the top and bottom is treated as the exception rather than the rule.

It may be true that the majority of interference sources would come from a horizontal plane, but there is no shortage of cases where devices which have a defined upright position during use (and thus not treated as operating horizontally) would easily be exposed from the top or bottom (e.g., any pole or wall mounted device could have sources mounted above or below it, and any cart or similar device could have sources sitting directly above it).

I'm inclined to think that there is a deficiency in the language of the standard that has led to an accepted practice which doesn't align with the intent of the test. But maybe I'm over-analyzing this.
 
Thread starter Similar threads Forum Replies Date
S IEC 60601-1-2 Radiated Immunity Test - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Radiated Immunity Testing as per IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1-2 Radiated E-Field Emission Testing - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P AIM 7351731 RFID reader immunity Other US Medical Device Regulations 3
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S IEC 60601-1-2 immunity to proximity fields Test setup - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1-2 - RF Conducted Immunity Test Setup - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
W IEC 60601-1-2 Susceptibility/Immunity for device intended to be used in patient body IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D IEC 60601-2-44: 202.101 Immunity Testing of Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Q Can a fail still be a pass? (Criteria Level B for 61000-4-6 Conducted Immunity) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Electrosurgery Immunity Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
N IEC 60601-1-2 ESD: Lower Compliance Level for ESD Immunity IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Impact using two different materials for product enclosure in radiation and immunity IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Y Is it weird for my conducted immunity (4-6) test result? Need help. CE Marking (Conformité Européene) / CB Scheme 1
J MD Directive vs EMC Directive - Totally excluded (for both emissions and immunity)? ISO 13485:2016 - Medical Device Quality Management Systems 1
T Device & Accessory, Components in device, or System EU Medical Device Regulations 0
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 0
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
M Sample record for verification performed by importers before placing a device on the market EU Medical Device Regulations 0
S IVD Device History Record ISO 13485:2016 - Medical Device Quality Management Systems 3
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
K How to handle GTINs for different configurations of one device with one part number? Other Medical Device Related Standards 0
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
C Multiple owners for one listed device Medical Device and FDA Regulations and Standards News 1
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
J Promoting and marketing of a non approved device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Russian Device to US US Food and Drug Administration (FDA) 5
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
S DHF/DMR/MDF for a software-only, cloud-based, single-instance device Medical Information Technology, Medical Software and Health Informatics 2
V MDSAP Requirements for Device Designer Canada Medical Device Regulations 1
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2

Similar threads

Top Bottom