Hello everyone!

I have a question regarding repackaging under the MDR.

Case study: A distributor purchases syringes from a manufacturer. The syringes come in boxes that contain 100pcs (secondary package). A pharmacy orders to the distributor 20pcs. The distributor takes 20pcs from the box of 100pcs and puts them in another box to send the order to the pharmacy. The primary packaging of the syringe remains intact. Is that considered a repackaging under the MDR?
 
Elsmar Forum Sponsor
#2
Hi, did you come to a conclusion for this? I have been working through 2 options:

1. This would be repackaging under Article 16 (due to change in pack size) as long as it could be shown that the secondary packaging is not required to maintain a sterile barrier and any relabelling reflects the original labelling if the primary packaging does not contain all required aspects. The process of repackaging should also not affect the integrity of the primary packaging.

2. If the secondary packaging has been designed by the manufacturer to protect the integrity of the sterile barrier (for example during transit) then further validation would be required on the new packaging leading to obligations that pertain to becoming the manufacturer.

It would be interesting to know how others view this example.
 

Cybel

Involved In Discussions
#3
Hi,
MDR, art. 16, 2 states:

2. For the purposes of point (c) of paragraph 1, the following shall not be considered to be a modification of a device that could affect its compliance with the applicable requirements:
(b) changes to the outer packaging of a device already placed on the market, including a change of pack size, if the repackaging is necessary in order to market the device in the relevant Member State and if it is carried out in such conditions that the original condition of the device cannot be affected by it. In the case of devices placed on the market in sterile condition, it shall be presumed that the original condition of the device is adversely affected if the packaging that is necessary for maintaining the sterile condition is opened, damaged or otherwise negatively affected by the repackaging.


So, your case study seems to be considered as repackaging but not as a repackaging affecting the device compliance to MDR.
However, it appears the point 4 of article 16 applies:

4. At least 28 days prior to making the relabelled or repackaged device available on the market, distributors or importers carrying out any of the activities mentioned in points (a) and (b) of paragraph 2 shall inform the manufacturer and the competent authority of the Member State in which they plan to make the device available of the intention to make the relabelled or repackaged device available and, upon request, shall provide the manufacturer and the competent authority with a sample or mock-up of the relabelled or repackaged device, including any translated label and instructions for use.
Within the same period of 28 days, the distributor or importer shall submit to the competent authority a certificate, issued by a notified body designated for the type of devices that are subject to activities mentioned in points (a) and (b) of paragraph 2, attesting that the quality management system of the distributer or importer complies with the requirements laid down in paragraph 3.


So, my additional questions, using your first example:
Case study: A distributor purchases syringes from a manufacturer. The syringes come in boxes that contain 100pcs (secondary package). A pharmacy orders to the distributor 20pcs. The distributor takes 20pcs from the box of 100pcs and puts them in another box to send the order to the pharmacy.
- distributors shall inform the manufacturer and the authority EACH time they perform this activity (for each order they receive from a pharmacy? for each LOT received by manufacturer?)
- distributors need to have their quality system certificated for this process, even if this process cannot affect the compliance of the device to MDR?
 

koulidis

Starting to get Involved
#4
After making some digging i found out that it dependeds on the packaging made by the manufacturer. If a distributor alters any packaging intended by the manufacturer for retail sale, it is considered as repackaging. If the distributor alters only the packaging that is only used for trasportation than it is not considered repackaging.

However i am still not sure. There should have been some clarification about that from the MDCG.
 
Thread starter Similar threads Forum Replies Date
Y Sterile device packaging under MDR EU Medical Device Regulations 3
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
M Packaging standards - ASTM vs ISTA Other Medical Device Related Standards 1
M Packaging studies - ISTA vs ASTM CE Marking (Conformité Européene) / CB Scheme 0
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 2
C Packaging Long Parts Manufacturing and Related Processes 3
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
V Packaging in Distributor Practices EU Medical Device Regulations 0
R Green dot required on packaging? Medical Device and FDA Regulations and Standards News 2
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Applicable ISO standards - packaging and others Medical Device and FDA Regulations and Standards News 4
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
C 21 CFR 803 malfunction detected during packaging - report? Other US Medical Device Regulations 1
E Latex in packaging Other Medical Device Related Standards 4
H Expired packaging material - Expired unused Tyvek pouches Other Medical Device and Orthopedic Related Topics 2
A GUDID - Packaging Levels 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A Packaging device with accessories already on the market Other Medical Device Regulations World-Wide 0
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 41
T Root Cause Failure Analysis - Not following Customer packaging Specification Problem Solving, Root Cause Fault and Failure Analysis 9
Q EU Rep for Packaging waste, WEEE and Battery disposal CE Marking (Conformité Européene) / CB Scheme 0
L AS5553 Clause 3.1.7 e "Control packaging material ..................reused" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
L Packaging unsterile medical devices ISO 13485:2016 - Medical Device Quality Management Systems 6
M MDR - Packaging - Secondary Medical Device Boxes EU Medical Device Regulations 2
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
M UDI size on packaging EU Medical Device Regulations 3
M UDI placement - Packaging levels EU Medical Device Regulations 2
E Unit Dose Packaging Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
K Packaging and Damaged Parts returned from a Service Provider (Anodize) - Who is Responsible? Manufacturing and Related Processes 20
B 510(k) clearance claim on medical device product packaging Other US Medical Device Regulations 6
E Packaging Specifications Mercedes-Benz U.S. International IATF 16949 - Automotive Quality Systems Standard 4
T Colour of symbols on Medical Device labelling and packaging Other Medical Device Related Standards 1
R Labels on sterile packaging EU Medical Device Regulations 5
A Interpretation of Article 16 (2b) - Packaging, samples and Certificate EU Medical Device Regulations 10
I Medical Device Packaging Multilingual Labeling EU Medical Device Regulations 3
N Process Flow diagram steps for inspection and packaging APQP and PPAP 4
C ISO 7010 m002 mandatory action - Color requirements on Packaging IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
shimonv GTIN Number on the Packaging of a Serviced Medical Device EU Medical Device Regulations 0
shimonv Is UDI required on the Packaging of a Serviced/Repaired Medical Device? Other US Medical Device Regulations 3
C QR Code on Medical Device Packaging and Notification to Notified Body EU Medical Device Regulations 1
TheMightyWife Address Change - Medical Device Packaging Labelling Requirements EU Medical Device Regulations 7
E Use of FDA Cleared logo on packaging US Food and Drug Administration (FDA) 8
V Tyvek packaging for Medical device process validation Qualification and Validation (including 21 CFR Part 11) 2
S Symbols - Device label versus Packaging label ISO 13485:2016 - Medical Device Quality Management Systems 1
A Packaging seal inspection Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
A Capability Analysis for Packaging Seal Strength with spec. >0.1 Kgf using Minitab Using Minitab Software 6
A Medical Device Class 2A, non sterile bodyworn packaging requirements CE Marking (Conformité Européene) / CB Scheme 7
N Medical device non sterile packaging requirement CE Marking (Conformité Européene) / CB Scheme 1
P Drug packaging / filling operations Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 8

Similar threads

Top Bottom