Re-validation of EtO sterilization after change of tape

Kees Fremerij

Starting to get Involved
#1
Íf a change in the secondairy packaging occurs, does the EtO sterilization process need to be revalidated? The change is that paper tape for closing the outer box will be used instead of plastic tape. However more tape will be used, also on the box corners. Paper is more penetratable I would say. Not relevant minor change? Thanks in advance!
 
Elsmar Forum Sponsor

planB

Trusted Information Resource
#2
Kees,

sure you have to re-validate your EtO sterilization process - the question is whether you can do this exercise by performing a documented adoption analysis or whether you have to do experimental work to demonstrate continued safety and effectiveness of your sterilization process.

Essentially you would have to answer: does the replacement tape create a greater challenge to the sterilization cycle in lethality and/or residuals?

To this end, you may walk through AAMI TIR28:2009, Product adoption and process equivalence for ethylene oxide sterilization, and specifically through the questions listed in Annex A, where it says "If the answer to any of the following questions is “yes,” further evaluation to include product qualification studies of the candidate product might be required".

HTH,
Gerhard
 
Last edited:
Thread starter Similar threads Forum Replies Date
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
S Fractional cycle fail in Eto sterilization validation ISO 13485:2016 - Medical Device Quality Management Systems 3
P ETO Sterilization Validation Process and an NCR - ISO 13485:2003 CMDCAS audit ISO 13485:2016 - Medical Device Quality Management Systems 1
N ETO Sterilisation Validation - EO Residual Minimum Sample Requirement ISO 13485:2016 - Medical Device Quality Management Systems 2
X EtO Steriliser Software (PACS) Update Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
A Eto Sterilisation Re-Validation Requirements Other Medical Device Related Standards 4
I ETO Sterilizer Validation (ISO 11135:2007) Other Medical Device Related Standards 23
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
I Guidelines that specify the frequency of ETO Sterilizer Re-Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 1
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Test Method Validation Qualification and Validation (including 21 CFR Part 11) 4
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
T Annual Validation as a detection mode on a PFMEA? FMEA and Control Plans 5
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 6
P Human Factors / Usability validation in the time of COVID Human Factors and Ergonomics in Engineering 14
C Template for Excel Validation Reliability Analysis - Predictions, Testing and Standards 6
M IT validation for a paper based MD repair company QMS ISO 13485:2016 - Medical Device Quality Management Systems 6
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
Stoic Manual soldering processes - 100% verifiable, or always requiring validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Rees System Validation Qualification and Validation (including 21 CFR Part 11) 1
K PQ validation qualification - Asked to write a PQ protocol ISO 13485:2016 - Medical Device Quality Management Systems 6
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9
Y Retrospective Validation - Class I device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3

Similar threads

Top Bottom