Reacting to Gage R&R Findings

M

mlowden

#1
Reaction to R&R findings

Question regarding R&R findings.
We have recently conducted an R&R on a digital balance that is used in our lab for weighting samples for dissolution to run on XRF. The procedure was developed in the early 80’s and specifies a sample size of 2g +/-.0005g. This sample is then digested and diluted to 100ml. When this procedure was developed, it was concluded that a variance in the sample of .002g was insignificant, in other words the XRF could not repeatedly discriminate. However, in an effort to reduce the amount of variation, the author of the procedure chose to limit the total sample variation to .001g.

The results of the R&R indicate the total R&R is 39% of the total tolerance, (.001g), which is above the allowable 30%.

The question is: Should the procedure be changed to allow a tolerance of +/-.001g for the sample, thus increasing the amount of total variation, but would cut the R&R findings to 19% of the total tolerance?

We are just wondering what direction to take in this matter.

Just for information, our plant is certified QS9000 and the lab is A2LA accredited. Thanks for any replies.
 
Elsmar Forum Sponsor
T

tomvehoski

#3
Who authored the specification? If it was your company I don't see any problem with changing it. You may have to get customer approval though if it is subject to PPAP or other requirements. I would make sure you have other data (experiments) showing the change does not affect the test in case you are challenged.

If it is a ANSI, ASTM or other industry standard, you probably cannot since it would affect 17025 requirements.

Tom
 
B

Benoit

#4
How do you performed the r&r study on scale?

I want to know how do you perform your r&r study on scale?

Regards
 
M
#5
If you have a product tolerance try that in your calculations; we do our GRR's based on product tolerance.
If the tolerance is internal, change it as long as you can meet the product tolerance. Be careful how much.

I'm not sure what the physical procedure is that you are using, but the below may give you some possible directions.

If the EV is high, review the equipment. It may have been able to hold the smaller tolerance fresh out of the box, but may have lost some of its discrimination over time. This would require re-hab at the manufacturer to fix. Avoid contamination of the samples used, from being exposed to heat or cold which may cause condensation. Use the exact same container(s) to avoid container variance. Add controls to be used in the study.

If the AV is high, then you have to fine tune the techniques used by all operators to insure that the item is placed at the same place on the scale(?), that an equal amount of time is spent for stabilization on the scale, that extra care is taken when placing and removing the item from the scale. This is also a good time to train others in 'best' procedures.
 
Thread starter Similar threads Forum Replies Date
F Should I perform several Gage R&R for every caliper to cover all specs range of use? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
T Gage R&R study - Ordinal data Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
Q Gage Pin Sets? Calibrate/Verify them all or just a sample size? General Measurement Device and Calibration Topics 16
S If 4 operate to measure, what is the value of k2?(MSA gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S Faro Arm, Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
8 MSA--Should I Use Parts or Gage Pins? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
C Gage Block Grade 1 - Calibration Measurement Uncertainty (MU) 15
R Gage Measuring 15 Parts at a Time Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
Q Gage calibration in disarray ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Gage R&R automated test equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
PQ Systems Recipe for Success! 10 Ingredients for a Healthy Gage Management System Using GAGEpack Software 0
D Gage Repeatability for Faro Arm Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S Comparing measurement results for a part from two ring gage General Measurement Device and Calibration Topics 3
M Quick Gage R&R Question Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
PQ Systems Best Practices In Gage Management Using GAGEpack Software 0
PQ Systems Gage R&R - 3 Part Series Using GAGEpack Software 0
I Excel based Gage R&R VS Minitab calculation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
R Repeatability Type 1 Gage and Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
W Proper gage to use for receiving inspection General Measurement Device and Calibration Topics 9
A GAGE R&R Binomial with master list (for method validation) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
A R chart from ANOVA gage R&R test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
M Type 1 Gage r study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
L Gage R&R TMV Acceptance Criteria Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
E Zero part to part variation - Gage R&R project Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
M Gage R&R and right way to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 19
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
C Gage Block Wringing General Measurement Device and Calibration Topics 3
B Gage R&R with NDC=1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Thread Plug Gage Pitch GO Diameter out of spec AQL - Acceptable Quality Level 8
G Gage R&R - Where am I going wrong? Part of a FAIR submission (Aerospace) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Gage R&R Excel templates Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Gage R&R with more than 3 appraisers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Anerol C Gage R&R Template AIAG 4th Edition IATF 16949 - Automotive Quality Systems Standard 3
T Attributes SPC study - Attributive control (Go gage) Statistical Analysis Tools, Techniques and SPC 5
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
V Gage Management and Gage R&R Software General Measurement Device and Calibration Topics 1
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1
S Gage R&R on alloy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10

Similar threads

Top Bottom