Reasonable GRR result for a 0-1" micrometer with a total tolerance of .002"

5

57jeeper

#1
Hello to the group.

Does anyone have any experience with expected GRR results for standard measuring equipment. What would be a reasonable GRR result for a 0-1" micrometer with a total tolerance of .002".

We performed a GRR on a rather small part and got 17%, which our customer is not satisfied with. They want less than 10%.

Is that an unreasonable requirement?

Thanks,
 
Elsmar Forum Sponsor
R

rsimano

#2
< 10% would indicate that variation is in the parts not the operator or equipment. Anything over 10% would indicate errors in the system that may or may not need to be repaired depending on importance(since the customer is not accepting 17% then it should be deemed important).

17% would indicate some operator or equipment error. You may need to evaluate your results to determine which. This may also indicate the operators were not measuring in the same location or using the same technique.

How many operators were used?
How many parts were measured?
How many measurement trials were done to each part?
Did all operators measure in the same location on the same part?
 

Ninja

Looking for Reality
Staff member
Super Moderator
#3
<snip>

We performed a GRR on a rather small part and got 17%, which our customer is not satisfied with. They want less than 10%.

Is that an unreasonable requirement?

Thanks,
Sounds pretty normal to me. <10% is the normal desire...

It would also be helpful to know what your spec range is....or is that what you meant by the 0.002" ?
 
5

57jeeper

#4
Here is the process we followed.

How many operators were used?
= 3 Operators
How many parts were measured?
= 10 parts
How many measurement trials were done to each part?
= 3 measurements
Did all operators measure in the same location on the same part?
= very close, the location was marked and they visually measured the same place.

Thanks,
 
R

rsimano

#6
IMHO. I would take a look at the gauge or the operator. See which one is giving the greater errors to the measurement variation. Your data should point you to the most likely cause for error.

1. Re-Verify the gauge is good.
2. Check operator measurement techniques to verify all are uniform and tool is used correctly.(micrometer can give incorrect data if the operator is torquing the gauge too tight to the part. I prefer a micrometer with a ratchet knob to reduce variation in torque between operators).
 
D

dickgent

#7
I would look at using a mic stand to better control the process and also assure that the repeatable pressure is applied to the thimble. If the gage does not have a ratchet stop, use one that does. Do you really need to measure your customers part? Since the GRR evaluates the measurement process I have told my customers that I do not need to measure their parts for the GRR. Some accepted that explanation .... some did not. Good Luck!
 

Reluctance

Involved In Discussions
#8
I have some additional questions about the 10 parts that you used for your gage study.

Were these 10 parts all "in tolerance"? Or put another way if you took the difference between the average of largest part and the average of the smallest part how does this compare to your .002" tolerance?

In my own experience I have attempted Gage R&R with parts all inside of the tolerance and come up with similar unsatisfactory results.

For the purposes of a Gage R&R you really need to have parts that can cover the full range of your tolerance and then some. The purpose of the study is to determine if your measurement process can tell you if a part is good or bad. If all you are measuring is "good parts" for the study you have no way of knowing whether or not your gage can tell the difference between a good or bad part.

So, did you include some "bad parts" in your Gage R&R?
 
R

rsimano

#9
Reluctance has a good point. :agree1:

A second question on top of that that i just thought about.
Are the parts from the same lot or from different lots over a period of time?

I have seen negative results come from a GR&R that was performed from samples from a single lot. Reluctance pointed out you need dimensions that cover the range of your tolerance and a single lot would only show part of that range.
 

Jim Wynne

Staff member
Admin
#10
Hello to the group.

Does anyone have any experience with expected GRR results for standard measuring equipment. What would be a reasonable GRR result for a 0-1" micrometer with a total tolerance of .002".

We performed a GRR on a rather small part and got 17%, which our customer is not satisfied with. They want less than 10%.

Is that an unreasonable requirement?

Thanks,
As others have indicated, there could be many reasons for the results you got. It would be best if you could post your data and let us try to see what's going on.

< 10% would indicate that variation is in the parts not the operator or equipment. Anything over 10% would indicate errors in the system that may or may not need to be repaired depending on importance(since the customer is not accepting 17% then it should be deemed important).

17% would indicate some operator or equipment error. You may need to evaluate your results to determine which. This may also indicate the operators were not measuring in the same location or using the same technique.
There could be many things going on with the OP's study and data, but a result <10% isn't necessarily an indication that only part variation is at work, nor is the 17% result necessarily an indication that there is operator or equipment error.

How many operators were used?
How many parts were measured?
How many measurement trials were done to each part?
Did all operators measure in the same location on the same part?
All good questions. :bigwave:

I would look at using a mic stand to better control the process and also assure that the repeatable pressure is applied to the thimble. If the gage does not have a ratchet stop, use one that does. Do you really need to measure your customers part? Since the GRR evaluates the measurement process I have told my customers that I do not need to measure their parts for the GRR. Some accepted that explanation .... some did not. Good Luck!
It would be good to use a stand only if that's the way the parts are normally measured. GR&R, and MSA in general, is a method for determining whether a given measurement system is appropriate for a specific measurement task, which means that it should always be performed on parts and the feature that is to measured in production.

I have some additional questions about the 10 parts that you used for your gage study.

Were these 10 parts all "in tolerance"? Or put another way if you took the difference between the average of largest part and the average of the smallest part how does this compare to your .002" tolerance?

In my own experience I have attempted Gage R&R with parts all inside of the tolerance and come up with similar unsatisfactory results.

For the purposes of a Gage R&R you really need to have parts that can cover the full range of your tolerance and then some. The purpose of the study is to determine if your measurement process can tell you if a part is good or bad. If all you are measuring is "good parts" for the study you have no way of knowing whether or not your gage can tell the difference between a good or bad part.

So, did you include some "bad parts" in your Gage R&R?
Unless one's doing an attributes study, there is no need to include bad parts. The purpose of GR&R is NOT to determine whether a measurement process can tell you if a part is good or bad. Again, that's only for attributes studies. In fact, depending on the purpose of the measurement system, it might not even be necessary for the measured parts to represent the full range of process variation.

I commend all of you to Miner's excellent blog series on MSA.
 
Thread starter Similar threads Forum Replies Date
M What is a reasonable MDSAP cost? ISO 13485:2016 - Medical Device Quality Management Systems 9
S Scheduling a Crisis Management Audit? Reasonable Request? Internal Auditing 2
D What is a reasonable consulting fee for preparation of a complete traditional 510k? Consultants and Consulting 11
C What is a reasonable level of QA Editorial Control? Quality Manager and Management Related Issues 20
A CAPA - Reasonable Timeframes for Effectiveness Evaluation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
N What is a ?Reasonable? failure rate for an electronic device in the 1st year? Quality Tools, Improvement and Analysis 7
S What is a reasonable Bioburden limit? ISO 13485:2016 - Medical Device Quality Management Systems 13
S Packaging Specification template and Reasonable dppm target for printed material? Document Control Systems, Procedures, Forms and Templates 1
F Travel Expenses for Assessors - Registrar states we pay "reasonable travel expenses" Registrars and Notified Bodies 27
Hershal Cutoff - Discussion regarding a reasonable break-even point for calibration General Measurement Device and Calibration Topics 3
G Reasonable time for ISO 9001:2000 implementation? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Icy Mountain Reasonable calibration? 50 foot measuring tape and other basic M&TE General Measurement Device and Calibration Topics 25
M Six Sigma - Where to get a 'reasonable' certification? Training - Internal, External, Online and Distance Learning 3
D What is a reasonable time frame for a system to be 'mature enough' to audit? Miscellaneous Environmental Standards and EMS Related Discussions 11
F Definition Reasonable - Defining a Reasonable Amount of Time Close Out CPAR's Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 6
A Cg, Cgk and GRR of a CMM Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
L GRR for a tolerance that has changed Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
K %GRR was between 10-30% so we have to have a "backup plan" per auditor IATF 16949 - Automotive Quality Systems Standard 15
V Minitab GRR crossed Xbar and r method macro needed Using Minitab Software 9
T GRR based on part tolerance or process variation. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
A Minitab 18: How to automate GRR Summary Table save Reliability Analysis - Predictions, Testing and Standards 1
J Question on very low NDC number with tolerable GRR ratio's Capability, Accuracy and Stability - Processes, Machines, etc. 7
C Evaluating GRR Scores %Tol vs %TV Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
S GRR on ID with set of pin gauges Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Q What does % study variation of 20% mean in GRR? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 17
C %GRR <10% not possible. How to get approval by auditor? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
V GRR Study on a Finished Product - Batch Manufacturing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S Script of R software to conduct Nested GRR for Destructive Testing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
A Hypothesis test for Bias, not for GRR. WHY ?? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B How to perform a MSA - GRR (Gage R&R) A&R/ANOVA Vs GUM/VDA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
J Calculating part variation from historic data for GRR study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S GRR (Gage Repeatability and Reproducibility) for Material Strength Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P New Part - How to choose what needs GRR (Gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
K The basics of GRR (Gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A Need MSA %GRR Interpretation Advice Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M GRR approach for off line Measuring Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P Does anyone have the GRR Formula? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
T MSA - Should I compare my GRR to tolerance or to my sample? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
D What the impact is of a high %GRR/TV Capability, Accuracy and Stability - Processes, Machines, etc. 6
V Sample Range for a valid GRR using ANOVA in Minitab Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
F GRR (Gage R&R) with Different Scales Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
R New to MSA - I was told to use ANOVA for the GRR's Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
J Type 1 GRR (Gage R&R) on a One Sided Specification Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Analysis of GRR studies, MSA 4th Ed, Chapter III section B - Rigid Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S Strange PT Ratio on an Excel spreadsheet for variable GRR ANOVA Method Excel .xls Spreadsheet Templates and Tools 2
A Customer Demands (MSA=Bias + GRR/2). Can somebody explain what it means? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
R ndc=0 and GRR=14% ??? Interpretation - basic question Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
I Minitab ANOVA Gage R&R Error (vs. EXCEL ANOVA GRR Calculation) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
R MSA - GRR (Gage R&R) Requirement for CMM Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 21
Rameshwar25 What is the significance of Spread in Calculating GRR (Gage R&R)? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1