Re: Recall vs field correction
Hi All,
There is a new draft FDA guidance Distinguishing Medical Device Recalls from Product Enhancements and Associated Reporting Requirements - Draft Guidance for
http :// www .fda. gov/MedicalDevices/DeviceRegulationandGuidance/GuidanceDocuments/ucm340518.htm?source=govdelivery - DEAD 404 LINK UNLINKED
This guidance introduces a new term: Product Enhancement - change or improvement to a non-violative device as part of continuous device improvement activities. Product enhancements include, but are not limited to, changes designed to better meet the needs of the user, changes to make the product easier to manufacture, and changes to the appearance of the device that do not affect its use. A product enhancement is both (1) a change to improve the performance or quality of a device, and (2) not a change to remedy a violation of the Federal Food, Drug, and Cosmetic Act (FD&C Act) [21 U.S.C. 321 et seq.] caused by the device. A product enhancement is not a medical device recall.
Other terms are further detailed as well:
Correction means repair, modification, adjustment, relabeling, destruction, or inspection (including patient monitoring) of a product without its physical removal to some other location (21 CFR 7.3(h) and 21 CFR 806.2(d)). Depending on the circumstances involved, a correction can be a recall or product enhancement.
Removal means the physical removal of a device from its point of use to some other location for repair, modification, adjustment, relabeling, destruction, or inspection. (21 CFR 806.2(i)). Depending on the circumstances involved, a removal can be a recall or product enhancement.