robert.beck

Involved In Discussions
#21
Hello Robert,
Interesting situation.

The actions you describe would not be considered a "Recall" per FDA regulations 21 CFR Part 7 or Part 810, since only FDA can determine that a Correction or Removal is a FDA Recall.

Regarding reportability of corrections or removals to FDA per 21 CFR Part 806....reports to FDA are required if the Correction/Removal is initiated to reduce a risk to health. You need to document a Health Hazard Evaluation (HHE) to make that determination.

When you replace the existing device with a new device, do you ask the customer to return the existing device?
If Yes, that would be a removal.
If No, your replacement action does not directly meet the definition for a Correction or Removal.

§806.10 Reports of corrections and removals.
(a) Each device manufacturer or importer shall submit a written report to FDA of any correction or removal of a device initiated by such manufacturer or importer if the correction or removal was initiated:

(1) To reduce a risk to health posed by the device; or

(2) To remedy a violation of the act caused by the device which may present a risk to health unless the information has already been provided as set forth in paragraph (f) of this section or the corrective or removal action is exempt from the reporting requirements under §806.1(b).

Regarding the "hotfix", the same requirements apply. You should document a HHE for the hotfix Correction, and if it meets FDA reportability criteria then you should report it to FDA.

You may find the attached presentation useful.
Thanks for the presentation. Slide #5 implies that there are corrections and removals that do not have to be reported and later slides go into more detail, identifying these situations as "exemptions." The company keeps detailed records on this and is planning to implement an improved tracking and record keeping system in the near future. I think we are in a grey area because most of the device replacements are done as the most time-efficient method of getting a working device into the patient's hands. A small percentage lead to device improvements, and an even smaller percentage could by described as "user error." I am trying to get them to stop doing hotfixes because they should be able to include the vast majority of these into a scheduled maintenance release. Because the device is so easy to use, it is prescribed about 100,000 times/year. Sometimes it is prescribed inappropriately, for instance to a patient with mental health issues, who took it off and threw it across the room.

I have another question along the lines of the same issue. This particular device is very safe and does nothing harmful. It is supposed to monitor some physiological parameters which are non-life threatening. It does nothing more than monitor and report. There is another set of similar physiological parameters, which indicate a serious potentially life-threatening condition, but this device is not intended for or indicated for monitoring of those parameters. The company certainly does not market it for this situation, but sometimes physicians use it that way because they cannot distinguish between the two situations because of their own deficiency as a physician or for convenience, and rarely this leads to a patient death. The company insists their device is not at fault and they cannot control physicians, and a certain amount of misuse is inevitable. The frequency is very low; a high estimate is 0.04% per year. What should be done about these events? My opinion is that they should at least be reported to MAUDE, but the management feels this will seriously and adversely affect their business and argues against it as I noted above. I'm considering leaving them because of this. Am I overreacting or not?
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#22
I have another question along the lines of the same issue. This particular device is very safe and does nothing harmful. It is supposed to monitor some physiological parameters which are non-life threatening. It does nothing more than monitor and report. There is another set of similar physiological parameters, which indicate a serious potentially life-threatening condition, but this device is not intended for or indicated for monitoring of those parameters. The company certainly does not market it for this situation, but sometimes physicians use it that way because they cannot distinguish between the two situations because of their own deficiency as a physician or for convenience, and rarely this leads to a patient death. The company insists their device is not at fault and they cannot control physicians, and a certain amount of misuse is inevitable. The frequency is very low; a high estimate is 0.04% per year. What should be done about these events? My opinion is that they should at least be reported to MAUDE, but the management feels this will seriously and adversely affect their business and argues against it as I noted above. I'm considering leaving them because of this. Am I overreacting or not?
If a manufacturer is well aware of an off-label use, and does nothing official to declare it a contraindication, it becomes part of the intended use even if the manufacturer did not intentionally endorse it. IMO they should clearly add it as a labelling contraindication; then if physicians still ignore it becomes abnormal use rather than misuse (provided that the warnings are prominent and no other practical means can prevent the unwanted use). 0.04% death rate for a device prescribed 100,000 times a year means 40 people dying this way each year! Not acceptable in my world.
 
Last edited:

robert.beck

Involved In Discussions
#23
There are no practical means to prevent this unwanted use, because physicians will make mistakes. This particular situation is clearly identified in the IFU as a contraindication, and in the approved intended use as "Not for ..." but it happens anyways and the company is ignoring it for business reasons which IMO are weak. The company is well-aware of this and other off-label uses. We are in the process of getting a new 510(k) to deal with one of the other less-critical off-label uses but not this one. So given that, what should be reported?
 

indubioush

Quite Involved in Discussions
#24
1. Report the adverse events now (read this) or
2. If you are not allowed to report the events, never ever sign any justification for not reporting.
3. Ensure your risk management file appropriately reflects this "reasonably foreseeable misuse."
 

robert.beck

Involved In Discussions
#25
This is good advice for a difficult situation, and I thank you. How does one mitigate or justify reasonably foreseeable misuse that is not due to the device directly but is a human mistake. I don't think it can be mitigated except by strenuous education and it will still happen to some extent. Do you see a justification based on a low incidence relative to a large number of patient benefiting from the device as valid?
 

Ronen E

Problem Solver
Staff member
Moderator
#26
There are no practical means to prevent this unwanted use, because physicians will make mistakes.
There are almost always ways to REDUCE (not necessarily completely eliminate) mistake rates - research "error proofing". Has any usability study been done on this? If I was auditing this situation the first thing I'd ask to see is a project/activity to methodically look into the issue and implement what's possible. Once that's done, you have grounds to claim "we did everything reasonable".
the company is ignoring it for business reasons which IMO are weak.
(emphasis added)
THAT is the issue. To me it doesn't sound acceptable, and I would seriously look at employment alternatives if I couldn't change that attitude.
So given that, what should be reported?
Reporting is important, but in cases involving death reporting is IMO secondary to attempting prevention. What good is it to the dead person, their family and friends, that the "incident" is reported?...
How does one mitigate or justify reasonably foreseeable misuse that is not due to the device directly but is a human mistake.
First, is it actually a "reasonably foreseeable misuse" or an abnormal use? I already listed the criteria for abnormal use, and I'm quite sure that if you'll look into them seriously you'll already be better positioned, both in terms of mitigating the risk and in terms of having a more solid justification for you actions (or lack thereof).

Second, it's almost always DUE TO THE DEVICE DIRECTLY. In usability terms, what matters is the User Interface, and that includes the labelling too. "Human mistakes" are encouraged by lacking UIs, and can be reduced through better ones. Once you have a serious usability study of the case done and dusted, you'd not only have a better chance of improving, but also be better positioned in claiming "everything reasonable has been done to prevent the abnormal use (=going against a prominent warning/contraindication)".
I don't think it can be mitigated except by strenuous education and it will still happen to some extent.
Do you have a methodical, team-based, documented analysis on file to support that statement, or is it only your own gut feeling? And if that's really the only option, has it been tried, to any extent? In terms of managing abnormal use (you can call it abuse), it makes a difference whether you've actually tried all reasonable means even if they ended up not solving the problem.

"Still happen to some extent" might be acceptable - depending on what that extent is, what's been done to achieve it, what needs to be done to further reduce it and by how much, etc.
 
Last edited:
Thread starter Similar threads Forum Replies Date
J Health Canada Field Action/Recall MDR Requirements Canada Medical Device Regulations 9
J Rework or Recall 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
D Shipping and recall questions ISO 13485:2016 - Medical Device Quality Management Systems 2
R Product Recall - medical devices in the hospital warehouse Canada Medical Device Regulations 2
J Mislabeling - Consider this an FDA notified recall? CFR 806.10 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Recall/Adverse event report procedures in Kuwait/Qatar/Kenya Other Medical Device and Orthopedic Related Topics 3
shimonv Recall - Is there a European regulation for recalls of products EU Medical Device Regulations 4
M Informational The FDA Takes Action to Protect Patients from Risk of Certain Textured Breast Implants; Requests Allergan Voluntarily Recall Certain Breast Implants a Medical Device and FDA Regulations and Standards News 0
S 21 CFR 810 - Medical Device Recall Authority - Contract Manufacturing Other US Medical Device Regulations 1
I Japanese medical device recall requirements & procedure for foreign manufacturers Japan Medical Device Regulations 4
J Recall Effectiveness - Accessory to Medical Devices shipped through Distributors Customer Complaints 3
chris1price Is there a regulatory (MDD, MDR, FDA, ISO, etc) requirement to perform a mock recall? CE Marking (Conformité Européene) / CB Scheme 4
B Health Canada Recall Definition - Seeking Clarification Canada Medical Device Regulations 5
J Calibration Recall - Lost Gages procedure example wanted General Measurement Device and Calibration Topics 4
B Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202? Canada Medical Device Regulations 4
Q Open Medical Device Recall: Are there consequences? Other US Medical Device Regulations 1
R Testing a Product Recall Process IEC 27001 - Information Security Management Systems (ISMS) 8
D Class II Sterile Product Shelf Life Labeling - Should we do a recall? Nonconformance and Corrective Action 4
T FDA Requirements - Class 1 Exempt Medical Device Recall ISO 13485:2016 - Medical Device Quality Management Systems 2
Marc NHTSA calls for nationwide Takata airbag recall World News 1
S Korean Post Market Surveillance and Recall Regulations Other Medical Device Regulations World-Wide 3
B Japan Medical Device Recall Requirements Japan Medical Device Regulations 4
A Medical Device Recall Responsibilities between different Company Groups ISO 13485:2016 - Medical Device Quality Management Systems 3
S Dimensional Issues because of the last years FORD Recall Manufacturing and Related Processes 2
S Medical Device Product Recall per EU Requirements EU Medical Device Regulations 2
M Writing a Manufacturing Recall Procedure Document Control Systems, Procedures, Forms and Templates 3
L Medical Device Reporting vs Recall/Correction: What action to take? US Food and Drug Administration (FDA) 5
Ronen E Abbott Voluntary Recall of FreeStyle lnsulinx? Blood Glucose Meters Other Medical Device and Orthopedic Related Topics 0
W Calibration Recall or Due Dates - Latitude Calibration Frequency (Interval) 14
S New Draft FDA Guidance to distinguish b/w Recall and Product Enhancement & Reporting Other US Medical Device Regulations 15
G Gage Recall Analysis - Threaded Plug Gage has failed Calibration General Measurement Device and Calibration Topics 10
AnaMariaVR2 FDA hits China Plant - OTC DRUG Nationwide Recall US Food and Drug Administration (FDA) 2
M GM Cruze Recall - Seems that the Cruze can catch fire World News 6
Ronen E Nidek Medical Class I recall: Medical Mark5 Nuvo Lite Oxygen Concentrators Other US Medical Device Regulations 0
W How do you determine the initial recall date for instrumentation? General Measurement Device and Calibration Topics 1
D Recall, Safety Notice or Training? - Medical Device with two identical complaints 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
G How to conduct a Toy Recall (College Question) Customer and Company Specific Requirements 2
E What is a "Recall form in a Manufacturing Facility"? Quality Manager and Management Related Issues 6
D Outside the US Recall (in the EU) on a CE Marked Product - File it in the USA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
E AS9100 Equipment Calibration Recall Methods and Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S SFDA Regulation of Medical Device Recall (Interim) - MOH decree No. 82 China Medical Device Regulations 1
A Missing expiry date on Class II Medical Device - Recall/Not? What are EU Requirements EU Medical Device Regulations 2
A Medical Device Safety Alert equal to a Recall? Other US Medical Device Regulations 3
K Vigilance, MDR, Recall, Complaint Procedures - Can this all be one single procedure ISO 13485:2016 - Medical Device Quality Management Systems 3
N Recall Procedure example wanted Document Control Systems, Procedures, Forms and Templates 3
T Is Positive Recall for in-process Product allowed in AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
Q General Medical Device Recall Flowchart wanted ISO 13485:2016 - Medical Device Quality Management Systems 3
M Definition Recall - Definition of AS9100 Section 7.6 F for Recall Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 15
K Product Recall and Withdrawal ? Example procedure wanted Food Safety - ISO 22000, HACCP (21 CFR 120) 7
K Are Withdrawal and Recall a kind of Control of Nonconformity? Food Safety - ISO 22000, HACCP (21 CFR 120) 5

Similar threads

Top Bottom