We all are familiar with the do's and don't of record entries for cGMP.
Don't scribble out, don't use white out, initial and date changes etc.
The question came up in a discussion as to where specially is this stated in the requirements.
We have searched CFR 820 and can't find anything.
Suggestions?
 
Elsmar Forum Sponsor

indubioush

Quite Involved in Discussions
#2
Many requirements come from good laboratory practices. 21CFR58

Sec. 58.130 Conduct of a nonclinical laboratory study.
(e) All data generated during the conduct of a nonclinical laboratory study, except those that are generated by automated data collection systems, shall be recorded directly, promptly, and legibly in ink. All data entries shall be dated on the date of entry and signed or initialed by the person entering the data. Any change in entries shall be made so as not to obscure the original entry, shall indicate the reason for such change, and shall be dated and signed or identified at the time of the change. In automated data collection systems, the individual responsible for direct data input shall be identified at the time of data input. Any change in automated data entries shall be made so as not to obscure the original entry, shall indicate the reason for change, shall be dated, and the responsible individual shall be identified.
 

chris1price

Trusted Information Resource
#4
Everywhere in 21CFR820 it says "the manufacturer shall establish and maintain...……" procedures, records, etc, I would take the word "maintain" to include the documents must be attributable, legible, etc. (ALCOA+), and follow good documentation practices.
 

SMBIZZQA

Starting to get Involved
#5
Thanks everyone for the information, we practice the guidelines for documentation but were never aware of where it was stated. As a converter we don't have any involvement with the pharmaceutical community where many of the guidelines initiate.
Thanks again
 

Schkund

Involved In Discussions
#6
I have some similar questions about this topic, specifically "Good Documentation Practices". My company has a very similar situation to the one that SMBIZZQA is describing... low risk Class I medical devices, no pharmaceutical involvement. The reference to 21 CFR 58 is extremely helpful and something I will definitely be looking into further. However, while the company I work for certainly performs activities that could be considered "nonclinical laboratory studies", we document data for many functions outside of the laboratory as well (especially production), and I'm not sure if the same regulations are necessary in those situations.

I also agree with your statement, chris1price, but I also understand some of the confusion that SMBIZZQA is experiencing. ALCOA+ clearly has very close ties with pharma practices, but the extent to which it should be implemented in the medical device world is a bit fuzzier. The same is true for good documentation practices... they both seem to ALWAYS be associated with pharma, and sometimes medical devices are mentioned on the side. It's very challenging to find resources on data documentation regulations that relate specifically to medical devices without it being a side-note to pharma practices.

So are ALCOA+ and GDP methods identical for medical devices and pharma? I've struggled with this question for some time and I'm not sure where to find a good answer. Maybe someone here on Elsmar knows...???
 

chris1price

Trusted Information Resource
#7
You are correct, the Data Integrity and ALCOA principles have come from the pharma arena. I have not seen an FDA warning letter for a device company discussing data integrity and I am not aware of anything FDA has published that is specific to devices and excludes pharma. Perhaps requiring ALCOA for a device company is a step too far. However if you follow the basic principles of ALCOA, your documentation practices will be very good.
 

yodon

Staff member
Super Moderator
#8
I agree with @chris1price - why WOULDN'T you want to ensure the highest degree of data integrity? The risks, to me, are too high to do otherwise. I have seen 483s basically for Good Documentation Practices, just framed around (lack of) control of records.
 

Schkund

Involved In Discussions
#9
A large part of why I am asking these questions is to defend our current procedure. Several of our upper management personnel come from industries outside of the medical device world and find our current procedures to be "excessive". I would like to say that our current procedures are somewhat necessary, but it's challenging to defend when there isn't much literature on good documentation practices specific to medical devices.

From the perspective of upper management, if we can take away some procedures that truly are excessive, it could free up a lot of time that production and quality have to spend ensuring that our DHRs are correctly completed.
 
Last edited:

yodon

Staff member
Super Moderator
#10
It should be unlikely that good record keeping is too burdensome so there may be some deeper issues. If, for example, to capture a record, operators have to make a lot of changes, what are the reasons? Are you trying to capture too much info? Is it possibly redundant? Do you have too many checkpoints that require too many signatures? If you made an investment in some software to better automate DHR capture, might the long-term savings outweigh the investment cost?

Are you able to have an "adult conversation" with management? What, specifically, do they believe is "excessive"? Specific areas are always better than generalizations.

Have you talked to the folks that are capturing the records to find out what they feel is excessive?
 
Thread starter Similar threads Forum Replies Date
J Incoming Inspection Records using Excel File ISO 13485:2016 - Medical Device Quality Management Systems 2
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
B Complaint Records - Accessing records on Easy Track Software Records and Data - Quality, Legal and Other Evidence 3
P IDEs and Clinical Investigators - Advice About Records? Other US Medical Device Regulations 3
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
Dazza 9001, 14001 and 45001 mandatory documents and records cross reference matrix Process Maps, Process Mapping and Turtle Diagrams 2
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
D Are Supplier Quality Agreements Quality Records ? Other Medical Device Regulations World-Wide 9
qualprod What records have to be maintained? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Retention period of documents and records ISO 13485:2016 - Medical Device Quality Management Systems 1
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
T Controlled Forms or documented requirements for records? ISO 13485:2016 - Medical Device Quality Management Systems 2
L Stage 2 audit - Requirement for 3 months of records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
qualprod Records where apply and stop recording everything? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
qualprod From paper to Electronic (records) in the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
J Controlled information versus defined documents / records ISO 13485:2016 - Medical Device Quality Management Systems 3
I When do FORMs and Log Files become Records? Document Control Systems, Procedures, Forms and Templates 9
S Validation Records - Very young QMS Qualification and Validation (including 21 CFR Part 11) 2
B Record Management - Does the QMS need to control templates of records? Records and Data - Quality, Legal and Other Evidence 17
J Integrity of electronic digital records - Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Do Management Review records have to be on a controlled form? ISO 13485:2016 - Medical Device Quality Management Systems 30
J Medical device related standards or regulations - Disposed records log ISO 13485:2016 - Medical Device Quality Management Systems 2
U Manufacturing solutions and keeping records on quality Content CE Marking (Conformité Européene) / CB Scheme 3
S Maintaining Equipment Records - We use a manual system ISO 17025 related Discussions 1
Q AS9120B flow down to external providers: Records Retention AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
I Date of Quality Management System - Training Records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
I Training records and "grandfather" clauses ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D AS9100 Inspection records - Is there a requirement to have gauge ID and calibration status AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
J Dr Training Records - Training doctors to use a PMA approved medical device Other US Medical Device Regulations 2
Q Documenting Customer Complaints (Records) Customer Complaints 6
M Informational EU – Hearing on the SCHEER preliminary Guidelines on the presence of Phthalates in certain medical devices (Brussels, 04 April 2019) – Summary records Medical Device and FDA Regulations and Standards News 0
I ISO 9001:2015 - On the Job Training and Records Training - Internal, External, Online and Distance Learning 4
A Records required for ISO 9001 clause 7.1.5.1 - Fitness for purpose of the monitoring and measurement resources ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
M Informational USFDA – Radiological Health Regulations; Amendments to Records and Reports for Radiation Emitting Electronic Products; Amendments to Performance Stand Medical Device and FDA Regulations and Standards News 0
L GMDN code wanted - Software for handling records Service Industry Specific Topics 9
I Training records and levels - When does training NOT need a record? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
Q Correction of records - FAR or DFAR? Various Other Specifications, Standards, and related Requirements 3
J Referencing Medical Device in FDA records - Package contains several other components Other Medical Device Regulations World-Wide 1
M AS9100D Competence and Training Records - X-Ray AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
C Karl Fischer Titrators and 21 CTR Part 11 and Paper Records Qualification and Validation (including 21 CFR Part 11) 6
J Calibration/Verification Records (IATF 16949 7.1.5.2.1) IATF 16949 - Automotive Quality Systems Standard 5
I GMP 21 CFR Part 11 Electronic Records Compliance Project Help Qualification and Validation (including 21 CFR Part 11) 9
P Minor Non-Conformance - Maintenance Records Not Fully Completed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
S Where to keep Enterprise Resource Planning software (ERP) Validation Records ISO 13485:2016 - Medical Device Quality Management Systems 1
K Do Forms and Templates ALWAYS become Records? Records and Data - Quality, Legal and Other Evidence 4
S What records are required to show compliance to ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 1
T MDR - Regarding the “Retention Period” of Documents and Records CE Marking (Conformité Européene) / CB Scheme 6

Similar threads

Top Bottom