Reference Only Gages - "For Reference Only" Rationale (Reasoning/Justification)

P

Phil Fields

#1
Hello everyone!
I am not trying to open Pandora’s Box here, I am just trying to prep for a meeting that I have next week. The meeting is to define how we document the rational for "Reference Only Gages".

We have been directed by an external auditor to define the rational used for “For Reference Only” gages.

At this point in time I am not looking for the philosophical discussion (I am sure there are many opinions concerning this issue), what I am looking for is how others have dealt with this.

Thanks,
Phil :thanx:
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#2
Re: "For Reference Only" Rational

The rational for reference only goes something like this:

Reference only stickers are used on any instrument which could elicit a value of some kind, but does not require being in the calibration/ performance verification program. Typically, these are instruments which are integral to some system/process (physically), and are difficult to remove/ isolate (physically).

The accuracy/error present in these instruments will not affect any process, system or quality decision.
 
Last edited:

Randy

Super Moderator
#3
Re: "For Reference Only" Rational

I think...dangerous ground here....that a primary question needing an answer from you is "Why do you need to reference the gages?" If the gages are involved in a those items listed by Brad then they are necessary and operational and not really "reference only" gages. On the other hand, if their use is along the line of a simple Go-No Go to aid someone prior to using the "actual" instruments for decison making then they could be safely assumed to be just "reference gages".

The answer to your question is purely dependant upon the intent and use of the instruments.

Did I confuse you?
 

Jim Wynne

Staff member
Admin
#4
Re: "For Reference Only" Rational

The rational for reference only goes something like this:

Reference only stickers are used on any instrument which could elicit a value of some kind, but does not require being in the calibration/ performance verification program. Typically, these are instruments which are integral to some system/process, and are difficult to remove/ isolate.

The accuracy/error present in these instruments will not affect any process, system or quality decision.
<pedantry>It's rationale. "Rational" is an adjective and "rationale" is a noun.</pedantry>

Although the OP asked to not get into the philosophical end of things, if a device will not affect any process, system or quality decision, why is it necessary?
 

BradM

Staff member
Admin
#5
Re: "For Reference Only" Rational

Randy/Jim, I did not do a good job in my previous post. I am having to go from memory on the verbiage, so I did not do a good job. (Does that excuse me from the wrong word usage also? :) ).

An injection molding machine is a good example. There are like 20 gauges on it that are not used for anything. It's difficult/impossible to remove them, but serve no real purpose. Hence they are classified reference only.

The difficult/impossible to remove is the part I meant by integral.
 
Last edited:
P

Phil Fields

#6
Re: "For Reference Only" Rational

Thank you for the replies so far. Like I said I am preparing for a meeting next week to discuss this issue. My 1st step would be to define what a REFERENCE gage is.
We have had some people here that place a reference only sticker on all gages not under calibration. This includes air and steam gages that are in areas where the is no manufacturing/inspection operations.
 

BradM

Staff member
Admin
#7
Re: "For Reference Only" Rational

Although the OP asked to not get into the philosophical end of things, if a device will not affect any process, system or quality decision, why is it necessary?
That's a really good question, and is a topic of discussion for another time. As I stated in prior conversations, we have had this discussion on the ASQ board, and Akio son steadfastly disagrees with Reference Only. Either check it, or get rid of it.

However, "reference only" does exist, arguably has a place/purpose, and is the subject of rationale. Another use of reference only are pressure gauges on regulators, where their value simply does not matter, and short of capping off the ports on the regulator, it's easier to just leave the gauges installed and marking them "for reference only".
 
M

madannc

#8
Re: Reference Only Gages - "For Reference Only" Rational

Back in 1998 I was working in a micro lab, they had a dedicated autoclave for sterilising, this autoclave had a had timer that at best was +10/-10 seconds out (worst case I remember was +25 secs) when used the parameters for sterilising plates/dishes etc was 8 - 12 mins (or somewhere around that) and we set the timer to 10 mins and labelled the timer for reference only. (knowing that the items would be in the autoclave for at least the minimum amount of time).

was this an incorrect use of the term "for reference only"?

The timer was intergal to the autoclave and could not easily be removed and was also old so timer replacement was not an option, the cost of replacement weighed against the risk did not (at the time) seem a good investment
 
Last edited by a moderator:

BradM

Staff member
Admin
#9
Re: Reference Only Gages - "For Reference Only" Rational

Back in 1998 I was working in a micro lab, they had a dedicated autoclave for sterilising, this autoclave had a had timer that at best was +10/-10 seconds out (worst case I remember was +25 secs) when used the parameters for sterilising plates/dishes etc was 8 - 12 mins (or somewhere around that) and we set the timer to 10 mins and labelled the timer for reference only. (knowing that the items would be in the autoclave for at least the minimum amount of time).

was this an incorrect use of the term "for reference only"?
My opinion only.... I would not have labeled it for reference only, and it would need to be verified. The value mattered; if the cycle went less than 10 minutes due to timer performance, then you would have a problem.

A timer on a microwave is a reference only device at my house. I don't ever know what to set it on, and I sit and look at it to heat to my desired level. The accuracy/error of it is irrelevant.
 

Jim Wynne

Staff member
Admin
#10
Re: "For Reference Only" Rational

That's a really good question, and is a topic of discussion for another time. As I stated in prior conversations, we have had this discussion on the ASQ board, and Akio son steadfastly disagrees with Reference Only. Either check it, or get rid of it.

However, "reference only" does exist, arguably has a place/purpose, and is the subject of rationale. Another use of reference only are pressure gauges on regulators, where their value simply does not matter, and short of capping off the ports on the regulator, it's easier to just leave the gauges installed and marking them "for reference only".
Insofar as the gauges you allude to are concerned, it's easier still not to mark them at all. The concept of rationale comes in when someone asks why they're not in the gage system. You describe what gages need to be in the system (and why), not what gages don't need to be in the system.
 
Thread starter Similar threads Forum Replies Date
P TS16949 - No "For Reference Only" Gages allowed? IATF 16949 - Automotive Quality Systems Standard 49
N For Reference Only gages to appear on Master List of Gages? General Measurement Device and Calibration Topics 10
M SOPs - Reference Documents Section Document Control Systems, Procedures, Forms and Templates 3
B Reference to IEC 60950-1 in Amendment 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
J Marking "Distributed By" to product with distributors' reference-code EU Medical Device Regulations 3
Dazza 9001, 14001 and 45001 mandatory documents and records cross reference matrix Process Maps, Process Mapping and Turtle Diagrams 2
T CSE 5.01 reference on drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
CycleMike GD&T - Hole pattern - Print (attached) has a single Datum Reference Frame Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
J Sample size for creating a data base as a reference to a tested variable Other Medical Device and Orthopedic Related Topics 6
R ASQ reference material clarification - Spiral bound materials allowed in ASQ Exam? Professional Certifications and Degrees 1
D Good Acceptance Activity/Inspection flowchart for reference - Wanted Please Process Maps, Process Mapping and Turtle Diagrams 2
G Customer complaint over a Reference Dimension Manufacturing and Related Processes 9
G Measure 3 times, for calibration report? Customer Owned Reference Standard General Measurement Device and Calibration Topics 4
I Data Backup Plan - Document Reference ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Q How to efficiently and compliantly reference external standards/regulations Other Medical Device Related Standards 2
S Lot assignment for secondary reference standard - Pharmaceuticals Manufacturing and Related Processes 0
V Inspection with 3D Scanning - Reference Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
C Do Calibration Reference Standards require an MSA study? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
J Purpose of Reference Section in SOPs Document Control Systems, Procedures, Forms and Templates 6
M Quality Policy - Standards and reference numbers ISO 13485:2016 - Medical Device Quality Management Systems 14
M Reference to IATF 16949 clause 8.4.3 in clause 8.3.4.4, Is it right? IATF 16949 - Automotive Quality Systems Standard 1
S Reference calibration tools (traceable to NIST) that are not used in production ISO 13485:2016 - Medical Device Quality Management Systems 4
S Contract Manufacturer Question - Whether or not the reference to the ex-CMO ISO 13485:2016 - Medical Device Quality Management Systems 6
B Need to convert Reference Only M&TE to a calibration schedule General Measurement Device and Calibration Topics 3
M Looking for Cross reference of AS9100:2016 to ISO13485 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Print only has Reference Dimensions Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
V QMS / QSR - Reference to Framework - Templates - Procedures for Medical Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
R How to Cross-Reference Controlled Documents Document Control Systems, Procedures, Forms and Templates 3
E IEC 60601-2-25 Reference to 60601-1 2nd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Visual Inspection with No Visual Reference General Measurement Device and Calibration Topics 2
M AS9100 Rev. D - Can anyone recommend a good reference book? Book, Video, Blog and Web Site Reviews and Recommendations 7
B When to use "For Reference Use Only" on documentation Document Control Systems, Procedures, Forms and Templates 4
G BCARs/EASA to ISO9001/AS9100 cross reference matrix/table wanted EASA and JAA Aviation Standards and Requirements 1
J In the ISO 9001:2008 company, can give reference to sister company also. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
E Use of Certified Reference Materials for Quality Assurance General Measurement Device and Calibration Topics 6
A Reference Source for Polarity Symbol Other ISO and International Standards and European Regulations 1
Q Multiple Quality Standards - AS9100 and ISO13485/9001 Cross Reference Matrix Quality Manager and Management Related Issues 3
E Reference Only on Gage or Visual Aid Calibration Frequency (Interval) 4
F Quality Objective Reference Chart Customer Complaints 4
V Good Documentation Practices - FDA Reference US Food and Drug Administration (FDA) 9
M Good reference sites or material to help me brush up quickly on Gage R&R Reliability Analysis - Predictions, Testing and Standards 7
J How does Layout Inspection determine Reference Value? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
W Do "Reference Only" documents have to be physically marked/labeled? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15