Reference Thermocouples are constructed from Thermocouple wires and copper wires. These junctions are insulated from each other and immersed into an Ice Bath so that mV values from a precision dc voltage source can be used to simulate a Thermocouple value.
Omega Engineering makes several Thermocouple Reference Probes (TRP) that are very convient for this task. My question is what is the best way to accurately check them (performance verification)?
Elsmar Forum Sponsor
BradM
Staff member
Admin
#2
Mike, it sounds like you are pretty well down the path already.:agree1: The best method I have found is to purchase a calibration dry block from a reputable dealer. Hart Scientific makes awfully good equipment.
Depending on how many ranges you verify, you might have to buy more than one dry block, as there is not magical piece of equipment. :D
With thermocouples, if you are going to read millivolt signals, you will need to compensate (or null out in an ice bath) the cold junction compensator.
Jerry Eldred
Forum Moderator
Super Moderator
#4
I've done them using GIDEP procedures. I'll have to go look to verify. But I believe it involves putting everything in an ice bath and verifying the mV output on the copper side is zero plus or minus a tolerance with the cold junction and the thermocouple side at zero. It may be later today or tomorrow before I'll be able to look it up.
M
Mike Boothe
#5
I've tried this method numerous times. You have to twist the thermocouple wires together and put this new thermocouple junction into the Ice Bath. I got varying results based on how well the thermocouple junction mechanical connection was made.
Is this an acceptable practice?
F
fast_hedgehog
#6
I have asked this question to so many people, and have never got a satisfactory answer...
The reference junction in the ice bath is itself a thermocouple junction, but no one seems to think they need verifying. Hopefully someone can help!
Thread starter Similar threads Forum Replies Date
N Digital Temperature Indicator Calibration (Thermocouple) - Using a reference junction General Measurement Device and Calibration Topics 1
M SOPs - Reference Documents Section Document Control Systems, Procedures, Forms and Templates 3
B Reference to IEC 60950-1 in Amendment 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
J Marking "Distributed By" to product with distributors' reference-code EU Medical Device Regulations 3
Dazza 9001, 14001 and 45001 mandatory documents and records cross reference matrix Process Maps, Process Mapping and Turtle Diagrams 2
T CSE 5.01 reference on drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
CycleMike GD&T - Hole pattern - Print (attached) has a single Datum Reference Frame Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
J Sample size for creating a data base as a reference to a tested variable Other Medical Device and Orthopedic Related Topics 6
R ASQ reference material clarification - Spiral bound materials allowed in ASQ Exam? Professional Certifications and Degrees 1
D Good Acceptance Activity/Inspection flowchart for reference - Wanted Please Process Maps, Process Mapping and Turtle Diagrams 2
G Customer complaint over a Reference Dimension Manufacturing and Related Processes 9
G Measure 3 times, for calibration report? Customer Owned Reference Standard General Measurement Device and Calibration Topics 4
I Data Backup Plan - Document Reference ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Q How to efficiently and compliantly reference external standards/regulations Other Medical Device Related Standards 2
S Lot assignment for secondary reference standard - Pharmaceuticals Manufacturing and Related Processes 0
V Inspection with 3D Scanning - Reference Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
C Do Calibration Reference Standards require an MSA study? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
J Purpose of Reference Section in SOPs Document Control Systems, Procedures, Forms and Templates 6
M Quality Policy - Standards and reference numbers ISO 13485:2016 - Medical Device Quality Management Systems 14
M Reference to IATF 16949 clause 8.4.3 in clause 8.3.4.4, Is it right? IATF 16949 - Automotive Quality Systems Standard 1
S Reference calibration tools (traceable to NIST) that are not used in production ISO 13485:2016 - Medical Device Quality Management Systems 4
S Contract Manufacturer Question - Whether or not the reference to the ex-CMO ISO 13485:2016 - Medical Device Quality Management Systems 6
B Need to convert Reference Only M&TE to a calibration schedule General Measurement Device and Calibration Topics 3
M Looking for Cross reference of AS9100:2016 to ISO13485 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Print only has Reference Dimensions Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
V QMS / QSR - Reference to Framework - Templates - Procedures for Medical Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
R How to Cross-Reference Controlled Documents Document Control Systems, Procedures, Forms and Templates 3
E IEC 60601-2-25 Reference to 60601-1 2nd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Visual Inspection with No Visual Reference General Measurement Device and Calibration Topics 2
M AS9100 Rev. D - Can anyone recommend a good reference book? Book, Video, Blog and Web Site Reviews and Recommendations 7
B When to use "For Reference Use Only" on documentation Document Control Systems, Procedures, Forms and Templates 4
G BCARs/EASA to ISO9001/AS9100 cross reference matrix/table wanted EASA and JAA Aviation Standards and Requirements 1
J In the ISO 9001:2008 company, can give reference to sister company also. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
E Use of Certified Reference Materials for Quality Assurance General Measurement Device and Calibration Topics 6
A Reference Source for Polarity Symbol Other ISO and International Standards and European Regulations 1
Q Multiple Quality Standards - AS9100 and ISO13485/9001 Cross Reference Matrix Quality Manager and Management Related Issues 3
E Reference Only on Gage or Visual Aid Calibration Frequency (Interval) 4
F Quality Objective Reference Chart Customer Complaints 4
V Good Documentation Practices - FDA Reference US Food and Drug Administration (FDA) 9
M Good reference sites or material to help me brush up quickly on Gage R&R Reliability Analysis - Predictions, Testing and Standards 7
J How does Layout Inspection determine Reference Value? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
W Do "Reference Only" documents have to be physically marked/labeled? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
S ISO 9001 Clauses which reference the 8 Principles of Quality Management ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Where could I get reference to an OQA Sampling Plan of 1 per 1000 ? Quality Manager and Management Related Issues 4
H Can I get some information... CRM (Certified Reference Materials) General Measurement Device and Calibration Topics 3
B Gage R&R Criteria on page 78 MSA Reference Manual Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
Similar threads
Top Bottom