Relevant to ISO11135:2014

Elsmar Forum Sponsor

chris1price

Trusted Information Resource
#2
Yes, take a look at Annex C.3, and the table C.3

The minimum recommended number of BI/PCDs to use is as follows:
a) For MPQ with a product load volume of up to 10 m3, use three BIs per m3 of product volume, with a minimum of five BIs.
b) For MPQ with a product load volume above 10 m3, use one additional BI per additional m3 beyond 10m3.
If BIs are used for routine control use half the number of BIs used during MPQ up to a maximum of 30.
The result of the calculation should be rounded to the next higher number.
 
#3
Yes, take a look at Annex C.3, and the table C.3

The minimum recommended number of BI/PCDs to use is as follows:
a) For MPQ with a product load volume of up to 10 m3, use three BIs per m3 of product volume, with a minimum of five BIs.
b) For MPQ with a product load volume above 10 m3, use one additional BI per additional m3 beyond 10m3.
If BIs are used for routine control use half the number of BIs used during MPQ up to a maximum of 30.
The result of the calculation should be rounded to the next higher number.
 
#4
Thankyou sir for your response.. but in Annexure C it is mentioned as product load volume instead of EtO sterilizer chamber size.. hence is it that product load volume is equivalent to chamber size?
 

chris1price

Trusted Information Resource
#5
That's a good point. I have always tried to fill the chamber with as much product as possible, and the volume should be consistent. If you base your calculation on the "usable chamber volume", you will be safe in your calculation.
 
Thread starter Similar threads Forum Replies Date
U Is there any website out there that summarizes all relevant standards? Medical Device and FDA Regulations and Standards News 1
DuncanGibbons Section 8.3 relevant for design organisations AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
Richard Regalado Informational ISO makes available relevant BCMS standards free for online access Business Continuity & Resiliency Planning (BCRP) 6
D BS EN 62304 - Medical-Relevant Data C.5 - Definition of IEC 62304 - Medical Device Software Life Cycle Processes 5
I ISO 9001:2015 Section 9.3.2 C1 "customer satisfaction and feedback from relevant interested parties" ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
I ISO-9001:2015 and "Relevant Parties" Management Review Meetings and related Processes 2
Ed Panek Relevant questions on Substantial Change Form CE Marking (Conformité Européene) / CB Scheme 3
Q Connecting AS9100 D Clause 4.2 to 9.3.2 b - Interested parties and relevant issues AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
S How to consider the relevant standards during development of ISO13482:2016 for IVD manufacturing Blood grouping Other Medical Device Related Standards 6
C Are zero defect sampling plans relevant to AS9120B Distributor AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
T Relevant internal and external communication - ISO 9001:2015 section 7.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
E IEC 60601-2-57 item 201.12.4.2 - Indication of parameters relevant to safety IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
H CER - Considering similar device/other medical devices for as relevant literature CE Marking (Conformité Européene) / CB Scheme 4
Q Thoughts on Communications relevant to the Quality Management System ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J What does "other relevant management roles" include? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
B Are the GHTF guidance documents still relevant? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 5
O Rebuilding Quality Manual to be more relevant to our business Quality Management System (QMS) Manuals 7
A Guidance for Cleanroom Qualification and Validation relevant for EU and USA ISO 13485:2016 - Medical Device Quality Management Systems 2
E Non-Conformances and the Relevant ISO 9001 Paragraph Nonconformance and Corrective Action 4
E Is clause 7.1 (product realisation) relevant to wholesaler? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
I Component Evaluation - Relevant Element of the Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M 8.3 Control of NC Product - Who is the Relevant Authority for Concession? IATF 16949 - Automotive Quality Systems Standard 10
A How relevant are ASQC CQT, CQE, CMI Completion Diplomas in the Current Work Place Career and Occupation Discussions 9
S How to set Quality Objectives for Relevant Functions Misc. Quality Assurance and Business Systems Related Topics 7
K Does each company department have relevant functions? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
R RAPS Certification Relevant to IT Business Analyst? Professional Certifications and Degrees 4
L Criteria to Establish a Relevant Due Date for Closing Corrective Actions Internal Auditing 8
G Permit to Fly - Is flight testing relevant to our production side? EASA and JAA Aviation Standards and Requirements 6
G 4.5.1 Performance Measurement & Monitoring - relevant information and analysis? Occupational Health & Safety Management Standards 3
W Laboratory Scope - Clause 7.6.3.1 - Traceable to relevant standards - Surface texture General Measurement Device and Calibration Topics 4
S Wrong material Ordered - Which Procedure and Form is relevant? Document Control Systems, Procedures, Forms and Templates 1
S Relevant Packaging Standards for Medical Device Package testing ISO 13485:2016 - Medical Device Quality Management Systems 10
D Definition Measurable Goal vs. Relevant Goal vs. Realistic Goal - Definitions Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 10
A What is "Relevant Functions and Levels" within the organization? Miscellaneous Environmental Standards and EMS Related Discussions 7
T Which ISO Standard is relevant for Refurbished Medical Devices? ISO 13485:2016 - Medical Device Quality Management Systems 3
P Which ISO guideline or system is relevant for Hospitals? Hospitals, Clinics & other Health Care Providers 5
A How to Design Good and Relevant Management System Forms? Misc. Quality Assurance and Business Systems Related Topics 2
Q Improvement Initiatives - what kind of meetings or activities are relevant? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C Flammability of medical textile - any relevant standards? ISO 13485:2016 - Medical Device Quality Management Systems 2
D Is there a need to tie NC found during internal audit to relevant ISO clauses? IATF 16949 - Automotive Quality Systems Standard 13
V Are there any other relevant Documents/Standards about RoHS apart from IEC QC 080000 RoHS, REACH, ELV, IMDS and Restricted Substances 9
T Relevant Standards for Product Standard, Testing, Specification Measurement, etc. ISO 13485:2016 - Medical Device Quality Management Systems 1
N Compliance with Relevant Regulatory Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
O Is there a Relevant Ink / Print Adhesion Standard (ASTM?) Standard? Various Other Specifications, Standards, and related Requirements 8
J Evidence - what you do when the relevant function is missing Records and Data - Quality, Legal and Other Evidence 5
B Logistical or Quality PPM's? Calculation of PPM relevant parts Quality Tools, Improvement and Analysis 28
M The definition of relevant communication - I need creative writing help Miscellaneous Environmental Standards and EMS Related Discussions 2
R Control of Documents 4.2.3 d "Relevant"? Form Revisions Document Control Systems, Procedures, Forms and Templates 5
Raffy Internal Audit - Covering all relevant areas - What areas might NOT be relevant? Internal Auditing 6
M Gauging equipment supplier to the auto industry - Is 17025 relevant to me? ISO 17025 related Discussions 2

Similar threads

Top Bottom