I am planning on running accelerated life cycle test with three different variables, temp, RH, speed of the pump. Is there any equation for this to calculate acceleration factor ...
Also I m planning to keep the pumps in chambers (accelerated temp 50c, 85%RH) for x hrs then take the pumps out of chamber and run the pumps for x hrs at room temp at max speed (observed in the field).... Is this the right approach or should I run the pumps while hey are in chamber ? I
Also I m planning to keep the pumps in chambers (accelerated temp 50c, 85%RH) for x hrs then take the pumps out of chamber and run the pumps for x hrs at room temp at max speed (observed in the field).... Is this the right approach or should I run the pumps while hey are in chamber ? I
It seems to be (as an engineer) that the pumps should be run in the chamber if you want to find out the affect of the environmental factors. If these are long lived units (on the order of at least days even in the "extreme" conditions), the pump will have returned to "normal" environmental levels pretty quickly in the test.