Repeatability Standard Deviation Formula in MSA 4th Edition
jelly1921
In MSA fourth version, Guidelines for determining bias:
analysis of results - numerical
sigma repeatability = sigma r = sum(Xi-Xbar)^2/(n-1)
Is it correct?
I think it should be = (sum(Xi-Xbar)^2/(n-1))^0.5
or (sigma r)^2 = sum(Xi-Xbar)^2/(n-1)
Because in the following page example-bias
sigma r = 0.212019765 [=(sum(Xi-Xbar)^2/(n-1))^0.5]
instead of 0.044952381 [=sum(Xi-Xbar)^2/(n-1)]
Jelly
Jen Kirley
Quality and Auditing Expert
I believe I have found the page in question.
On page 88 (Analysis of results - Numerical) the sigma formula listed in 6) is intended for use with an independent sample.
7) goes on to direct us to determine if the repeatability is acceptable by calculating %EV. It says "If the %EV is large then the measurement system variation may be unacceptable. Since the bias analysis assumes that the repeatability is acceptable, continuing the analysis with a measurement system with a large %EV will lead to misleading and confusing results." The t-statistic formula you cite with its .05 (95% confidence) is then introduced in 8) (pg. 88) and detailed in 9) (pg. 88-89), and is used to produce the results discussed on page 91.
Does this make sense?
jelly1921
Thank Jennifer Kirley
The way is clear enough, no problem no that.
My qestion is that the formula in 6) of page 88 seems not right.
Jen Kirley
I anticipated a further question about it, so I made, and attached a calculator to exhibit the Bias Study in MSA Manual (4th ed) pp 87-92.
You can look at the formula for σb and see that it, and the result agrees with the MSA Manual. The math works out, I believe the formulas are correct.
The tool I have attached is meant to give you the perspective that the manual lacks - it does not work through the problem for us, it just shows you Table III-B 2 with results on page 92. If you know anything about Excel you can look at the formulas and see how they work.
The data array for the histogram is behind the graph.
I hope this helps!
jelly1921
I hope this helps!
I do appreciate what you've done!
In your attachment, you use function to calculate standard deviation, instead of step by step without using the equation shown in 6) of page 88. I put the steps in the attachment.
G Best way to calc uncert for measuring microscope without duplicating repeatability? Measurement Uncertainty (MU) 2
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
G Is it common to do a repeatability check during calibration of an instrument? General Measurement Device and Calibration Topics 5
S "X-MR charts do not separate piece-to-piece repeatability of the process" Statistical Analysis Tools, Techniques and SPC 2
S GRR (Gage Repeatability and Reproducibility) for Material Strength Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
M Repeatability issue in Gage R&R study for inner diameter of a bushing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
B How to Demonstrate Stability and Repeatability in very Short Run Process AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
R Repeatability in MSA - Is repeatability a measure of Location Error Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
T Method Repeatability for MU (Measurement Uncertainty) Measurement Uncertainty (MU) 1
K Wilson Hardness Tester Model 525 Accuracy/Repeatability issues help General Measurement Device and Calibration Topics 4
B Repeatability vs. Instrument Specifications Measurement Uncertainty (MU) 6
Jerry Eldred New ILAC Requirements Regarding Repeatability - Your thoughts on this please Measurement Uncertainty (MU) 11
T Value of K1, K2, K3 in GR&R - I want to measure Repeatability and Reproducibility Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
C How to use GR&R for Machine Repeatability Study when there are no specs? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S Weld Cell Repeatability Studies - Origin - The 'drilled hole panel' Methodology Manufacturing and Related Processes 3
M Minitab Type 1 Repeatability & GRR% of Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
T Gage Repeatability for an Automated Leak Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
S Weight Scale Gage R&R (Gage Repeatability and Reproducibility) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
U Gauge repeatability and reproducibility (Gage R&R) worksheets Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
M Determining Repeatability without Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 22
D Gage R&R Repeatability - How do I know my gauge is in specification Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
A Gage Reproducibility & Repeatability for POKA-YOKE (pass/fail) gage? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
M Gage R&R (Reproducibility and Repeatability) using Minitab 14 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
U Gage R&R (Reproducibility and Repeatability) Understanding for the Layperson Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
E Repeatability for an Electrical Screwdriver Torque Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
F How would you Verify Repeatability and Reproducibility of Measurement Systems? Quality Tools, Improvement and Analysis 5
S Operator Capability - Gage R&R Poor Repeatability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
J Measuring Tape Repeatability and Accuracy - Types of tapes - ECIII, ECII and ECI Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
M Gage repeatability and reproducibility - GRP% Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R MSA Attribute gage repeatability and reproducibility IATF 16949 - Automotive Quality Systems Standard 1
R How to Determine CMM Gage R&R - Accuracy, Repeatability, Reproducibility? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Y Product Stability (Machine Capability) vs. Equipment Repeatability (Gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
N Repeatability Variation and Constant d2* - What is d2* and how is it calculated? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
A Gage R&R - Repeatability and Reproducibility of a newly bought CMM machine Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
D Gage R&R (Repeatability and Reproducibility) for Dummies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 67
P Gage Repeatability (1x1x10) form - 1 part measured 10 times by 1 appraiser Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
P Gage R&R (Repeatability and Reproducability): What are the calculation restrictions Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
D Determining Gauge R&R (Repeatability and Reproducibility) Interval Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A Automated Testing Gage R&R (Repeatability and Reproducibility) Issues Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Courier Gage R&R (Reproducibility and Repeatability) - Automotive Requirements Basics Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M Gage R&R (Repeatability and Reproducibility) Study with one appraiser Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
B Gage R&R (Repeatability and Reproducibility) Destructive Test Method Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
K Repeatability after CMM repair work Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J What is an acceptable Gage R&R (Repeatability and Reproducability) system? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
L Bias, Stability and Repeatability - What is an acceptable percent % ? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Marc Gage R&R (Repeatability and Reproducability) - Tolerance vs. Variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
R Gauge Repeatability and Reproducibility (Gage R&R) on all shop floor gages? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
A Is there a standard on tolerances of substances ? Medical Device and FDA Regulations and Standards News 0