We typically use microscopes to locate the edge of images to determine size. These might be moving microscopes on a CMM, or fixed short distance like filar type eyepieces. Most of the time we measure chrome on glass images (very thin evaporated chromium on a glass substrate) and have well characterized uncertainties for measuring to this type of edge.

My question is: when a customer sends in another type of material (such as images on photo film, or photo paper) , we do not have an uncertainty for that specific material. What is normally done in such a case? Would it be reasonable to have one operator (the tech doing the measurement) measure a feature 10 times and use the std dev from that in the uncertainty calc to determine the k=2 uncertainty for that specific measurement? Or does it need to be more extensive?
 

dwperron

Trusted Information Resource
It would be similar to the uncertainty budget you are already using. Yes, finding the standard deviation would give you a Type A contributor for repeatability. You will still need to include the usual suspects: your measuring device's accuracy, its calibration uncertainty, resolution, environmental contributors, etc.
 

greif

Involved In Discussions
Oh good- simpler than I thought it might be. Thought maybe an R&R and multiple users might be standard for this sort of thing.
 