Hi,

Does the MDR specify that we MUST re-do aging/stability testing if there is a design change? If the design changes, for example the material used for the device is now different and certain specifications are changed to meet newly noted market demand, do we need to re-do aging? Or does some rationalization exist or some guidance document to say its fine to not re-do aging? I'm looking for anything in the MDR and FDA guidance documents?
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
The MDR is a law that is intended to cover ALL medical devices in Europe. It does not have absolute statements like the one you are asking about.
Every product design or design change is unique. You need to evaluate the impact of the change on the shelf-life and the service life of the device.
The evaluation should be part of the engineering changer order process, with engineering rationale and link to risk management and V&V files.
 

Ed Panek

QA RA Small Med Dev Company
Trusted Information Resource
#3
FDA requires any shelf life changes to use the same evaluation methods as in your original submission for approval 510(k). Accelerated aging is ok but must be followed up with real-time aged sample testing.
 

Sammy23

Starting to get Involved
#4
The MDR is a law that is intended to cover ALL medical devices in Europe. It does not have absolute statements like the one you are asking about.
Every product design or design change is unique. You need to evaluate the impact of the change on the shelf-life and the service life of the device.
The evaluation should be part of the engineering changer order process, with engineering rationale and link to risk management and V&V files.
That makes sense. Thank you! I'll refer to the change assessment guidance.
 
Thread starter Similar threads Forum Replies Date
Ed Panek Does this FDA Requirement Apply to international (not USA) distributors for USA based manufacturing companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
C Requirement for Revision of Standard in all Documentation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K ROHs compliance requirement REACH and RoHS Conversations 1
E Language requirement DoC EU Medical Device Regulations 3
I QMS monitoring, measurement, analysis and evaluation requirement - Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
J Requirement for Signature on Document Document Control Systems, Procedures, Forms and Templates 10
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
L Water requirement for Non-sterile topical OTCs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
W Pfmea function requirement and failure mode FMEA and Control Plans 6
Y 510k requirement for bringing in gloves Medical Device and FDA Regulations and Standards News 0
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 2
J When PPAPs are not a contractual requirement APQP and PPAP 9
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 8
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
A The requirement of the PMS plan--suitable indicators and threshold values EU Medical Device Regulations 0
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
K Regulatory requirement of SaMD with machine learning component IEC 62304 - Medical Device Software Life Cycle Processes 3
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
P Regulatory or registration requirement for disinfectant in ASEAN Registrars and Notified Bodies 1
P IATF 16949 requirement - error-proofing in control plan IATF 16949 - Automotive Quality Systems Standard 3
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
B IATF16949 audit requirement - Auditor request UCL and LCL must be show Xbar-R, IATF 16949 - Automotive Quality Systems Standard 7
J 21 CFR 821 Medical Device Tracking Requirement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
W Misinterpretation of requirement acceptable as root cause? Problem Solving, Root Cause Fault and Failure Analysis 19
S ECG Cable Banana to Snap or Tab electrode Adapters -- ISO 10993 Requirement Other Medical Device Related Standards 0
D IATF 16949 Requirement for CMMI in a Global Company Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L Stage 2 audit - Requirement for 3 months of records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Requirement to manufacture under a quality management system EU Medical Device Regulations 4
A Risk Number for each software requirement IEC 62304 - Medical Device Software Life Cycle Processes 7
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
A Several sections of traditional 510(K) have the same or similar requirement of content. US Food and Drug Administration (FDA) 4
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
Q User Requirement Specification for HR (Human Resource Management System) Manufacturing and Related Processes 1
P Obstruction alarm requirement ISO 80601-2-12 Other Medical Device Related Standards 2
P Electrosurgical Device User Need: Cord Flexibility -> Requirement Other Medical Device and Orthopedic Related Topics 4
C Industrial scales and MSA (IATF 16949 requirement 7.1.5.1.1) IATF 16949 - Automotive Quality Systems Standard 30
C ISO 13485 Requirement in Australia/NZ for class 1? ISO 13485:2016 - Medical Device Quality Management Systems 1
N Label Placement Requirement in MDD/MDR EU Medical Device Regulations 3
adir88 MDR requirement: Risk Management Plan for "each device" ISO 14971 - Medical Device Risk Management 5
S Article 61 - MDR Clinical Investigation Requirement EU Medical Device Regulations 9

Similar threads

Top Bottom