Requirements for PCB Component Replacements in an IEC 60601-1 Compliant Device

N

nbarabino

Hello

We have an IEC 60601-1 Third Edition compliant product (USB powered) which has to have one PCB mounted component replaced due to stock concerns. We have found a substitute with identical electrical function and package (DIP).

Is there any compliance test that has to be repeated? Upon our assesment there is nothing that may change the test results with the new component.

Thank you for your help

Regards
Nicol
 
N

nbarabino

Hi

Thanks for your answer.

In fact, no, it's a voltage regulator.

Regards
 
N

nbarabino

Hi, I'm sorry I should have given more details on the device.

I said no because the device has no patient connections (it's a wireless transmitter to communicate with an implantable device), an therefore the failure of the regulator may only have effect on the wireless link, which has been rated as not critical.

We haven't consulted yet our test house.
 

somashekar

Leader
Admin
Hi, I'm sorry I should have given more details on the device.

I said no because the device has no patient connections (it's a wireless transmitter to communicate with an implantable device), an therefore the failure of the regulator may only have effect on the wireless link, which has been rated as not critical.

We haven't consulted yet our test house.
If this determines performance of an implant device, then performance failure is safety critical to the patient. Under your circumstances, I would discuss at length with the test house.
Note: Are both the earlier device and the replaced device carry some sort of safety mark like the UL mark ? Has this been one of the requirements written in your design dossier.
 
N

nbarabino

I'm checking now, but I am pretty sure that the telemetry is not safety critical for the patient. In this case will still be needed to repeat tests?

Regarding UL mark: The new device has it, not the old one, and it was not on the requirements.
 
D

Dimitri2

Just if the telemetry would be safety critical, the whole thing would IMHO cramble down.
You can never rely on radio link for safety. And if you do... Well :)

So, I think you will not have to repeat tests. But you will have to explain in the risk assesment why you think it is not critical. Ask your auditing person (we have a company which visits us once a year just to confirm our overall certification). At the end, this company says to us if tests are sufficient and new are needed.
 

Peter Selvey

Leader
Super Moderator
I think there may be a basic point of confusion here. For a medical device, just about every part is "safety critical" because it relates somehow to the function.

But, for product certification, they have a "scope of test" and/or scope of certification, which is usually far more narrow. For starters, in UL 60601 the scope is limited to direct hazards (see the definition of safety in the standard), which means cases where harm is caused by a physical "thing" coming out of the device such as electricity, heat, fire or mechanical movement. Indirect harm, such as failure to alarm or misdiagnosis are not covered (it is in the 3rd edition, but that's a different story).

Next, UL's product certification scope is further limited to "ELECTRICAL SHOCK, FIRE AND MECHANICAL HAZARDS ONLY" (have a look at your UL Online Certification, and click on the link "See General Information ... ").

So, while the part in question could be safety critical, it's not likely to be something UL is interested in. Keep in mind, just testing to UL 60601-1 itself is already a large amount of work, and certification agencies like UL don't really have the time, resources or expertise to be handling "safety" in the full regulatory context. They are interested in changes to parts which form part of electrical isolation barriers, limit fire, leakage current, over temperature and so on, i.e. direct hazards.
 
N

nbarabino

Thank you both.

I will further analyse the associated risk and let you know if retesting is needed.

Regards
 
Thread starter Similar threads Forum Replies Date
B China GB4943.1 Safety Requirements - PCB Markings China Medical Device Regulations 0
A IEC 60601 Creepage and Clearance Requirements on PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Qualification and Acceptance Requirements of PCB's from Multiple Suppliers Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
JoCam PCB Functional Test Equipment Regulations and Requirements Occupational Health & Safety Management Standards 3
S PCB Handling Requirements - What to do if we drop a PCB during the assembly Manufacturing and Related Processes 4
S Are the standard measurement system requirements too demanding? Manufacturing and Related Processes 0
S Pass-Thru PPAP Requirements Manufacturing and Related Processes 2
Javier Guan CGMP records review requirements? US Food and Drug Administration (FDA) 0
P What are VDA FMEA Requirements for Critical Characteristics? FMEA and Control Plans 0
I Labeling requirements for Type Testing China Medical Device Regulations 1
E MSA requirements for testers running software IATF 16949 - Automotive Quality Systems Standard 2
R Supplied Kit of Part FAI requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
R Regarding safety requirements of the Lid latch mechanism CE Marking (Conformité Européene) / CB Scheme 1
H CZ - Common Zone requirements Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
S What is the relation between EU Annex 1 and ISO 14644 requirements? Manufacturing and Related Processes 2
T Clarity on requirements for indirect suppliers per AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
J US based company no longer selling in the EU: Clinical Evaluation Report and PMS requirements EU Medical Device Regulations 1
T Mandatory requirements for design verification sample 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
R IVDR software requirements CE Marking (Conformité Européene) / CB Scheme 2
B FDA Requirements for medical device to be used for training purposes only US Food and Drug Administration (FDA) 9
T Clarification on requirements IATF 16949 - Automotive Quality Systems Standard 3
Q ISO 13485 Contract Manufacturer Controls and Quality System Requirements ISO 13485:2016 - Medical Device Quality Management Systems 4
J Document Control Requirements within Oracle System Document Control Systems, Procedures, Forms and Templates 1
G FDA requirements for risk analysis US Food and Drug Administration (FDA) 2
Q IATF / AIAG: Control Plans and CC / SC requirements FMEA and Control Plans 4
LincolnQA AS9100D, Clause 8.4.1.1d - actions to take when external providers do not meet requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
N White Label Vs Private Label Requirements Canada Medical Device Regulations 0
T SaMD labeling requirements - MDR IEC 62304 - Medical Device Software Life Cycle Processes 3
R Licensing Requirements for Third Party Logistics Providers Japan Medical Device Regulations 1
P MDR Reporting Requirements for Devices Already covered in a correction under 21 CFR 806 US Food and Drug Administration (FDA) 0
L Meeting training requirements for IATF Manufacturing and Related Processes 4
D Swiss Documentation Requirements for Class 1 Medical Device Other Medical Device Related Standards 0
M Medical Devices Regulations and labelling requirements UAE Other Medical Device Regulations World-Wide 0
J Forging - Cracks requirements Manufacturing and Related Processes 17
O MDR language requirements EU Medical Device Regulations 12
R Does anyone know how ASQ exam requirements are verified? Professional Certifications and Degrees 3
J Bioburden requirements for Procedure Packs under Article 22? EU Medical Device Regulations 3
Sidney Vianna Informational Being developed ISO 30201 Human Resource Management System — Requirements Other ISO and International Standards and European Regulations 6
D PRRC requirements EU Medical Device Regulations 6
Q MSA requirements for IATF IATF 16949 - Automotive Quality Systems Standard 12
B AS9100 Certification Audit Accounting for Additional Customer QMS Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
K Device part of combination product certification requirements Medical Device and FDA Regulations and Standards News 3
L "to be treated as APPLIED PART": is shopping in requirements allowed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Incoming Certification Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
Ed Panek ISO 13485 Reporting Requirements ISO 13485:2016 - Medical Device Quality Management Systems 4
D 8.3.2.1 Review of the Requirements for Products and Services. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Battery Labeling Requirements US Medical Device Regulations 0
D 8.2.3.1 Review of the Requirements for Products and Services AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B Multisite Certification Requirements for ISO 27001 IEC 27001 - Information Security Management Systems (ISMS) 8

Similar threads

Top Bottom