Another idea I'm mulling over, and would appreciate feedback....

I'm considering creating design documents in such a way that, at the end of the design, the requirements, verification protocol and test-results are included in a single document, which is approved in stages.
(NOTE: this approach is probably only viable for relatively simple devices, and assumes all tests are done internally).

The idea is as follows:
  1. Stage 1. The document starts with a table(s) listing the design requirements. This gets approved.
  2. Stage 2. The document gets revised to add table columns to describe the verification instructions for each requirement. This gets approved.
  3. Stage 3. The document gets revised to add table columns to document test activities.

So, for example, a row in the table may evolve as follows:

Stage 1 (two columns per row):
ID: DR-2.3.1
Requirement: Total packaged system weight shall be less than 500g

Stage 2 (add the following columns):
Verification Method: Weight the packaged system using (equipment reference).
Acceptance Criteria: Measured weight is less than 500g.

Stage 3 (add the following columns):
EUT: Device TPS-1023
Tested By/Date: Joe Bloggs, 2014-09-29
Pass/Fail/NA: Pass
Notes/Observations: Measured weight: 472 grams

Advantages:
- Obvious traceability between requirements and verification activities.
- Only one document to maintain.

(potential) Disadvantages:
- This would be both a document and a record.

Thoughts? Is this approach viable/acceptable?
 
Elsmar Forum Sponsor

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#2
Hi Mark,

I see records that also contain instructions and criteria all the time. In fact, I support their premise because the needed information is where it's needed, when it's needed, and presumably current as the document would be getting multiple reviews throughout. What's not to like?
 

Mark Meer

Trusted Information Resource
#3
I support their premise because the needed information is where it's needed, when it's needed...
My thoughts exactly. :)
In previous projects we've had:
1. A design requirements document
2. Several verification protocol documents
3. A test report record for each verification test set
4. A traceability matrix to ensure that all requirements in (1) have been verified by someone in (3) using the methods in (2).

At the end of the day it's a lot to maintain. Particularly when there are changes to a design requirement, nearly all the above needs to be updated.

Furthermore, the "traceability matrix" is a burdensome task in cross-referencing, which would be totally unnecessary if everything was all together in a single table...
 
#4
One thing you need to remember and watch out for with the "one document" approach is how much do you want to put in front of an auditor? The more information presented, the higher risk of generating questions. That's why I would suggest keep your "stage 3" seperate from your one document.

Have your design input requirement, verification method, and design output with a reference to your source of the results (verification report).

An auditor would more likely to spot check and ask for a few reports versus being able to review everything.
 

Helmut Jilling

Auditor / Consultant
#5
One thing you need to remember and watch out for with the "one document" approach is how much do you want to put in front of an auditor? The more information presented, the higher risk of generating questions. That's why I would suggest keep your "stage 3" seperate from your one document.

Have your design input requirement, verification method, and design output with a reference to your source of the results (verification report).

An auditor would more likely to spot check and ask for a few reports versus being able to review everything.
NO, NO, NO...please do not build your system to hide things from an auditor... completely backward approach....


The auditor is there to help you be compliant to requirements, customer requirements, and to help you achieve improvement, better performance and customer satisfaction...by identifying areas which are not compliant or effective.


If you hide things from your auditor because he might say it is a finding, then you are short-circuiting this process. It would be like hiding things from your doctor because you are afraid he might say you are sick...why go to a doctor if you don't want to get well? Why do an audit if you don't want to improve?


If you do not have this kind of relationship with your auditor...then get abetter auditor....
 

matkins

Starting to get Involved
#6
Mark,
Did you ever go further with this idea? If so, would you mind posting it? I'm looking for a means to document design gates and approvals for new and changed products. I like the idea of having it all in one document, showing the complete status of the project.
 

Mark Meer

Trusted Information Resource
#7
For our new projects, we intend to do this, but presently I've got nothing to share.

The idea is however, as described.

Maintain a controlled spreadsheet that is you would add columns to at each stage.
Also, there is a process for revisioning the document:
- v0.x = requirements only stage
- vx.x = requirements + verification protocol stage
- Revision A (B, C, ....) = final design document (requirement+v.protocol+test-record)

So, for example, a single row of this project design document might look as follows:

Stage 1: Design Inputs
|| Req.ID || Design Requirement ||
| 1.2 | Device shall measure 2.4 (+/- 2) mm in diameter |

Stage 2: Verification Planning
|| Req.ID || Design Requirement || Verification Plan || Acceptance Criteria ||
| 1.2 | Device shall measure 2.4 (+/- 2) mm in diameter | Use calibrated calipers to measure the diameter| Diameter measures 2.4(+/-2) mm |

Stage 3: Verification Testing
|| Req.ID || Design Requirement || Verification Plan || Acceptance Criteria || Tested by/date || Equipment Used || EUT || Results ||
| 1.2 | Device shall measure 2.4 (+/- 2) mm in diameter | Use calibrated calipers to measure the diameter| Diameter measures 2.4(+/-2) mm | Jo Personnel, yyyy-mm-dd | (equipment reference) | Unit XXXX | Pass (measured 2.5 mm) |

...That's the idea anyway.
(P.S. hope you can understand my table syntax...pity you can't make proper tables on these forums...)
 
Thread starter Similar threads Forum Replies Date
R Supplier related drawings and verification of process requirements - Source Inspection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Prashant G AS9100 Requirement - Verification of characteristics results with drawing requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
Sidney Vianna Informational AS13100 A - Delegated Product Verification Training Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
Ed Panek Verification & Validation requirements - Patch that adheres to skin ISO 13485:2016 - Medical Device Quality Management Systems 2
E IATF 16949 Cl. 7.1.5.2.1 - Calibration and Verification Records Requirements IATF 16949 - Automotive Quality Systems Standard 1
R CAPA Verification - ISO 13485:2016 Requirements and Objective Evidence ISO 13485:2016 - Medical Device Quality Management Systems 2
D 820.120 UDI (Unique Device Identification) Labeling Verification Requirements Other US Medical Device Regulations 11
D User Requirements Tracing - Verification and Validation Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
E How to write requirements for Medical App Validation and Verification IEC 62366 - Medical Device Usability Engineering 7
M UDI (Unique Device Identification) Verification Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
H Form that will satisfy Design Review and Design Verification Requirements (7.3) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Meeting 7.4.3 Verification of Purchased Product Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
G AS9100 Clause 7.5.1.1 Production Process Verification Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
J Medical Device Software Verification Requirements Design and Development of Products and Processes 7
D Verification Requirements in a Machining Process Qualification and Validation (including 21 CFR Part 11) 6
Q ISO 62304 (Medical Device Software Development) Verification Requirements IEC 62304 - Medical Device Software Life Cycle Processes 1
P Record Retention requirements of Calibration and Verification Service Providers ISO 13485:2016 - Medical Device Quality Management Systems 11
6 Verification Requirements Matrix for ISO 22000 Food Safety - ISO 22000, HACCP (21 CFR 120) 4
S Audit and Calibration question - does verification alone meet the requirements? General Auditing Discussions 5
P Internal Calibration or Verification - Personnel Requirements IATF 16949 - Automotive Quality Systems Standard 3
T Meeting AS9100 Clause 7.4.3 e) Delegation of Verification requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
B Design Verification - Design outputs not verified against customer requirements Design and Development of Products and Processes 6
C Record Requirements in AS 9100 7.4.3 Verification of purchased product Records and Data - Quality, Legal and Other Evidence 5
A Verification of GM Customer Specific Requirements Customer and Company Specific Requirements 3
M Effect of Sampling on Setup Verification Requirements Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
T 21 CFR 820.20 - Quality Planning Requirements? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Composites - Freezer sensor requirements Manufacturing and Related Processes 2
A Requirements for Contract Manufacturers Other Medical Device and Orthopedic Related Topics 3
O Informational Ford Motor Company Customer Specific Requirements for IATF 16949:2016 - 08 Jan 2021 Customer and Company Specific Requirements 0
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
J Requirements for registering Alcohol Wipes (70% IPA) in Canada Medical Device and FDA Regulations and Standards News 4
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
N German National Requirements - Safety Officer/Authorised Rep Other Medical Device Regulations World-Wide 0
S Traceability of requirements to design and risk Design and Development of Products and Processes 3
F USB powered handheld medical device - Isolation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Class I (exempt) testing requirements Other Medical Device Related Standards 0
J Saudi Arabia Technical File Requirements Other Medical Device Regulations World-Wide 1
N Help with basic understanding of Competence requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A Test requirements for marketing in US when RATED values are not of US IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Registration requirements for custom branded device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
A User interface Language Requirements - EU MDR EU Medical Device Regulations 1
A Which guidance do you use for guiding the requirements of IFU for 510(k) submission? Medical Device and FDA Regulations and Standards News 5
O IATF 16949 News Ford Motors Customer Specific Requirements Update - Nov 2020 IATF 16949 - Automotive Quality Systems Standard 5
S ISO 9001 Clause 8.2.3 - Review of the requirements for products and services in a Cafe ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
E Contract manufacturer FDA requirements foreign company US Food and Drug Administration (FDA) 6
JoCam Rental, service & installation - regulatory requirements EU Medical Device Regulations 2
G MDR translation requirements EU Medical Device Regulations 3