Retrospective DHF for an old non-registered device

K

Krthna

#1
Hello!

My backstory is a that we're a very small medical device manufacturer that manufactures (what will be) Class I and II devices (very low-risk), and I have no experience in the Quality Field at all. We haven't had a documented quality system and the required FDA files (DMRs, DHRs, and DHFs).

We're putting those together and it wouldn't be possible without this forum. So much of the Quality Industry is implicit knowledge and experience-based and this forum really helps! So, thank you!!!

I have two questions-

The FDA has us on its radar and we have a warning letter. We are going to be applying for 510(k)s and DeNovos (or PreDeNovos) in the 4th quarter this year.

Some of our products have been on the market for a long time.

We don't have everything that went into the design back then and we do have a number of changes that have occurred since that first product(s) went on sale.

1. The retrospective DHF - does anyone have any advice on where to begin?
Should we begin with the current iteration of the device (fresh start) or should we attempt to piece it together to the best of our ability?
In the former's case - should the primary design input be the older design? Do we still need a planning stage and documentation?

2. We're a small (tiny) company and changes are very rapid in terms of design and process. Once we have a DHF in place (and the DMR), and we apply for and register the product with the FDA, how do we incorporate these changes (assuming we do the documented input, review, output process) before we have to file changes before the FDA?
Should we attempt having more static products and processes after a 510(k)?
For example, we make minor modifications like adding digital output, or an extra input port; or assembly procedure may change based on who's doing what and what they learn on the job.
I understand that all the little changes eventually add up, but it's, sort of like, company culture (we're 3 people strong here) to change things ever so often.

I hope the questions make sense. Any help will be much appreciated!

Regards,
Krthna.
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Oh no, what a mess...

First thing first:

If you have a warning letter from the FDA, that should be top priority.
(a) understand what the letter actually says (and what it doesn't).
(b) respond in a timely manner (15 business days fom receiving the letter), and to-the-point.
(c) prepare a written work plan to implement corrective actions (rather than just corrections), to eliminate all concerns expressed in the letter. Share with the FDA (if you can, in your initial response).
(d) implement the plan and gather evidence, including for effectiveness of corrective actions (= that root causes have been eliminated), and keep the FDA posted.

I really hope you don't intend to continue "business as usual" until Q4. Selling medical devices without proper regulatory clearance is a serious offence. If you don't have your devices on the market since 1976 or earlier (pre-amendments) then having them already on the market doesn't provide any concession; on the contrary.

Second, sort out and document your medical devices scope, and ascertain the classification of each. Class I requirements may be far lesser than class II (BTW, class II is usually referred to as medium risk; the risk as you perceive it has very little to do with regulatory obligations). For instance, many class I devices are exempt from 510k submission and Design Controls (meaning, they don't require a DHF at all). Once you have the ProCodes (FDA's 3-letter code for a specific device type) for each and every medical device you intend to market, you will be able to tell exactly what requirements apply to each. There could be a big difference.

Third, if you make class II devices, you most likely need a QMS compliant with 21 CFR part 820 ("QSR"). In creating one, you will address issues such as Design Control (including coverage of DHF etc.) and Change Control, among many others. You could start here: http://www.fda.gov/medicaldevices/d...medicaldevicequalitysystemsmanual/default.htm

Regarding the retro-DHF: You could go as if you are just introducing a new device to the market, if you made sure that there are no older version on the market (including in the distribution channels). If this is impossible, the DHF's scope would have to include any version that will be left out there. Wherever you begin, start with a new page and follow through the guidance on Design Control (820.30). Don't skip anything - planning, inputs etc. - as awkward as it may feel. If you absolutely can't retrieve or fabricate a certain piece, explain in detail and try to justify. Many times when one tries to justify in detail why something is impossible, they actually realize that a way exists to achieve the same...

...changes are very rapid in terms of design and process.
As a medical devices manufacturer, this is a privilege you will mostly have to give up.

Should we attempt having more static products and processes after a 510(k)?
Yes, definitely.

Once you will have a part-820-compliant Change Control SOP, you will have to follow it for every change. Doesn't necessarily mean that all changes - big or small - are handled at the same level; the SOP should provide decision making tools for keeping controls commensurate with change characteristics. These would also address the question of "should we notify / resubmit to the FDA?".

If your company is too small / lacks regulatory knowledge and experience, perhaps you better hire someone to help you with that. When a company gets a Warning Letter, it's typically a sign they need outside help, IMHO.

Cheers,
Ronen.
 
Last edited:
K

Krthna

#3
Thank you Ronen! That was a very informative reply!

Our problem is that we have no Substantially Equivalent predicate devices and applicable Procodes; the lowered requirements won't work, I think.

Mostly, we sold our products to researchers until recently, when our "rapid design changes" :) allowed us to begin selling to a broader market.

But we will persevere!
 

Ronen E

Problem Solver
Staff member
Moderator
#4
My pleasure.

...Mostly, we sold our products to researchers until recently...
Just to clarify, it doesn't really matter. The only thing that matters is what they are used for (as long as this is in line with your expressed intentions, not against them). If those researchers use them for clinical purposes (either on humans or other animals) they are likely to be deemed Medical Devices.

A device is:

"an instrument, apparatus, implement, machine, contrivance, implant, in vitro reagent, or other similar or related article, including a component part, or accessory which is:

- recognized in the official National Formulary, or the United States Pharmacopoeia, or any supplement to them,
- intended for use in the diagnosis of disease or other conditions, or in the cure, mitigation, treatment, or prevention of disease, in man or other animals, or
- intended to affect the structure or any function of the body of man or other animals,

and which does not achieve its primary intended purposes through chemical action within or on the body of man or other animals and which is not dependent upon being metabolized for the achievement of any of its primary intended purposes."
http://www.fda.gov/medicaldevices/d...nce/overview/classifyyourdevice/ucm051512.htm

Cheers,
Ronen.
 
Thread starter Similar threads Forum Replies Date
Y Retrospective Validation - Class I device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
M Medical Device News FDA News - 03-10-18 - CDRH FY 2019 Proposed Guidance Development and Focused Retrospective Review of Final Guidance Other US Medical Device Regulations 0
P Retrospective Plastics Injection Molding Process Validation Quality Manager and Management Related Issues 2
A Retrospective Effectivness Review of Closed CAPAs ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
D Retrospective (Retroactive) Supplier Qualifications Supplier Quality Assurance and other Supplier Issues 5
P Retrospective Application of EN 60601-1:2006 - Harmonized Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 27
V Representing the Risk Assessment - Prospective or Retrospective and Summarization ISO 14971 - Medical Device Risk Management 2
D Retrospective Validation - Producing Product Over 10 Years Qualification and Validation (including 21 CFR Part 11) 7
L Retrospective Validation of the Performance of Manufacturing Qualification and Validation (including 21 CFR Part 11) 8
S Retrospective Validation of Injection Moulded Class I Medical Devices Qualification and Validation (including 21 CFR Part 11) 2
S Retrospective Medical Device Certification ISO 13485:2016 - Medical Device Quality Management Systems 1
S Asked by customer to supply Retrospective Capability Data IATF 16949 - Automotive Quality Systems Standard 12
Marc OS/2 Warp - A Retrospective After Work and Weekend Discussion Topics 0
A FDA and NB audit of Engineering Drawings in DHF and DMR. Medical Device and FDA Regulations and Standards News 1
D DHF Responsibility after design handover Document Control Systems, Procedures, Forms and Templates 4
R Suggestions for putting together a DHF (Design History File) ISO 13485:2016 - Medical Device Quality Management Systems 3
P DHF, DMR and DHR in MDR? EU Medical Device Regulations 6
A DHF File - Shall the document cover all the variants? Design and Development of Products and Processes 2
S Our products borrow DHF documents from a previous DHF for another product ISO 13485:2016 - Medical Device Quality Management Systems 2
S Is an index necessary for DMR or DHF? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
S DHF File Change Control Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
C Per IEC 62304, are DHF documents Configuration Items? IEC 62304 - Medical Device Software Life Cycle Processes 5
A Hundreds of Medical Devices & Accessories - 3 DHF's? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
E Updating DHF (Design History File) to include Alternative Component 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
V What is required in terms of the DHR, DMR, DHF? Other Medical Device and Orthopedic Related Topics 4
K ISO 13485 and DHF (Design History File) requirements ISO 13485:2016 - Medical Device Quality Management Systems 9
O Distributors and ISO 13485 - DHF and DMR Responsibilities ISO 13485:2016 - Medical Device Quality Management Systems 7
M Design History File (DHF) for New Generation of Product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
E Design Dossier/DHF Updates - What to Keep EU Medical Device Regulations 2
N 510k without DHF (Design History File) that we bought a 3 years ago US Food and Drug Administration (FDA) 4
D Incomplete DHF: Re-do or plug gaps? Design and Development of Products and Processes 6
A DHF (Design History File) for a Legacy Product (Class iii Medical Device) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
B When To Share Your DHF (Design History File) With A Customer US Food and Drug Administration (FDA) 1
J How to structure a DHF when all manufacturing is done by Contract Maufacturers 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S Cost Targets as Design Inputs in your DHF (Design History File) US Food and Drug Administration (FDA) 7
renenatasha Usage of web articles in DHF (Design History File) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
A Medical Device DHF (Design History File) - One per Design Change? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
L Creating DHF (Design History File) for Medical Device Systems Design and Development of Products and Processes 8
K How to start a Design History File (DHF) for Medical devices Company 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R Design History File DHF Practices and Contract Manufacturer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
D Design History File (DHF) - Buy-For-Resell Products 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
D DHF Review - When does the FDA review the DHF? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M What are typical Roles & Responsibilities of DHF (Design History File) Librarian? US Food and Drug Administration (FDA) 6
N Control of Medical Device DHF (Device History File) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
H Paper based DHF (Design History File) updates for Software Updates 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Q DMR-Device Master Record vs DHF-Design History File vs DHR-Device History Record 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 50
C Remediating a Design History File (DHF) for an IVD developed prior to 1997 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Design History File (DHF) Maintenance Information wanted 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
O DHF (Design History File) Template Other US Medical Device Regulations 13
H Working from a previous design history. How can I best leverage the exisitng DHF? Other US Medical Device Regulations 4