Revalidation under VDmax25 following material change
Hi all
We are looking at a material change to a single component in one of our devices due to supply issues.
The gamma sterilisation process has been validated using VDmax25.
The new material is a slightly different hardness of the same material grade used in the original. The volume and surface area of the component (and device) have not changed and the difference in density between the original and proposed components is <1%.
The manufacturer, process, manufacturing environment, handling etc. will not change.
What I was wondering is whether this would/should trigger a sterilisation revalidation?
Thanks in advance.
planB
Have a look into ISO 11137-1, section 12.5.2, Assessment of change:
A change in product, its package or the presentation of product for sterilization shall be assessed for its effect on the appropriateness of the sterilization process. Those parts of process definition or PQ that have to be undertaken shall be determined based on the nature of the change. The outcome of the assessment, including the rationale for decisions reached, shall be recorded (see 4.1.2).
From what you write, you seem to have gathered evidence that you do not significantly change "the presentation of product". You may also want to demonstrate that your change in material has no significant impact on the bioburden of your final device, either with a rationale or e.g. by a dose audit together with a bioburden determination of your changed final device.
Hope this helps,
Philip B
You could also talk to your test lab, they often have their own guidelines on what changes would necessitate a new sterilisation validation.
I What revalidation is necessary if we change parts in a WFI loop ? Qualification and Validation (including 21 CFR Part 11) 1
C Revalidation of Process Equipment for Equipment Transfered to New Facility. ISO 13485:2016 - Medical Device Quality Management Systems 5
R Revalidation machine line and Stability Studies impact. Medical Device and FDA Regulations and Standards News 3
V How to Add Annual Revalidation requirements to Control Plan? FMEA and Control Plans 5
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
A Revalidation of Training - ISO 17025 ISO 17025 related Discussions 4
V Plating Supplier PPAP Revalidation Requirements - A-Coat GM 9984132 Plating Process APQP and PPAP 1
J Yearly Cleanroom Revalidation Test Requirements - Questions for users clean room ISO7 Qualification and Validation (including 21 CFR Part 11) 5
L EO Sterilization Cycle Revalidation Intervals ISO 13485:2016 - Medical Device Quality Management Systems 2
O Requalification/Revalidation Activities - Dietary Supplement Manufacturing Equipment Design and Development of Products and Processes 3
B Autoclave Revalidation Methods Qualification and Validation (including 21 CFR Part 11) 5
A Moving Equipment (Injection Molding Machines) in the Plant - Revalidation Required? ISO 13485:2016 - Medical Device Quality Management Systems 7
Y Difference between Cl. 7.4.1 Supplier Re-Evaluation and Cl. 7.5.2 Revalidation Supplier Quality Assurance and other Supplier Issues 4
S Gamma Sterilization: Change in Bulk Density vs. Revalidation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Is Annual Revalidation Required? ISO 13485 and FDA ISO 13485:2016 - Medical Device Quality Management Systems 11
F Equipment Revalidation - Refrigerated rooms (2?C ? 8?C) Food Safety - ISO 22000, HACCP (21 CFR 120) 1
Douglas E. Purdy 7.5.2 e) Revalidation - Let me run this by you for verification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
J Can we register non-device clinical decision support software under draft guidance? Other US Medical Device Regulations 5
L Significant subcontractor update, still under MDD Certification, possible? CE Marking (Conformité Européene) / CB Scheme 3
C UDI for consumable/replacement components under the MDR EU Medical Device Regulations 2
R How to obtain OBL licence under Indian MDR, 2017 Other Medical Device Related Standards 2
M Classification of Instruments under EU MDR EU Medical Device Regulations 1
A % of defects on the whole batch based on result from inspection under AQL Level II Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
A Transactions under MDR Medical Device and FDA Regulations and Standards News 3
H Still having trouble as to what equipment falls under 8.5.1.1 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M Post Market Safety Update Report for devices that will be up-classified under MDR but are currently under MDD EU Medical Device Regulations 6
chris1price Notified Body responsible for product safety under MDD? EU Medical Device Regulations 4
K Re-packaging under MDR EU Medical Device Regulations 3
C Requirements for distributors under MDR: translation EU Medical Device Regulations 0
A Can anyone share a Distribution Agreement template under MDR 2017/745? EU Medical Device Regulations 0
N OBL under the MDR. Which are the obligations of the OEM? EU Medical Device Regulations 0
U Distributor obligations under MDR EU Medical Device Regulations 5
M Extension to shelf life of new product under MDR EU Medical Device Regulations 1
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4
V New Product Model CE Mark Certification under MDR EU Medical Device Regulations 0
lanley liao How to keep the manufacturing capability under the API monogram Oil and Gas Industry Standards and Regulations 6
S Are EC type examinations still being conducted under MDR? EU Medical Device Regulations 5
V Sister companies selling same medical device under different names ISO 13485:2016 - Medical Device Quality Management Systems 3
Sidney Vianna What ISO Standard (under the TC 176) supports the UN Sustainable Development Goal #10? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 11
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
F Plastics cans shelf life determination under static load Manufacturing and Related Processes 1
L Classification under MDR EU Medical Device Regulations 1
K Biological indicator system classification under MDR EU Medical Device Regulations 2
K Advertising and promotion of device under investigation China Medical Device Regulations 1
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
A First CE-mark class III implantable device under MDD EU Medical Device Regulations 8
M Requirement to manufacture under a quality management system EU Medical Device Regulations 4