Risk Control Implemented in Software
yodon
Scratching my head over 60601-1, 14.6.2 (PEMS), Risk Control. For risk controls implemented in software, the standard says:
Suitably validated tools and PROCEDURES shall be selected and identified to implement each RISK CONTROL measure. These tools and PROCEDURES shall be appropriate to assure that each RISK CONTROL measure satisfactorily reduces the identified RISK(S).
I'm having trouble thinking of what tools or procedures would be suitable for implementing a risk control in software. I can think of some possibilities for procedures (maybe an 'elevated' review and test process) but the only "tools" I can think of would be a compiler (but I'm in the camp that you can't really validate a compiler - and the best proof of output is all the testing that IS done). Examples would be appreciated.
yodon
Thanks, QAMTY. I don't have those standards... would you mind posting maybe 1 or 2 examples cited? I'm just trying to get a feel for what they were thinking.
Q
QAMTY
Im posting some portions of the text as an example.
xxxxx
B.6.2 Use
The HAZOP technique was initially developed to analyse chemical process systems, but has
been extended to other types of systems and complex operations. These include mechanical
and electronic systems, procedures, and software systems, and even to organizational
changes and to legal contract design and review.
The HAZOP process can deal with all forms of deviation from design intent due to deficiencies
in the design, component(s), planned procedures and human actions.
It is widely used for software design review. When applied to safety critical instrument control
and computer systems it may be known as CHAZOP (Control HAzards and OPerability
Analysis or computer hazard and operability analysis).
XXXXX
B.13.5 Outputs
The primary output of FMEA is a list of failure modes, the failure mechanisms and effects for
each component or step of a system or process (which may include information on the
likelihood of failure). Information is also given on the causes of failure and the consequences
to the system as a whole. The output from FMECA includes a rating of importance based on
the likelihood that the system will fail, the level of risk resulting from the failure mode or a
combination of the level of risk and the ‘detectability’ of the failure mode.
FMECA can give a quantitative output if suitable failure rate data and quantitative
consequences are used.
B.13.6 Strengths and limitations
The strengths of FMEA/FMECA are as follows:
widely applicable to human, equipment and system failure modes and to hardware,
software and procedures;
identify component failure modes, their causes and their effects on the system, and
present them in an easily readable format;
avoid the need for costly equipment modifications in service by identifying problems
early in the design process;
identify single point failure modes and requirements for redundancy or safety systems;
provide input to the development monitoring programmes by highlighting key features
to be monitored.
Limitations include:
they can only be used to identify single failure modes, not combinations of failure
modes;
unless adequately controlled and focussed, the studies can be time consuming and
costly;
they can be difficult and tedious for complex multi-layered systems.
B.13.7 Reference document
IEC 60812, Analysis techniques for system reliability – Procedures for failure mode and effect
analysis (FMEA)
XXXXXX
B.20.4 Process
The HRA process is as follows:
Problem definition, what types of human involvements are to be investigated/assessed?
Task analysis, how will the task be performed and what type of aids will be needed to
support performance?
Human error analysis, how can task performance fail: what errors can occur and how
can they be recovered?
Representation, how can these errors or task performance failures be integrated with
other hardware, software, and environmental events to enable overall system failure
probabilities to be calculated?
Screening, are there any errors or tasks that do not require detailed quantification?
Quantification, how likely are individual errors and failures of tasks?
Impact assessment, which errors or tasks are most important, i.e. which ones have the
highest contribution to reliability or risk?
Error reduction, how can higher human reliability be achieved?
Documentation, what details of the HRA need to be documented?
In practice, the HRA process proceeds step-wise although sometimes with parts (e.g. tasks
analysis and error identification) proceeding in parallel with one another.
B.20.5 Output
Outputs include:
a list of errors that may occur and methods by which they can be reduced – preferably
through redesign of the system;
error
Xxxx
B.23 Sneak analysis (SA)and sneak circuit analysis (SCI)
B.23.1 Overview
Sneak analysis (SA) is a methodology for identifying design errors. A sneak condition is a
latent hardware, software or integrated condition that may cause an unwanted event to occur
or may inhibit a desired event and is not caused by component failure. These conditions are
characterized by their random nature and ability to escape detection during the most rigorous
of standardized system tests. Sneak conditions can cause improper operation, loss of system
availability, program delays, or even death or injury to personnel.
B.23.2 Use
Sneak circuit analysis (SCA) was developed in the late 1960s for NASA to verify the integrity
and functionality of their designs. It served as a useful tool for discovering unintentional
electrical circuit paths, and assisted in devising solutions to isolate each function. However,
as technology advanced, the tools for sneak circuit analysis also had to advance. Sneak
analysis includes and far exceeds the coverage of sneak circuit analysis. It can locate
problems in both hardware and software using any technology. The sneak analysis tools can
integrate several analyses such as fault trees, failure mode and effects analysis (FMEA),
reliability estimates, etc. into a single analysis saving time and project expenses.
B.23.3 Input
Sneak analysis is unique from the design process in that it uses different tools (network trees,
forests, and clues or questions to help the analyst identify sneak conditions) to find a specific
type of problem. The network trees and forests are topological groupings of the actual system.
Each network tree represents a sub-function and shows all inputs that may affect the subfunction
output. Forests are constructed by combining the network trees that contribute to a
particular system output. A proper forest shows a system output in terms of all of its related
inputs. These, along with others, become the input to the analysis.
B.23.4 Process
The basic steps in performing a sneak analysis consist of:
data preparation;
construction of the network tree;
evaluation of network paths;
final recommendations and report.
B.23.5 Output
A sneak circuit is an unexpected path or logic flow within a system which, under certain
conditions, can initiate an undesired function or inhibit a desired function. The path may
consist of hardware, software, operator actions, or combinations of these elements. Sneak
circuits are not the result of hardware failure but are latent conditions, inadvertently designed
into the system, coded into the software program, or triggered by human error. There are four
categories of sneak circuits:
On the web you may find examples
Hope this helps
Marcelo
Compilers are one example, other are requirements traceability and configuration management tools (these are mentioned in 80002-1.
Please note that the focus on the requirement you mentioned is the need for validation of these procedures and tools, so as to use procedures and tools that are know to be good and fit for purpose.
Peter Selvey
I found that IEC 62304, while often mis-interpreted, is a fairly well written document, in that authors avoided wishful thinking type of requirements and had some experience in real design.
IEC 60601-1 Clause 14 is the opposite. It is poorly written, many wishful thinking items and has mistakes. For example, the clause for architecture looks like something that would fit specifications.
In the case of Clause 14.6.2 it is a wishful thinking kind of statement that might have a place in a guide but does not fit a standard. And, if you read it closely, you will find that no records are required.
In practice, manufacturers will use a mix of well established technology and home baked technology. The home baked stuff should be treated with a bit more suspicion in the design process. That's really what the clause is getting at - more of a hint towards using well established technology rather than home baked.
The idea that you can "validate" every tool and procedure used is not practical. Hence, the standard doesn't require any records to be kept.
yodon
Thanks, Peter, I agree on every point you made. My 'challenge' is that the test lab (assessing compliance to the standard) is asking how we fulfilled this clause. I've got my tap shoes all laced up and am starting to dance. :)
E Do anyone have document of automotive production risk and control of risk? Lean in Manufacturing and Service Industries 1
S Risk control through Information for safety ISO 14971 - Medical Device Risk Management 12
T IEC 62304 : Risk control for SaMD IEC 62304 - Medical Device Software Life Cycle Processes 8
M Does 4.5 - Alternative RISK CONTROL apply to the Particular Standards? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Use Error Risk Controls and Control Verification ISO 14971 - Medical Device Risk Management 6
K Defining risk control measures IEC 62304 - Medical Device Software Life Cycle Processes 14
M ISO14971:2019 - Verification of implementation and effectiveness of risk control ISO 14971 - Medical Device Risk Management 14
R Risk control measures as per ISO 14971 ISO 14971 - Medical Device Risk Management 6
adir88 Documenting Risk Control Option Analysis ISO 14971 - Medical Device Risk Management 8
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
D Performance specification as a Risk Control Measure, EN 14971 ISO 14971 - Medical Device Risk Management 7
Ashok sunder Is it possible to reduce Risk likelihood and impact Post control Ranking after corrective action taken for risk? FMEA and Control Plans 1
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
qualprod ISO 9001 Risk control method - What could be the better way to control risks? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Can Scope of Equipment Control be Tied to Risk? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Y Training as a risk control for ISO 14971 ISO 14971 - Medical Device Risk Management 13
Q How to Analyze Risk if is out of your control ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
F Medical Device HACCP (Hazard Analysis and Critical Control Point) Risk Management ISO 14971 - Medical Device Risk Management 2
K Risk Reduction by Risk Control: IEC:62304-Class C ISO 14971 - Medical Device Risk Management 15
A 5.5.3 - Software Unit Acceptance Criteria (Risk Control Measures) IEC 62304 - Medical Device Software Life Cycle Processes 3
A Effectiveness of Risk Control Measures ISO 14971 - Medical Device Risk Management 4
M Control Measures for Hazards already deemed Low Risk ISO 14971 - Medical Device Risk Management 6
M ISO 14971:2012 - Verification of Implementation of Risk Control Measures ISO 14971 - Medical Device Risk Management 12
Q Applying Risk to the Medical Device Document Control Program Document Control Systems, Procedures, Forms and Templates 3
S What to do if no further control possible to reduce the OHS risk? Occupational Health & Safety Management Standards 16
S What does "Operational Review" mean as tool - Effectiveness of Internal Control Risk Management Review Meetings and related Processes 3
S Risk Management and Revision Control ISO 14971 - Medical Device Risk Management 1
Marc Millions of Ford vehicles have fire risk part - Cruise-control deactivation switch World News 0
A Maturity Model of Organisations - Administrative Risk Control And CI Misc. Quality Assurance and Business Systems Related Topics 0
T AS9100D Risk-Based Internal Audit Schedule AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
thisby_ Installation Related Issues and Risk Management ISO 14971 - Medical Device Risk Management 5
W Reconciling FMEA RPN ratings with Risk Acceptability ISO 14971 - Medical Device Risk Management 11
D How to address the content deviation of 'cannot apply criteria of risk acceptability prior to...' ISO 14971 - Medical Device Risk Management 1
Doninina Risk management file according MDR or ISO 14971:P2019 ? EU Medical Device Regulations 2
T Risk based CA AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
T IVD Risk - destruction of patient samples - Harm to property? ISO 14971 - Medical Device Risk Management 5
R Using RPN to Confirm Risk Reduced to an Acceptable Level Risk Management Principles and Generic Guidelines 12
T IVD Device Software - Risk Classification IEC 62304 - Medical Device Software Life Cycle Processes 16
G Help:Risk Management - Accessories US Food and Drug Administration (FDA) 1
N Writing Risk Management procedure for small manufacturing and we don't know where to start. Manufacturing and Related Processes 9
E How to risk assess tooling? For a medical device and is it needed??? Manufacturing and Related Processes 2
M Clinical evaluation interface with the risk management process EU Medical Device Regulations 9
L Risk analysis Manufacturing and Related Processes 4
J Risk Analysis for Proficiency Testing Reliability Analysis - Predictions, Testing and Standards 1
J ISO 10993-1:2018 Format to Perform Risk Management Process US Food and Drug Administration (FDA) 1
B Risk Management Procedure updates needed for 14971:2019 ISO 14971 - Medical Device Risk Management 11
M What is the Risk of Using Obsolete Versions of C=0 & ANSI/ ASQ Z1.4 Sampling Plans? ISO 13485:2016 - Medical Device Quality Management Systems 8
D AS9100D 8.4.2 Note 2 Significant Operational Risk AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
A Calculating Risk Estimation ISO 14971 - Medical Device Risk Management 29
M Intended Use vs Actual Use and Scope of Risk Management EU Medical Device Regulations 8