Risk Evaluation Matrix-Product Portfolio

alimary15

Involved In Discussions
Hello,

a quick question :)

Does it make sense to have the same Risk Evaluation Matrix for completely different products?

Shouldn?t different products ( such as Apps, or heavy machines) have different tresholds of acceptable/unacceptable risks?

What is the best practice? Is it better to your experience to have 1 Evaluation Matrix for all, or to define 1 evaluation Matrix for each single product/device?

Thank you very much!
 

Marcelo

Inactive Registered Visitor
The evaluation matrix you mentioned is the graphical interpretation of the risk acceptability criteria.

The criteria should be based on the risk management policy, and should be created for each device.

So yes, you should have different ones for different devices. The only rationale for having the same for different devices is if the devices have generally the same risks, for example, if all are lung ventilators. If you manufacture a lung ventilator and an implant, or a lung ventilator and ultra-sound machine, there?s no way the risk acceptability criteria would be the same.

However, most of the time people use only one. This is common practice, not best practice, thought.
 
Last edited:

alimary15

Involved In Discussions
The evaluation matrix you mentioned is the graphical interpretation of the risk acceptability criteria.

The criteria should be based on the risk management policy, and should be created for each device.

So yes, you should have different ones for different devices. The only rationale for having the same for different devices is if the devices have generally the same risks, for example, if all are lung ventilators. If you manufacture a lung ventilator and an implant, or a lung ventilator and ultra-sound machine, there?s no way the risk acceptability criteria would be the same.

However, most of the time people use only one. This is common practice, not best practice, thought.
Thanks for the feedback !

In case the devices are totally different I would go for the option to have different acceptability criterias. However, if having a general and "good-for-all" policy, would it then make any sense to discuss possible residual risks in a risk/benefit analysis?

By having an acceptability criteria that is unique for all, wouldn?t there be a risk of classifyng as acceptable risks that might not be acceptable? I am concerned about and if it is possible to lose any kind of information within the RM process by doing so.

What is your opinion?

Thanks
 

Marcelo

Inactive Registered Visitor
However, if having a general and "good-for-all" policy, would it then make any sense to discuss possible residual risks in a risk/benefit analysis?

The risk management policy is the high-level policy on how the enterprise deal with risk. Policies are direction, but really not directly actionable. That why you need to define the criteria for each device based on the policy. Please see an example of policy discussion in this thread - http://elsmar.com/Forums/showthread.php?t=52843

By having an acceptability criteria that is unique for all, wouldn?t there be a risk of classifyng as acceptable risks that might not be acceptable? I am concerned about and if it is possible to lose any kind of information within the RM process by doing so.

Yes, that why is makes no sense to have a risk acceptability criteria for all devices, as I mentioned. This problem might even happen if the devices are similar but have modifications that can impact the risks.
 

alimary15

Involved In Discussions
The risk management policy is the high-level policy on how the enterprise deal with risk. Policies are direction, but really not directly actionable. That why you need to define the criteria for each device based on the policy. Please see an example of policy discussion in this threadhttp://elsmar.com/Forums/showthread.php?t=52843

Sorry for the bad use of words- I was always speaking about the risk criteria of acceptability!

Thanks for the help!
 

alimary15

Involved In Discussions
This problem might even happen if the devices are similar but have modifications that can impact the risks.

In this case for example, when upgrading the same device, it is practice to start from the previous RA and then move forward to identify new possibile risks that are coming from the new modification.

My only concern is using the same acceptability criteria for software, hardware, and different kind of devices. I think it would make things way more general.

I will try to discuss this further within my team and try to change things.

Thanks for the confirmation :thanx:
 
Last edited by a moderator:

Marcelo

Inactive Registered Visitor
Ah, ok, well, people generally use those concepts interchangeably, but they are very different, which causes confusion in the discussion and understanding.

So, if your question is

However, if having a general and "good-for-all" risk acceptability criteria, would it then make any sense to discuss possible residual risks in a risk/benefit analysis?

Yes, it would, because the residual risk are related to risk control measure, and even if similar devices would have similar risks, it does not mean that they would have the same risk control measures (in fact, newer devices in a device family usually have different risk controls measures to some risks because that?s how technology evolution works.
 

Marcelo

Inactive Registered Visitor
In this case for example, when upgrading the same device, it is practice to start from the previous RA and then move forward to identify new possibile risks that are coming from the new modification.

Yes, and this is acceptable by ISO 14971 IF you perform and evaluation and concludes that you can use the proviso RA.

My only concern is using the same acceptability criteria for software, hardware, and different kind of devices. I think it would make things way more general.

Exactly. They probably need separate ones, based on the same high level risk management policy.
 

alimary15

Involved In Discussions
Ok great thank you!

Then I might have a more complex question:

Let?s consider a scenario where we have different modules that could be or could be not present on the field and that are sold as a complete system to the costumer.

Then what is it the best practice? Make RA for each single module? Shall RA also be made to the complete and overall system to include possible risks arising from interactions between each single module?

What?s the best practice in this case?

Thanks
 

Marcelo

Inactive Registered Visitor
Then I might have a more complex question:

Let?s consider a scenario where we have different modules that could be or could be not present on the field and that are sold as a complete system to the costumer.

Then what is it the best practice? Make RA for each single module? Shall RA also be made to the complete and overall system to include possible risks arising from interactions between each single module?

What?s the best practice in this case?

Thanks

Risk management (as defined in ISO 14971 and as required by most regulations) is applicable to the medical device as a whole.

So yes, you would need to focus on the complete device, and as you mentioned, part of the focus is on interactions between each single module - in fact, think of it as interactions in the different configurations.
 
Thread starter Similar threads Forum Replies Date
M Clinical evaluation interface with the risk management process EU Medical Device Regulations 9
T Biological Evaluation (10993) & Risk Management ISO 14971 - Medical Device Risk Management 9
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
U Changes to Internal Processes and Risk Evaluation - Mitigations Risk Management Principles and Generic Guidelines 10
M Informational Work in progress at the FDA for biological evaluation – Color Hazard and RISk calculator (CHRIS) Medical Device and FDA Regulations and Standards News 0
Q Evidence of precautions (clinical evaluation report, risk management report) EU Medical Device Regulations 6
qualprod Qualitative or Quantitative? Risk method evaluation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Rincewind Clinical Evaluation of very low risk medical devices EU Medical Device Regulations 3
J Simple supplier evaluation qualification process form that includes Risk Assessment Document Control Systems, Procedures, Forms and Templates 2
Sam Lazzara ISO 14971 Clause 7 - Evaluation of Overall Residual Risk Acceptability ISO 14971 - Medical Device Risk Management 3
J FMEA Evaluation of "User Risk" - Wording FMEA 3rd party advisors ISO 14971 - Medical Device Risk Management 3
S Risk Analysis and Risk Evaluation - Biomedical company ISO 14971 - Medical Device Risk Management 6
B Chrysler requirement of Initial Risk Evaluation IATF 16949 - Automotive Quality Systems Standard 2
M Q: In what circumstances can a clinical investigator submit an EFS/IDE for the Significant Risk Device? US Medical Device Regulations 1
T Article 22 procedure pack risk analysis EU Medical Device Regulations 1
S Risk analysis on patient monitor ISO 14971 - Medical Device Risk Management 5
John Broomfield NATURAL CAPITAL AT RISK Misc. Quality Assurance and Business Systems Related Topics 7
MaHoDie How deep should be risk control tracebility IEC 62304 - Medical Device Software Life Cycle Processes 3
D Objective Measures for Risk Acceptability? ISO 14971 - Medical Device Risk Management 6
D Risk Based Sample Size and Standards Compliance ISO 14971 - Medical Device Risk Management 2
austin_howell Labeling Controls affecting Probability and Risk ISO 14971 - Medical Device Risk Management 7
J Determination of software safety class (62304) prior to software risk analysis ISO 14971 - Medical Device Risk Management 3
Q Risk acceptance for missing DHR info and saving records ISO 13485:2016 - Medical Device Quality Management Systems 11
G FDA requirements for risk analysis US Food and Drug Administration (FDA) 2
T Corrective Action Risk Matrix AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
B Fluorescent Video System risk class EU Medical Device Regulations 6
A Risk Assessment for ISO 13485:2016 section 7?? ISO 13485:2016 - Medical Device Quality Management Systems 11
L HA vs risk analysis ISO 14971 - Medical Device Risk Management 2
Q Risk Controls in PFMEA ISO 14971 - Medical Device Risk Management 12
D What do you think of Chat GPTs answer to this Risk Acceptability question? ISO 14971 - Medical Device Risk Management 5
Richard Regalado Two risk assessments for ISMS IEC 27001 - Information Security Management Systems (ISMS) 0
M Risk-based approach to Test Method Validation for Design Verification? US Medical Device Regulations 5
N Effective use of a Risk Register - Bumper sticker or Mission Control ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
Q Risk Management ISO 14971 - Probability of Occurrence ISO 14971 - Medical Device Risk Management 8
Z Risk Management SOP ISO 14971 ISO 14971 - Medical Device Risk Management 1
M Risk Management Plan ISO 14971 - Medical Device Risk Management 13
J Risk, contingency, and MOC. General Auditing Discussions 1
K Help with ISO 14971: Benefit-Risk Analysis ISO 14971 - Medical Device Risk Management 3
T AS9100D Risk-Based Internal Audit Schedule AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
thisby_ Installation Related Issues and Risk Management ISO 14971 - Medical Device Risk Management 5
W Reconciling FMEA RPN ratings with Risk Acceptability ISO 14971 - Medical Device Risk Management 22
D How to address the content deviation of 'cannot apply criteria of risk acceptability prior to...' ISO 14971 - Medical Device Risk Management 1
Doninina Risk management file according MDR or ISO 14971:P2019 ? EU Medical Device Regulations 2
T Risk based CA AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
T IVD Risk - destruction of patient samples - Harm to property? ISO 14971 - Medical Device Risk Management 5
E Do anyone have document of automotive production risk and control of risk? Lean in Manufacturing and Service Industries 1
R Using RPN to Confirm Risk Reduced to an Acceptable Level Risk Management Principles and Generic Guidelines 12
T IVD Device Software - Risk Classification IEC 62304 - Medical Device Software Life Cycle Processes 16
G Help:Risk Management - Accessories US Food and Drug Administration (FDA) 1
N Writing Risk Management procedure for small manufacturing and we don't know where to start. Manufacturing and Related Processes 9

Similar threads

Top Bottom