Hello,

a quick question :)

Does it make sense to have the same Risk Evaluation Matrix for completely different products?

Shouldn?t different products ( such as Apps, or heavy machines) have different tresholds of acceptable/unacceptable risks?

What is the best practice? Is it better to your experience to have 1 Evaluation Matrix for all, or to define 1 evaluation Matrix for each single product/device?

Thank you very much!
 
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#2
The evaluation matrix you mentioned is the graphical interpretation of the risk acceptability criteria.

The criteria should be based on the risk management policy, and should be created for each device.

So yes, you should have different ones for different devices. The only rationale for having the same for different devices is if the devices have generally the same risks, for example, if all are lung ventilators. If you manufacture a lung ventilator and an implant, or a lung ventilator and ultra-sound machine, there?s no way the risk acceptability criteria would be the same.

However, most of the time people use only one. This is common practice, not best practice, thought.
 
Last edited:

alimary15

Involved In Discussions
#3
The evaluation matrix you mentioned is the graphical interpretation of the risk acceptability criteria.

The criteria should be based on the risk management policy, and should be created for each device.

So yes, you should have different ones for different devices. The only rationale for having the same for different devices is if the devices have generally the same risks, for example, if all are lung ventilators. If you manufacture a lung ventilator and an implant, or a lung ventilator and ultra-sound machine, there?s no way the risk acceptability criteria would be the same.

However, most of the time people use only one. This is common practice, not best practice, thought.
Thanks for the feedback !

In case the devices are totally different I would go for the option to have different acceptability criterias. However, if having a general and "good-for-all" policy, would it then make any sense to discuss possible residual risks in a risk/benefit analysis?

By having an acceptability criteria that is unique for all, wouldn?t there be a risk of classifyng as acceptable risks that might not be acceptable? I am concerned about and if it is possible to lose any kind of information within the RM process by doing so.

What is your opinion?

Thanks
 

Marcelo

Inactive Registered Visitor
#4
However, if having a general and "good-for-all" policy, would it then make any sense to discuss possible residual risks in a risk/benefit analysis?
The risk management policy is the high-level policy on how the enterprise deal with risk. Policies are direction, but really not directly actionable. That why you need to define the criteria for each device based on the policy. Please see an example of policy discussion in this thread - http://elsmar.com/Forums/showthread.php?t=52843

By having an acceptability criteria that is unique for all, wouldn?t there be a risk of classifyng as acceptable risks that might not be acceptable? I am concerned about and if it is possible to lose any kind of information within the RM process by doing so.
Yes, that why is makes no sense to have a risk acceptability criteria for all devices, as I mentioned. This problem might even happen if the devices are similar but have modifications that can impact the risks.
 

alimary15

Involved In Discussions
#5
The risk management policy is the high-level policy on how the enterprise deal with risk. Policies are direction, but really not directly actionable. That why you need to define the criteria for each device based on the policy. Please see an example of policy discussion in this threadhttp://elsmar.com/Forums/showthread.php?t=52843
Sorry for the bad use of words- I was always speaking about the risk criteria of acceptability!

Thanks for the help!
 

alimary15

Involved In Discussions
#6
This problem might even happen if the devices are similar but have modifications that can impact the risks.
In this case for example, when upgrading the same device, it is practice to start from the previous RA and then move forward to identify new possibile risks that are coming from the new modification.

My only concern is using the same acceptability criteria for software, hardware, and different kind of devices. I think it would make things way more general.

I will try to discuss this further within my team and try to change things.

Thanks for the confirmation :thanx:
 
Last edited by a moderator:

Marcelo

Inactive Registered Visitor
#7
Ah, ok, well, people generally use those concepts interchangeably, but they are very different, which causes confusion in the discussion and understanding.

So, if your question is

However, if having a general and "good-for-all" risk acceptability criteria, would it then make any sense to discuss possible residual risks in a risk/benefit analysis?

Yes, it would, because the residual risk are related to risk control measure, and even if similar devices would have similar risks, it does not mean that they would have the same risk control measures (in fact, newer devices in a device family usually have different risk controls measures to some risks because that?s how technology evolution works.
 

Marcelo

Inactive Registered Visitor
#8
In this case for example, when upgrading the same device, it is practice to start from the previous RA and then move forward to identify new possibile risks that are coming from the new modification.
Yes, and this is acceptable by ISO 14971 IF you perform and evaluation and concludes that you can use the proviso RA.

My only concern is using the same acceptability criteria for software, hardware, and different kind of devices. I think it would make things way more general.
Exactly. They probably need separate ones, based on the same high level risk management policy.
 

alimary15

Involved In Discussions
#9
Ok great thank you!

Then I might have a more complex question:

Let?s consider a scenario where we have different modules that could be or could be not present on the field and that are sold as a complete system to the costumer.

Then what is it the best practice? Make RA for each single module? Shall RA also be made to the complete and overall system to include possible risks arising from interactions between each single module?

What?s the best practice in this case?

Thanks
 

Marcelo

Inactive Registered Visitor
#10
Then I might have a more complex question:

Let?s consider a scenario where we have different modules that could be or could be not present on the field and that are sold as a complete system to the costumer.

Then what is it the best practice? Make RA for each single module? Shall RA also be made to the complete and overall system to include possible risks arising from interactions between each single module?

What?s the best practice in this case?

Thanks
Risk management (as defined in ISO 14971 and as required by most regulations) is applicable to the medical device as a whole.

So yes, you would need to focus on the complete device, and as you mentioned, part of the focus is on interactions between each single module - in fact, think of it as interactions in the different configurations.
 
Thread starter Similar threads Forum Replies Date
T Biological Evaluation (10993) & Risk Management ISO 14971 - Medical Device Risk Management 9
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
U Changes to Internal Processes and Risk Evaluation - Mitigations Risk Management Principles and Generic Guidelines 10
M Informational Work in progress at the FDA for biological evaluation – Color Hazard and RISk calculator (CHRIS) Medical Device and FDA Regulations and Standards News 0
Q Evidence of precautions (clinical evaluation report, risk management report) EU Medical Device Regulations 6
qualprod Qualitative or Quantitative? Risk method evaluation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Rincewind Clinical Evaluation of very low risk medical devices EU Medical Device Regulations 3
J Simple supplier evaluation qualification process form that includes Risk Assessment Document Control Systems, Procedures, Forms and Templates 2
Sam Lazzara ISO 14971 Clause 7 - Evaluation of Overall Residual Risk Acceptability ISO 14971 - Medical Device Risk Management 3
J FMEA Evaluation of "User Risk" - Wording FMEA 3rd party advisors ISO 14971 - Medical Device Risk Management 3
S Risk Analysis and Risk Evaluation - Biomedical company ISO 14971 - Medical Device Risk Management 6
B Chrysler requirement of Initial Risk Evaluation IATF 16949 - Automotive Quality Systems Standard 2
E Supply chain main policies ,scope, risk assessments & relavant KPI Supply Chain Security Management Systems 2
D Use Error Risk Controls and Control Verification ISO 14971 - Medical Device Risk Management 5
J Risk Assessment of Lithium Ion Batteries FMEA and Control Plans 3
Melissa Risk Management Process, How far do I need to go? ISO 14971 - Medical Device Risk Management 10
D Does Risk Management apply to re-labeler (MDR) EU Medical Device Regulations 1
H Risk Management Plan in agile process ISO 14971 - Medical Device Risk Management 11
H Risk Analysis and Probability of Occurrence ISO 14971 - Medical Device Risk Management 3
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
P Benefit risk analysis on pFMEA ISO 14971 - Medical Device Risk Management 9
B AS9102 - 3D printing a special tool required for assembly (counterfeit risk?) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
K Defining risk control measures IEC 62304 - Medical Device Software Life Cycle Processes 14
U Supply risk management Manufacturing and Related Processes 4
D Cybersecurity and Risk Management: Loss of confidentiality IEC 62304 - Medical Device Software Life Cycle Processes 5
Q FMEA and Risk assessment in Microsoft Access FMEA and Control Plans 6
I Realization processes input into overall risk ISO 14971 - Medical Device Risk Management 2
M Need Help With Information Security Asset Risk Register IEC 27001 - Information Security Management Systems (ISMS) 2
thisby_ Post Market/Production Risk Assessment ISO 14971 - Medical Device Risk Management 0
S Risk Management Review ISO 14971 - Medical Device Risk Management 4
D Low risk IVD study in the UK, do I need MHRA approval? UK Medical Device Regulations 1
S Risk Management and other Files ISO 14971 - Medical Device Risk Management 8
N ISO 27001 for Jumb Burger - Risk Assessment sheet IEC 27001 - Information Security Management Systems (ISMS) 11
C Risk Assessment Tools ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
qualprod Examples to mitigate risk from Covid ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
G Risk of stopping your customer's line IATF 16949 - Automotive Quality Systems Standard 4
C Risk Matrix vs FMEAs ISO 14971 - Medical Device Risk Management 12
S IVD risk class II devices for Brazil and MDSAP Other Medical Device Regulations World-Wide 0
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 11
M ISO14971:2019 - Verification of implementation and effectiveness of risk control ISO 14971 - Medical Device Risk Management 3
Aymaneh Medical Device Cybersecurity Risk Management IEC 27001 - Information Security Management Systems (ISMS) 2
S Traceability of requirements to design and risk Design and Development of Products and Processes 3
R Risk control measures as per ISO 14971 ISO 14971 - Medical Device Risk Management 6
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
R The term "Benefit Risk Ratio" in EU MDR, do I need to present benefit risk analysis as a RATIO Risk Management Principles and Generic Guidelines 4
_robinsingh Security Risk Assessment Tool IEC 27001 - Information Security Management Systems (ISMS) 0
A 21 CFR 820 - Risk Management - Looking for some guidance US Food and Drug Administration (FDA) 3
bryan willemot Contract Review and risk managment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Risk Analysis using Monte Carlo Simulation instead of Scoring and Heat Map Risk Management Principles and Generic Guidelines 2
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8

Similar threads

Top Bottom