We have been recently certified to ISO 13485 and 9001, and in our most recent external audit our auditor found that our risk management procedure did not meet the requirements of 14971, and wrote this up as a Level 1 nonconformance. I was not at the audit, and our Director of Operations was not able to determine (from the auditor) why he felt we needed to comply when 14791 is a note in the standard and not (apparently) a requirement of 13485. I have been tasked with revising our current risk management procedure to comply with 14971. I was wondering if anyone had run into this situation, and if there are any suggestions (examples?) of a risk management procedure which would conform.
Our company is a contract manufacturing concern, and we have been asked to build a non-invasive device by a design firm which has designed the device. Thanks for your thoughts!
arios
I don't understand either why the comment sounds like you have to implement ISO 14971 unless it has been declared as required in other parts of your QMS or in a Technical documentation. What is true is that you have to implement risk management in yous system and have records.
There is a guidance document of the GHTF which can serve you: http :// www .ghtf.org/documents/sg3/sg3n15r82005.pdf - OBSOLETE BROKEN 404 LINK(s) UNLINKED - PLEASE HELP - REPORT POSTS WITH BROKEN LINKS
On the other hand, ISO 14971 is a very complete standard, supported with very good information and annexes that help you understand how to implement a good Risk Management process. My advice for you is to acquire and get familiar with this standard. Risk management is not only a requirement, but also makes good business sense, and overall it help us think more on the patient on a proactive manner.
MRWardell
Arios - yes, 14971 is a good standard and I have been reviewing it in order to respond to the nonconformance. Our risk management procedure under 13485 is a little weak, but I guess my concern was that the auditor required our procedure to conform to 14971, while I did not see how he could do that when 13485 does not appear, at least to me, to make that requirement. Thanks for your input and the link.:thanx:
DrM2u
I suggest that you dispute the finding with the registrar. Note 3 under 7.1 says to look at ISO 14971 for guidance. Guidance is needed to compensate for lack of knowledge. There is no requirement in ISO 13485 that the organization has to comply with the 'guidance' of ISO 14971 or with auditor's interpretations of the standard.
To shine some light on WHY the finding was issued ... and this is just my educated guess ... The auditor gets to do extra work (and doesnt get paid for) if the review board rejects the audit package or asks for additional information. The chances are that the review board uses ISO 14671 for 'guidance' and hystorically they are a lot pickier than ISO 9001 review boards for example. Therefore the auditor probably figured that if your organization complies with ISO 14671 then he and the organization avoids a potential hurdle in the registration process. A good auditor should have explained this potential scenario and maybe documented an opportunity for improvement or area of concern, but not a finding.
Any additional feedback from other auditors or auditees?!?
arios
I agree with DrM2U. You could further inquiry about the auditor's finding,
and if not fair challenge the auditor's conclusion. If you wish to appeal the finding also check ASAP the timing with your registrar to do that process. Like mentioned earlier you should also consider that the auditor may have raised the finding based on other technical reasons, which would be important to find out if there is any. It is sad he apparently did not mention the reasons.
John Martinez
Ok, please define "external auditor". If a Certification Body (registrar) then challenge the finding.
If Notified Body (CE Mark or CMDCS) then they may be accurate.
If a regulator or customer, then you may have to comply to continue business.
I'm not a Medical Device expert. Has the US CFR changed to require this specific ISO standard? If so, that would be a switch, since most regulators seem to be allergic to ISO standards, even though their CFR's reflect the Standard's requirements.
MRWardell
John - the external auditor is our ISO 13485 registrar. I think the time to challenge the finding is past, and I don't think our Director of Operations wants to take that course if we can address the finding and satisfy the auditor. I am working on a revised Risk Management procedure which I think addresses the requirements of 14971, but I guess we'll see. There have not been any changes to the standard, just this auditors view of what we need to do to comply. Thanks for your thoughts.
John Martinez
By accreditation standards, there are formal contestment processes in place. Your auditor is REQUIRED to advise you of this during the closing meeting.
If there is a non-conformance and you have not answered it or closed it yet, then the l time to contest is not too late.
Certainally, you can make a call to your registrar's technical department and inquire as to when this became a requirement. Perhaps is it an accreditation body's requirement that the registrar is pushing down.
Bottom line is, if you do not have all of the information; how do you exptect your response to this NC to be effective?
I'd make the call. If not as a contestment, then as an iquiry as to where this requirement came from so that you can address it properly.
I'm not stating that this document is not useful. It is. I'm stating that, obviously you are having a diffucult time with this, so utilize your registrar's technical department's insight.
DrM2u
To add to John's comments, your job is not to satisfy the auditor but to meet the requirements of the standard in a way that is beneficial to the organization. You can always ask for another auditor if you believe that your current auditor is not professional or has difficulties understanding your system. Keep in mind that you are the customer for the registrar, not the other way around!
jkuil
Documented requirements for risk management are obligatory per ISO 13485:2003 7.1.
ISO 14971:2007 is the EU harmonized / US, Canada, Japan... recognized standard to meet the essential requirements for risk management. This means that when you claim conformance to this standard authorities will accept this as conforming to the essential requirements for safety and efficacy as specified in legislation (e.g. MDD annex I).
However, standards are voluntary and you can use other standards or your own business standards to meet the requirements. However, that will not result in easy approval by the notified body and/or authorities as you experienced. They should verify that your QMS complies with the requirements of the recognized standard (as this is the expected minimum) even when you do not claim conformance. It is easier for them to require conformance to ISO 14971, but inappropiate.
Still, conforming to ISO 14971 is a sensible thing to do, as it will facilitate marketing approvals. So your operational director took a right decisiion.
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
R Identify Medical Device characterstics as Annex C of ISO 14971 Risk Management ISO 14971 - Medical Device Risk Management 5
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
S ISO 14971 Risk Management - Questions for Hazard identification ISO 14971 - Medical Device Risk Management 2
D Risk management according to ISO 14971 - When to document risk controls? ISO 14971 - Medical Device Risk Management 10
M Informational ISO TC 210 IEC SC 62A JWG 1 Medical device risk management – São Paulo meeting 2019 Medical Device and FDA Regulations and Standards News 6
M Medical Device News ISO TC 210 IEC SC 62A JWG 1 Medical device risk management – São Paulo meeting 2019 Medical Device and FDA Regulations and Standards News 0
D Where does FMEA fit in your ISO 14971 Risk Management process? ISO 14971 - Medical Device Risk Management 13
M Informational ISO TC 210 JWG 1 meeting in São Paulo – Revision of ISO 14971 and ISO TR 24971 – Medical Device Risk Management Medical Device and FDA Regulations and Standards News 0
S ISO 13485:2016 - How I can integrate a risk management approach in our SOPs ISO 13485:2016 - Medical Device Quality Management Systems 1
A Including all Processes in Risk Management - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 8
T Using Risk Management in ISO 10993 - Medical Device Accessory 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Q Risk Management - Additional Process in ISO 9001:2015? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
alonFAI How to define a Risk Based Approach for Supplier Management per ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
L Risk Management in an IVD, ISO 13485 certified company ISO 14971 - Medical Device Risk Management 2
S Informational Risk Management Implementation for ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 60
A Informational Risk Management (and Risk Based Thinking) in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 56
A Is Risk Management Process compliant to ISO 14971 in absence of Hazardous Situations? ISO 14971 - Medical Device Risk Management 5
Marc Are you looking for ISO 31000 - Risk Management Principles and Generic Guidelines? ISO 14971 - Medical Device Risk Management 1
Marc Are you looking for ISO 14971 - Medical Device Risk Management? Risk Management Principles and Generic Guidelines 1
C ISO 13485 - Documented Requirements for Risk Management ISO 13485:2016 - Medical Device Quality Management Systems 6
M Risk Management (ISO 14971:2007) Internal Audit Checklist ISO 14971 - Medical Device Risk Management 7
somashekar What are the ISO 13485 documented requirements for Risk Management? ISO 13485:2016 - Medical Device Quality Management Systems 13
Y ISO 31004 - Risk Management Implementation Guidance Risk Management Principles and Generic Guidelines 2
N ISO 14971:2007 vs. 2009 - Which Risk Management Standard is still accepted in the EU Other ISO and International Standards and European Regulations 2
E ISO 14971:2009 Risk Management Requirements CE Marking (Conformité Européene) / CB Scheme 2
G ISO Registrars/CBs citing for use of "Detection" in Risk Management ISO 14971 - Medical Device Risk Management 27
M ISO 14971 Medical Device Risk Management FAQ ISO 14971 - Medical Device Risk Management 38
M Risk Management Plan Template - ISO 14971:2007 Compliant ISO 14971 - Medical Device Risk Management 13
C Section 7.1 of ISO 13485 - Risk Management throughout Product Realization ISO 13485:2016 - Medical Device Quality Management Systems 7
D ISO 17025 and Calibration Laboratory Risk Management Requirements ISO 17025 related Discussions 1
C Scope of Risk Management in ISO13485 vs. ISO 14971/EU MDD ISO 14971 - Medical Device Risk Management 2
C ISO 14971 Clause 9 Requirements - Post-Production Monitoring and Risk Management ISO 14971 - Medical Device Risk Management 7
A Where can I buy EN ISO 14971:2009 (Medical Device Risk Management)? ISO 14971 - Medical Device Risk Management 11
Q ISO 14971 Class II Medical Devices - Product Realization & Risk Management ISO 14971 - Medical Device Risk Management 5
K ISO 62304 Software Risk Management and Medical Device Class IEC 62304 - Medical Device Software Life Cycle Processes 5
K Risk Management (ISO 31000) and AS9100C - Gap Analysis Template Needed Risk Management Principles and Generic Guidelines 2
K ISO 31000 - Implementing Risk Management in the Construction Industry? Risk Management Principles and Generic Guidelines 5
B ISO 9001:2008 and Risk Management ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
B Application of Risk Management - ISO 14971 for a Tooling Manufacturer ISO 14971 - Medical Device Risk Management 18
A ISO 9001 Project Management and Risk Analysis Requirements - Construction ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
R ISO 10993-1:2009 incorporating Risk Management published by ISO ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 14971:2007 Risk Management - Class I A Sterile Medical Device ISO 14971 - Medical Device Risk Management 4
M Implementing Risk Management for Contract Manufacturers - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 12
K What is 'output of risk management' in ISO 13485, Cl 7.3.2(e).? ISO 13485:2016 - Medical Device Quality Management Systems 7
M ISO 14971 Risk Management Qualification of Personnel ISO 14971 - Medical Device Risk Management 5
Q ISO 14971 for Contract Manufacturers - Extensive risk management ISO 14971 - Medical Device Risk Management 16
V ISO 14971:2007 Application of Risk Management to Medical Devices ISO 14971 - Medical Device Risk Management 9
M ISO 14971 Risk Management questions and comments ISO 14971 - Medical Device Risk Management 38
V ISO 14971 Risk Management training courses in Montreal or Ottawa ISO 14971 - Medical Device Risk Management 1