Where in the MDR is there a requirement for a risk management report? I've never seen any anywhere.
Below listed are risk management requirements as per MDR Annex I - SECTION 3, 4 & 5
3. Manufacturers shall establish, implement, document and maintain a risk management system. Risk management shall be understood as a continuous iterative process throughout the entire lifecycle of a device, requiring regular systematic updating. In carrying out risk management manufacturers shall: (a) establish and document a risk management plan for each device; (b) identify and analyse the known and foreseeable hazards associated with each device; (c) estimate and evaluate the risks associated with, and occurring during, the intended use and during reasonably foreseeable misuse; (d) eliminate or control the risks referred to in point (c) in accordance with the requirements of Section 4; (e) evaluate the impact of information from the production phase and, in particular, from the post-market surveillance system, on hazards and the frequency of occurrence thereof, on estimates of their associated risks, as well as on the overall risk, benefit-risk ratio and risk acceptability; and (f) based on the evaluation of the impact of the information referred to in point (e), if necessary amend control measures in line with the requirements of Section 4.
4. Risk control measures adopted by manufacturers for the design and manufacture of the devices shall conform to safety principles, taking account of the generally acknowledged state of the art. To reduce risks, Manufacturers shall manage risks so that the residual risk associated with each hazard as well as the overall residual risk is judged acceptable. In selecting the most appropriate solutions, manufacturers shall, in the following order of priority: (a) eliminate or reduce risks as far as possible through safe design and manufacture; (b) where appropriate, take adequate protection measures, including alarms if necessary, in relation to risks that cannot be eliminated; and (c) provide information for safety (warnings/precautions/contra-indications) and, where appropriate, training to users. Manufacturers shall inform users of any residual risks.
5. In eliminating or reducing risks related to use error, the manufacturer shall: (a) reduce as far as possible the risks related to the ergonomic features of the device and the environment in which the device is intended to be used (design for patient safety), and (b) give consideration to the technical knowledge, experience, education, training and use environment, where applicable, and the medical and physical conditions of intended users (design for lay, professional, disabled or other users).