Routine Testing for Class I Medical Device

S

sara_jg

#1
Hi,

For one of our product (class I device) we have already done type testing as per IEC 60601-1.

For doing routine testing in all our products before dispatching whether we need to do high voltage testing(dielectric test) and insulation resistance (IR test) testing as per 60601-1 std. or shall we define our own routine tests keeping 60601-1 and 62353 as reference??

please help.

Thank you in advance
saran
 
Elsmar Forum Sponsor
#2
The tests and requirements set out in 60601-1 are, as you identify, Type Tests.

It is generally understood that some of these tests can be damaging to the unit under test. For instance, the impact test tends to dent panels and the marking durability tests can scuff labels. Neither necessarily constitutes test failure - but leave the unit in an unsaleable condition.

This accumulation of damage is why the sequence of tests (see Annex B) is defined. Your device should still be functional and safe after all he tests have been completed - but it may look a little worse for wear.

The dielectric strength test places significant stress upon electrical insulation. It may even permanently damage the insulation. And that damage may not be visible. So, it is not possible to guarantee that a test unit has undamaged insulation - so should never enter the market.

So what tests should be used on production?

Clause 5.1 describes the use of Type Tests. The 'Note 2' to that clause identifies IEC 62354 Annex K as guidance for 'Production Tests' and IEC62353 as guidance for 'Recurrent Tests'. Both have alternate methods for both the dielectric tests and various leakage current tests.

The guidance rationale for Clause 5.1 is useful:
Production tests need not be identical with TYPE TESTS, but can be adapted to manufacturing conditions and possibly invoking less RISK for the quality of the insulation or other characteristics...
One option for the dielectric tests that is quite commonly used, for instance by manufacturer's of transformers, is to perform the full voltage, full 1 minute, test on a sample to 'Type Test' it but then, for production, test at full voltage - but only for 5 seconds. This is, basically, the test we perform for production.

One thing to highlight - the dielectric strength tests only apply to solid insulations. If you are using 'creepage and clearance' insulations then you use a ruler as your test equipment - not the dielectric tester!

Also, when it comes to production dielectric tests, it is often a mistake to test from the 'back to the front'. By that I mean testing from the mains input to the front panel connectors. What happens then is that the test voltage stress is applied across more than one insulation and will be 'shared out' unevenly and unpredictably.
 

phase90

Starting to get Involved
#3
What are the requirements for dielectric tests during production? I have had conflicting information. As an example, the test house that certified a product for 60601 requires us to test with full voltage (1500 VAC) for 60 seconds (3000 VAC for applied parts to mains). The notified body that audits us to keep our CE certification tells us we are testing it too long and could damage the insulation. They recommend 5 seconds. :confused:
 
#4
The test house is only used to doing Type Tests - hence their suggestion of 60s dielectric test.

For the production test 5 seconds is preferable. But, even then, make sure that you are only testing solid insulation and you are not stressing parallel insulations (where the test voltage would be unevenly divided).
 

phase90

Starting to get Involved
#5
Thanks Pads.

What I have been told now is that it can be tested for a shorter period of time, but at a higher voltage. (1000 + 2x line rating) * 1.2 for 1 second. (1776V, since they are using 230 VAC in the calculation, rather than 250V)
 
R

Randy956

#6
Hello,

If testing at hipot for 5 seconds, what IEC standard would support this time frame?

I've seen hipot being applied to production line units for 60 seconds and I think that's too long. I need some sort of "official" reference I can point to before I can suggest shortening that dwell time. Any suggestions on what document I can use as reference?

Thanks!

Randy
 
#7
Hi Randy and welcome to the forums.

The 'Type Tests' for medical electrical equipment are what is in IEC 60601-1 (plus collateral and particular parts).

For 'Production Tests' the technical report IEC 62354 provides guidance.

For 'Recurrent Tests' (perhaps those done by a hospital's equipment management service each year) the technical report IEC 62353 provides guidance.
 
S

sara_jg

#8
Hi

it would be of great help if any body could share a copy of IEC 62354 standard.

Thank you
sara
 

Marcelo

Inactive Registered Visitor
#9
<snip> it would be of great help if any body could share a copy of IEC 62354 standard.
Any deliverable from IEC (IEC TR 62354 is in fact not a standard, but a technical report) is copyrighted and must be bought. Please refrain from asking for pirated copies of copyrighted material in this forum.
 
Thread starter Similar threads Forum Replies Date
M Routine testing of medical electrical systems - What specific electrical safety tests should be performed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
T Nonconforming product at customer detected by a routine inspection by field service Nonconformance and Corrective Action 9
N Referencing of a standard's year in routine regulatory compliance documents Other Medical Device Related Standards 3
M Medical Device News FDA - Eliminating Routine FDA news 14-09-18 - Re-Review of Third Party 510(k) Reviews Other US Medical Device Regulations 0
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
C Dose Mapping - How to relate back to Routine Monitoring Position ISO 13485:2016 - Medical Device Quality Management Systems 3
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
J FDA Routine Inspection Pre-Anouncement - What will be checked? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
M Identifying Routine and Non-Routine Activities: OSHA OHSAS 18001 Occupational Health & Safety Management Standards 6
S How you define 'Normal & Abnormal' Conditions and 'Routine & Non-routine' Activities ISO 14001:2015 Specific Discussions 8
J Routine test for MD finished product IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
RoxaneB Environmental Objectives - Do they need to be linked into with Routine Indicators? Miscellaneous Environmental Standards and EMS Related Discussions 6
RoxaneB Search for another name for "Standardization Process" or "Routine Management" Quality Tools, Improvement and Analysis 4
D Line clearance - Routine components in the prep room for universal use ISO 13485:2016 - Medical Device Quality Management Systems 8
R Inspection and Detection - Kit to production - Need an inspection routine Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 5
B NIOSH Approval for Surgical N95 Respirators - Required testing US Food and Drug Administration (FDA) 2
M ECG lead leakage currents - How to specify ECG leads during electrical safety testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
D CFR Title 14: Aeronautics and Space PART 120—DRUG AND ALCOHOL TESTING PROGRAM Federal Aviation Administration (FAA) Standards and Requirements 3
lanley liao Purchase Acceptance Criteria - Tensile testing Oil and Gas Industry Standards and Regulations 2
M Device mounted at IV pole - what about mechanical stability testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Cochlear Implant Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Class I (exempt) testing requirements Other Medical Device Related Standards 0
JoCam Electrical Testing for Japan, PSE or CB Scheme Other Medical Device Regulations World-Wide 0
M Who are the go to companies for non-destructive hardness testing? General Measurement Device and Calibration Topics 3
M Determining if an Insulin Pen Testing Machine is a Medical Device? EU Medical Device Regulations 4
P Testing cloud-based backups IT (Information Technology) Service Management 7
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
N Chemical Testing on Medical Devices - Solutions in a container closure system (bag) EU Medical Device Regulations 1
M Comparing data from destructive testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T Flammability testing Reliability Analysis - Predictions, Testing and Standards 0
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
K When is Bioburden Testing Required? Other Medical Device Related Standards 4
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
A ANSI/AAMI versions of 60601-1-2 and related testing requirements Other Medical Device Related Standards 3
C Surgical mask stability testing (CE mark) EU Medical Device Regulations 2
Beliz Biocompatibility Testing for Laser Epilation Device EU Medical Device Regulations 2
C One Time Service Supplier - Temperature and Humidity Testing Service ISO 13485:2016 - Medical Device Quality Management Systems 5
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
P Flammability Testing of Aircraft Interior Materials Federal Aviation Administration (FAA) Standards and Requirements 0
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 8
E ASTM F2118 - Fatigue testing of bone cement - Changes between the 2003 and the 2014? Other Medical Device Related Standards 1
K Biocompatibility Testing - Multile products of different sizes and shapes US Food and Drug Administration (FDA) 2
S Requirement to Conduct New Shelf-life Testing? (re-do testing for design change) EU Medical Device Regulations 3
JoCam Mobile Patient Hoists and Electrical Testing Other Medical Device Related Standards 0
T Interlaboratory comparison or proficiency testing in destructive testing of welded joints ISO 17025 related Discussions 3

Similar threads

Top Bottom