Hi everybody. I'm new to this forum and would really appreciate any help with regards to the following. If we are using IEC 60601 as a basis for routine testing of medical electrical systems, what specific electrical safety tests should be performed?

From Clause 16 of 60601, I can find reference to these tests:

- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)

If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?

- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?

Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?

Many thanks for any help.

Michael
 
Last edited:
Elsmar Forum Sponsor

Ahang

Registered
#2
Hi everybody. I'm new to this forum and would really appreciate any help with regards to the following. If we are using IEC 60601 as a basis for routine testing of medical electrical systems, what specific electrical safety tests should be performed?

From Clause 16 of 60601, I can find reference to these tests:

- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)

If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?

- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?

Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?

Many thanks for any help.

Michael
Hello Michael,

- Answer to your first question:
As stated in sub-clause 16.1,
1- you should first specify that which part(s) of your me system is(are) installed in the patient environment and which is(are) outside of this environment.
2- For those parts which are in the patient environment whether they are medical devices or non-medical devices (e.g. a laptop connected to the medical device in order to extract data), all me equipment relevant requirements shall be applied according to 60601-1 (e.g. all parts of clause 8).
3- For those parts which are not in the patient environment, their related safety standards shall be applied.
As an example imagine that you connect an active medical device (which is connected to the patient) through a network to a monitor in the nurse station, in this case, the medical device is in the patient environment and shall be tested according all applicable requirements of IEC 60601-1. But the monitor is outside of the patient environment and shall be in compliance with ITE safety standards (e.g. IEC 60950)

Answer to your 2'nd question:
Refer to IEC 60601-1, 8.7.1 b) 3’rd dash, all leakage currents including touch current shall be measured in both normal condition and single fault condition.

Ahang Rabiee
 

swathi.sp

Starting to get Involved
#3
Dear Folks,

Further to Leakage current test as per IEC 60601-1 , Clause 8.7, we have our own designed human body impedence circuit "as illustrated below representead as "MD in 60601-1. Can anybody help me to understand how to analyse the frequency characteristics of the Measuring device.



1568107273602.png
 

Benjamin Weber

Trusted Information Resource
#5
Dear Folks,

Further to Leakage current test as per IEC 60601-1 , Clause 8.7, we have our own designed human body impedence circuit "as illustrated below representead as "MD in 60601-1. Can anybody help me to understand how to analyse the frequency characteristics of the Measuring device.



View attachment 26052
Well, actually frequency analysis comes for free here. Just look at the right figure, that is the frequency response of the low pass filter / measuring device in the left.

It is a simple first order low-pass with a cut-off frequency of 1061 Hz (-3db, defined by R1 and C1). The resistor R2 is the shunt resistor through which the leakage current is flowing.

More Info:
Low-pass filter - Wikipedia -> Electronic low-pass filter
 

swathi.sp

Starting to get Involved
#6
Well, actually frequency analysis comes for free here. Just look at the right figure, that is the frequency response of the low pass filter / measuring device in the left.

It is a simple first order low-pass with a cut-off frequency of 1061 Hz (-3db, defined by R1 and C1). The resistor R2 is the shunt resistor through which the leakage current is flowing.

More Info:
Low-pass filter - Wikipedia -> Electronic low-pass filter
Many thanks Benjamin. Your response is so helpful for me to understand further.
 
Thread starter Similar threads Forum Replies Date
S Routine Testing for Class I Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
T Nonconforming product at customer detected by a routine inspection by field service Nonconformance and Corrective Action 9
N Referencing of a standard's year in routine regulatory compliance documents Other Medical Device Related Standards 3
M Medical Device News FDA - Eliminating Routine FDA news 14-09-18 - Re-Review of Third Party 510(k) Reviews Other US Medical Device Regulations 0
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
C Dose Mapping - How to relate back to Routine Monitoring Position ISO 13485:2016 - Medical Device Quality Management Systems 3
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
J FDA Routine Inspection Pre-Anouncement - What will be checked? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
M Identifying Routine and Non-Routine Activities: OSHA OHSAS 18001 Occupational Health & Safety Management Standards 6
S How you define 'Normal & Abnormal' Conditions and 'Routine & Non-routine' Activities ISO 14001:2015 Specific Discussions 8
J Routine test for MD finished product IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
RoxaneB Environmental Objectives - Do they need to be linked into with Routine Indicators? Miscellaneous Environmental Standards and EMS Related Discussions 6
RoxaneB Search for another name for "Standardization Process" or "Routine Management" Quality Tools, Improvement and Analysis 4
D Line clearance - Routine components in the prep room for universal use ISO 13485:2016 - Medical Device Quality Management Systems 8
R Inspection and Detection - Kit to production - Need an inspection routine Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
K When is Bioburden Testing Required? Other Medical Device Related Standards 3
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
A ANSI/AAMI versions of 60601-1-2 and related testing requirements Other Medical Device Related Standards 0
C Surgical mask stability testing (CE mark) EU Medical Device Regulations 1
Beliz Biocompatibility Testing for Laser Epilation Device EU Medical Device Regulations 2
C One Time Service Supplier - Temperature and Humidity Testing Service ISO 13485:2016 - Medical Device Quality Management Systems 5
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
P Flammability Testing of Aircraft Interior Materials Federal Aviation Administration (FAA) Standards and Requirements 0
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 3
E ASTM F2118 - Fatigue testing of bone cement - Changes between the 2003 and the 2014? Other Medical Device Related Standards 1
K Biocompatibility Testing - Multile products of different sizes and shapes US Food and Drug Administration (FDA) 2
S Requirement to Conduct New Shelf-life Testing? (re-do testing for design change) EU Medical Device Regulations 3
JoCam Mobile Patient Hoists and Electrical Testing Other Medical Device Related Standards 0
T Interlaboratory comparison or proficiency testing in destructive testing of welded joints ISO 17025 related Discussions 2
B ASTM E18-2020 - Rockwell testing standard changes? General Measurement Device and Calibration Topics 2
U Medical Device Design finalization testing ISO 13485:2016 - Medical Device Quality Management Systems 2
Jane's Like-for-like critical raw material change qualification - type of testing/ number of lots required ISO 13485:2016 - Medical Device Quality Management Systems 4
J Conflict of Interest Registrar/Notified Body/Testing House Quality Manager and Management Related Issues 4
M Inter-operator Variability Testing - Requirements for EU Medical Device Regulations 5
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
M Production approval testing - Alternative ideas for Validation Reliability Analysis - Predictions, Testing and Standards 4
JoCam MDL in Canada without Canadian Electrical Testing Canada Medical Device Regulations 0
T HF testing / Summative evaluation for MDDS class I necessary? Human Factors and Ergonomics in Engineering 2
Kuldeep Singh Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Workplace Drug Testing in 2020 Misc. Quality Assurance and Business Systems Related Topics 9
W ASTM F1929 dye penetration test - Validation for in-house testing ISO 13485:2016 - Medical Device Quality Management Systems 13
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 11
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 3
T Qualification testing of Lead acid batteries Reliability Analysis - Predictions, Testing and Standards 0
D Sterility and BioBurden Testing for Plastics ISO 13485:2016 - Medical Device Quality Management Systems 5
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Correct way to certify hydrostatic testing when it is not 100% (and Sample Size) Various Other Specifications, Standards, and related Requirements 6

Similar threads

Top Bottom