Routine testing of medical electrical systems - What specific electrical safety tests should be performed?

Michael Ayers

Starting to get Involved
#1
Hi everybody. I'm new to this forum and would really appreciate any help with regards to the following. If we are using IEC 60601 as a basis for routine testing of medical electrical systems, what specific electrical safety tests should be performed?

From Clause 16 of 60601, I can find reference to these tests:

- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)

If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?

- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?

Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?

Many thanks for any help.

Michael
 
Last edited:
Elsmar Forum Sponsor

Ahang

Registered
#2
Hi everybody. I'm new to this forum and would really appreciate any help with regards to the following. If we are using IEC 60601 as a basis for routine testing of medical electrical systems, what specific electrical safety tests should be performed?

From Clause 16 of 60601, I can find reference to these tests:

- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)

If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?

- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?

Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?

Many thanks for any help.

Michael
Hello Michael,

- Answer to your first question:
As stated in sub-clause 16.1,
1- you should first specify that which part(s) of your me system is(are) installed in the patient environment and which is(are) outside of this environment.
2- For those parts which are in the patient environment whether they are medical devices or non-medical devices (e.g. a laptop connected to the medical device in order to extract data), all me equipment relevant requirements shall be applied according to 60601-1 (e.g. all parts of clause 8).
3- For those parts which are not in the patient environment, their related safety standards shall be applied.
As an example imagine that you connect an active medical device (which is connected to the patient) through a network to a monitor in the nurse station, in this case, the medical device is in the patient environment and shall be tested according all applicable requirements of IEC 60601-1. But the monitor is outside of the patient environment and shall be in compliance with ITE safety standards (e.g. IEC 60950)

Answer to your 2'nd question:
Refer to IEC 60601-1, 8.7.1 b) 3’rd dash, all leakage currents including touch current shall be measured in both normal condition and single fault condition.

Ahang Rabiee
 

swathi.sp

Involved In Discussions
#3
Dear Folks,

Further to Leakage current test as per IEC 60601-1 , Clause 8.7, we have our own designed human body impedence circuit "as illustrated below representead as "MD in 60601-1. Can anybody help me to understand how to analyse the frequency characteristics of the Measuring device.



1568107273602.png
 

Benjamin Weber

Trusted Information Resource
#5
Dear Folks,

Further to Leakage current test as per IEC 60601-1 , Clause 8.7, we have our own designed human body impedence circuit "as illustrated below representead as "MD in 60601-1. Can anybody help me to understand how to analyse the frequency characteristics of the Measuring device.



View attachment 26052
Well, actually frequency analysis comes for free here. Just look at the right figure, that is the frequency response of the low pass filter / measuring device in the left.

It is a simple first order low-pass with a cut-off frequency of 1061 Hz (-3db, defined by R1 and C1). The resistor R2 is the shunt resistor through which the leakage current is flowing.

More Info:
Low-pass filter - Wikipedia -> Electronic low-pass filter
 

swathi.sp

Involved In Discussions
#6
Well, actually frequency analysis comes for free here. Just look at the right figure, that is the frequency response of the low pass filter / measuring device in the left.

It is a simple first order low-pass with a cut-off frequency of 1061 Hz (-3db, defined by R1 and C1). The resistor R2 is the shunt resistor through which the leakage current is flowing.

More Info:
Low-pass filter - Wikipedia -> Electronic low-pass filter
Many thanks Benjamin. Your response is so helpful for me to understand further.
 
Thread starter Similar threads Forum Replies Date
S Routine Testing for Class I Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 24
A Earth (ground) leakage current routine test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Routine Bioburden Monitoring ISO 13485:2016 - Medical Device Quality Management Systems 5
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
T Nonconforming product at customer detected by a routine inspection by field service Nonconformance and Corrective Action 9
N Referencing of a standard's year in routine regulatory compliance documents Other Medical Device Related Standards 3
M Medical Device News FDA - Eliminating Routine FDA news 14-09-18 - Re-Review of Third Party 510(k) Reviews Other US Medical Device Regulations 0
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
C Dose Mapping - How to relate back to Routine Monitoring Position ISO 13485:2016 - Medical Device Quality Management Systems 3
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
J FDA Routine Inspection Pre-Anouncement - What will be checked? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
M Identifying Routine and Non-Routine Activities: OSHA OHSAS 18001 Occupational Health & Safety Management Standards 15
S How you define 'Normal & Abnormal' Conditions and 'Routine & Non-routine' Activities ISO 14001:2015 Specific Discussions 8
J Routine test for MD finished product IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
RoxaneB Environmental Objectives - Do they need to be linked into with Routine Indicators? Miscellaneous Environmental Standards and EMS Related Discussions 6
RoxaneB Search for another name for "Standardization Process" or "Routine Management" Quality Tools, Improvement and Analysis 4
D Line clearance - Routine components in the prep room for universal use ISO 13485:2016 - Medical Device Quality Management Systems 8
R Inspection and Detection - Kit to production - Need an inspection routine Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
C Biocompatability Testing - Electric Breast Pumps , Silicone Component EU Medical Device Regulations 3
L Proficiency Testing Statistical Analysis ISO 17025 related Discussions 0
T Changes to product IFU & Summative Testing IEC 62366 - Medical Device Usability Engineering 1
A Chemotherapy Resistance Testing for EU EU Medical Device Regulations 1
N FDA mechanical bench testing Manufacturing and Related Processes 5
E Microbial Barrier Testing for Medical Devices Qualification and Validation (including 21 CFR Part 11) 2
A Safety testing of Medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
C When should I take endotoxin testing? EU Medical Device Regulations 3
A Safety testing for Medical devices to be sold in India ISO 13485:2016 - Medical Device Quality Management Systems 0
M Test Lab Recommends ANSI C63.27 Wireless Coexistence Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
P Biocompatibility testing ISO-10993 for FDA submission Other Medical Device Related Standards 7
R Tips on transitioning out of testing lab Career and Occupation Discussions 0
T Aircraft GAPP Software Testing Compliance Question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Usability for Near Patient Testing Test Kit Human Factors and Ergonomics in Engineering 2
J Single Fault Testing for Secondary Circuits (60601-1 section 13.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Medical device testing lab/service standard Other Medical Device Related Standards 1
T Cytoxtoxicity Testing Dilution Other Medical Device Related Standards 1
M Altitude testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
H EMC & Safety Testing of the product with the same Sr. No. for EC Declaration CE Marking (Conformité Européene) / CB Scheme 3
A Non-medical device testing in the medical system IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W CQI-9 Testing Fluids Manufacturing and Related Processes 0
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
T Testing for proving food safety on USP class vi material RoHS, REACH, ELV, IMDS and Restricted Substances 0
J Risk Analysis for Proficiency Testing Reliability Analysis - Predictions, Testing and Standards 1
Rene Minassian Control of (TMME) Testing and Measurement Equipment Oil and Gas Industry Standards and Regulations 2
M Is complete testing required as per ISO 10993 for materials used in orthopedic implants or is literature review route possible Other Medical Device Related Standards 3
B A testing lab within an organization ISO 17025 related Discussions 10
B Package integrity Testing failure ISO 13485:2016 - Medical Device Quality Management Systems 7
R Basic Lab Practices - related to ISO 17025 testing labs ISO 17025 related Discussions 1

Similar threads

Top Bottom