Hi everybody. I'm new to this forum and would really appreciate any help with regards to the following. If we are using IEC 60601 as a basis for routine testing of medical electrical systems, what specific electrical safety tests should be performed?

From Clause 16 of 60601, I can find reference to these tests:

- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)

If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?

- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?

Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?

Many thanks for any help.

Michael
 
Last edited:
Elsmar Forum Sponsor

Ahang

Registered
#2
Hi everybody. I'm new to this forum and would really appreciate any help with regards to the following. If we are using IEC 60601 as a basis for routine testing of medical electrical systems, what specific electrical safety tests should be performed?

From Clause 16 of 60601, I can find reference to these tests:

- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)

If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?

- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?

Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?

Many thanks for any help.

Michael
Hello Michael,

- Answer to your first question:
As stated in sub-clause 16.1,
1- you should first specify that which part(s) of your me system is(are) installed in the patient environment and which is(are) outside of this environment.
2- For those parts which are in the patient environment whether they are medical devices or non-medical devices (e.g. a laptop connected to the medical device in order to extract data), all me equipment relevant requirements shall be applied according to 60601-1 (e.g. all parts of clause 8).
3- For those parts which are not in the patient environment, their related safety standards shall be applied.
As an example imagine that you connect an active medical device (which is connected to the patient) through a network to a monitor in the nurse station, in this case, the medical device is in the patient environment and shall be tested according all applicable requirements of IEC 60601-1. But the monitor is outside of the patient environment and shall be in compliance with ITE safety standards (e.g. IEC 60950)

Answer to your 2'nd question:
Refer to IEC 60601-1, 8.7.1 b) 3’rd dash, all leakage currents including touch current shall be measured in both normal condition and single fault condition.

Ahang Rabiee
 

swathi.sp

Involved In Discussions
#3
Dear Folks,

Further to Leakage current test as per IEC 60601-1 , Clause 8.7, we have our own designed human body impedence circuit "as illustrated below representead as "MD in 60601-1. Can anybody help me to understand how to analyse the frequency characteristics of the Measuring device.



1568107273602.png
 

Benjamin Weber

Trusted Information Resource
#5
Dear Folks,

Further to Leakage current test as per IEC 60601-1 , Clause 8.7, we have our own designed human body impedence circuit "as illustrated below representead as "MD in 60601-1. Can anybody help me to understand how to analyse the frequency characteristics of the Measuring device.



View attachment 26052
Well, actually frequency analysis comes for free here. Just look at the right figure, that is the frequency response of the low pass filter / measuring device in the left.

It is a simple first order low-pass with a cut-off frequency of 1061 Hz (-3db, defined by R1 and C1). The resistor R2 is the shunt resistor through which the leakage current is flowing.

More Info:
Low-pass filter - Wikipedia -> Electronic low-pass filter
 

swathi.sp

Involved In Discussions
#6
Well, actually frequency analysis comes for free here. Just look at the right figure, that is the frequency response of the low pass filter / measuring device in the left.

It is a simple first order low-pass with a cut-off frequency of 1061 Hz (-3db, defined by R1 and C1). The resistor R2 is the shunt resistor through which the leakage current is flowing.

More Info:
Low-pass filter - Wikipedia -> Electronic low-pass filter
Many thanks Benjamin. Your response is so helpful for me to understand further.
 
Thread starter Similar threads Forum Replies Date
S Routine Testing for Class I Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 24
B Routine Bioburden Monitoring ISO 13485:2016 - Medical Device Quality Management Systems 5
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
T Nonconforming product at customer detected by a routine inspection by field service Nonconformance and Corrective Action 9
N Referencing of a standard's year in routine regulatory compliance documents Other Medical Device Related Standards 3
M Medical Device News FDA - Eliminating Routine FDA news 14-09-18 - Re-Review of Third Party 510(k) Reviews Other US Medical Device Regulations 0
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
C Dose Mapping - How to relate back to Routine Monitoring Position ISO 13485:2016 - Medical Device Quality Management Systems 3
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
J FDA Routine Inspection Pre-Anouncement - What will be checked? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
M Identifying Routine and Non-Routine Activities: OSHA OHSAS 18001 Occupational Health & Safety Management Standards 6
S How you define 'Normal & Abnormal' Conditions and 'Routine & Non-routine' Activities ISO 14001:2015 Specific Discussions 8
J Routine test for MD finished product IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
RoxaneB Environmental Objectives - Do they need to be linked into with Routine Indicators? Miscellaneous Environmental Standards and EMS Related Discussions 6
RoxaneB Search for another name for "Standardization Process" or "Routine Management" Quality Tools, Improvement and Analysis 4
D Line clearance - Routine components in the prep room for universal use ISO 13485:2016 - Medical Device Quality Management Systems 8
R Inspection and Detection - Kit to production - Need an inspection routine Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
V How to select the tests for proficiency testing Measurement Uncertainty (MU) 0
B Customer Preference Testing Customer and Company Specific Requirements 2
N Pre-Clinical Performance Testing Design and Development of Products and Processes 3
I Software (SaMD) mobile application verification testing: objective evidence Medical Information Technology, Medical Software and Health Informatics 2
I How to find required testing for a specific device? Other US Medical Device Regulations 3
J CQI-9v4 Alternative TUS Testing Methods Various Other Specifications, Standards, and related Requirements 0
M ASTM F1717 Static and Dynamic Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T Mold Limits for Microbiology Testing Other Medical Device Related Standards 3
S Orthopedic implant sterility testing Medical Device and FDA Regulations and Standards News 1
H Independence between the development and testing IEC 62304 - Medical Device Software Life Cycle Processes 6
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Testing criteria - where to place Design and Development of Products and Processes 2
E 60601-1 - Tilt testing - Tensile safety factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
gramps What do you think about automated QA testing For software app industry? Misc. Quality Assurance and Business Systems Related Topics 5
B In house NIOSH pre Testing accepted by NIOSH? US Food and Drug Administration (FDA) 1
M Bacteriostasis/Fungistasis Testing Other Medical Device and Orthopedic Related Topics 6
P Sample Size for Distribution Simulation Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
N EN 813, EN 12277, EN 1497 - Testing some harness prototypes to an EN standard Various Other Specifications, Standards, and related Requirements 0
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 9
B NIOSH Approval for Surgical N95 Respirators - Required testing US Food and Drug Administration (FDA) 2
M ECG lead leakage currents - How to specify ECG leads during electrical safety testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
D CFR Title 14: Aeronautics and Space PART 120—DRUG AND ALCOHOL TESTING PROGRAM Federal Aviation Administration (FAA) Standards and Requirements 3
lanley liao Purchase Acceptance Criteria - Tensile testing Oil and Gas Industry Standards and Regulations 2
M Device mounted at IV pole - what about mechanical stability testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Cochlear Implant Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Class I (exempt) testing requirements Other Medical Device Related Standards 1

Similar threads

Top Bottom