Hi everybody. I'm new to this forum and would really appreciate any help with regards to the following. If we are using IEC 60601 as a basis for routine testing of medical electrical systems, what specific electrical safety tests should be performed?
From Clause 16 of 60601, I can find reference to these tests:
- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)
If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?
- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?
Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?
Many thanks for any help.
Michael
From Clause 16 of 60601, I can find reference to these tests:
- protective earth resistance (16.9.2.2)
- earth leakage in normal condition (16.6.2)
- touch current in normal condition and with earth disconnected (16.6.1)
- patient leakage testing in normal condition (16.6.3)
If it is assumed that all relevant parts of the system have already been tested to 60601 at an equipment level, what other tests should we be doing to ensure the system is safe?
- Insulation testing?
- Measuring voltage on accessible parts?
- Patient auxiliary current?
- Mains on applied parts?
Also, why is the touch current test the only test where singe fault condition testing is required (and then only with earth conductor disconnection, not neutral conductor disconnection)?
Many thanks for any help.
Michael