Safety & EMC certificate updated when MCU (MIcro-Controller) is changed?

B

bennylee

We have the 3rd version of Safety 60601-1 & EMC 60601-1-2 certificate

As our current-used MCU will be faded out, we need to change to other MCU (micro-controller) with same function.
For this change, it will cause
  • PCB layout change only &
  • Accordingly, some component in PCBA was changed
(Note: The main point is no change in Intended for use, performance, output, outercase)

Certificate need to re-apply ?
For this case, do we need to update the Safety & EMC certificate?
In case if not update, what documentation is required?
For our planned action is to prepare some documentation by comparing the difference between OLD MCU & NEW MCU, & providing the related internal test & external test (For external test, we concern for EMC, we will have test from 3rd party), is it enough?

FDA 510(k) need to re-apply ?
I have checked the “Deciding When to Submit a 510(k) for a Change to an Existing Device”, for our case, for class 2 medical device products, it seems that the above change are required to “documentation” only, is it correct ?
 
P

PaulGr

With respect to your 510(k): as this is about replacing an obsolete component, this is almost certainly a change that does not require submission of a new 510(k). Please note that it is still the same amount of work (documentation) but no need to submit to FDA.

With respect to IEC60601: you have to demonstrate why your 60601-1 and 60601-1-2 testreports (and derived certificates) are still valid after applying the change. Small layout changes can have impact on EMC compliance just as introducing a new processor running on different clock frequencies. I have even seen an example where a 'small layout change' reduced creepage on a board from 10mm to 2mm where 8mm was required leading to a large recall.

As whether or not to retest depends on many other factors (architecture, critical parts changed, were you close to the limits in the current design, etc). Many testlabs are willing to discuss with you if retesting is required although you (legal manufacturer) will always be finally responsible.
 
Thread starter Similar threads Forum Replies Date
H EMC & Safety Testing of the product with the same Sr. No. for EC Declaration CE Marking (Conformité Européene) / CB Scheme 3
H Regarding the validity of EMC & Safety test reports without NABL & ILAC-MRA mark. CE Marking (Conformité Européene) / CB Scheme 2
H Regarding EMC & Safety retesting requirements of the product CE Marking (Conformité Européene) / CB Scheme 2
M EMC Directive and product safety standards CE Marking (Conformité Européene) / CB Scheme 1
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
O Guide on EMC for Functional Safety CE Marking (Conformité Européene) / CB Scheme 3
N Device with Safety & EMC Test Certificates from different Laboratories ISO 13485:2016 - Medical Device Quality Management Systems 22
M Combining EcoDesign with Compliance (Product Safety/EMC) Engineer Career and Occupation Discussions 2
Y New Japanese EMC/electrical safety standards - Additional product testing required? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P EN 62304 Safety Class of an ECG device designed to be used in clinical situations when patient is in immediate danger. IEC 62304 - Medical Device Software Life Cycle Processes 3
R Regarding safety requirements of the Lid latch mechanism CE Marking (Conformité Européene) / CB Scheme 1
D FDA combo product-need to define direct & indirect impact regarding safety,quality & regulatory.Can anyone point me in right direction-SupplierMgmt 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
E QMS references in Safety Assurance Case? US Food and Drug Administration (FDA) 4
F Safety program The Reading Room 1
J Determination of software safety class (62304) prior to software risk analysis ISO 14971 - Medical Device Risk Management 3
M Software Safety Classification and Legacy Software IEC 62304 - Medical Device Software Life Cycle Processes 4
A Safety testing of Medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
A Safety testing for Medical devices to be sold in India ISO 13485:2016 - Medical Device Quality Management Systems 0
Q Where to place our Safety Program in our QMS? Occupational Health & Safety Management Standards 9
Q Safety Training Program Training - Internal, External, Online and Distance Learning 8
T Book Excerpt: Drugs and the FDA: Safety, Efficacy, and the Public’s Trust Book, Video, Blog and Web Site Reviews and Recommendations 0
M Safety assessment of cosmetic product EU Medical Device Regulations 1
A Quality and Safety in one. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
D ISO 14001 Finding - Missing Safety Data Sheets ISO 14001:2015 Specific Discussions 2
V Functional safety for Fuses and Fuse boxes ISO 26262 - Road vehicles – Functional safety 0
armani Monitoring and measuring equipment for occupational health and safety purposes Occupational Health & Safety Management Standards 10
T Testing for proving food safety on USP class vi material RoHS, REACH, ELV, IMDS and Restricted Substances 0
T Internal and external communication procedure for Food Safety Food Safety - ISO 22000, HACCP (21 CFR 120) 2
H Product Safety & Conformity Representative (PSCR) Training Company IATF 16949 - Automotive Quality Systems Standard 0
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
K Electrical Safety Tester ISO 13485:2016 - Medical Device Quality Management Systems 7
H Harmonized Standard for the refrigerated lab device safety CE Marking (Conformité Européene) / CB Scheme 2
L Article 87 Field Safety Corrective Actions in a third country EU Medical Device Regulations 2
S Risk control through Information for safety ISO 14971 - Medical Device Risk Management 12
Ron Rompen 4.4.1.2 Product Safety Compliance IATF 16949 - Automotive Quality Systems Standard 30
B QMS Functional Safety Checklist ISO 26262 - Road vehicles – Functional safety 0
H Characteristics related to safety - format of documentation ISO 14971 - Medical Device Risk Management 3
Watchcat Raising new questions of safety and effectiveness Other US Medical Device Regulations 4
Anerol C In Charge of Safety - New- Safety program Example Occupational Health & Safety Management Standards 6
Aymaneh Summary of safety and performance (SSP) IVDR 2017/746 EU Medical Device Regulations 1
S Summary of safety and clinical performance and Article 120(3) of the MDR EU Medical Device Regulations 4
Sidney Vianna Food & Beverage Safety Lead Auditor Open Position - USA Job Openings, Consulting and Employment Opportunities 0
D ISO 26262 functional safety : how to get started Other ISO and International Standards and European Regulations 8
A What is different between Field Safety Notice and Advisory Notice in U.S and EU? Document Control Systems, Procedures, Forms and Templates 7
A TGA - Annual Report or Periodic Safety Update Report? Other Medical Device Regulations World-Wide 0
B IVDR Electrical safety standard EU Medical Device Regulations 0
Z Choice of PTC heater voltage to maximize user safety and ease of CE compliance to LVD CE Marking (Conformité Européene) / CB Scheme 0
M Post Market Safety Update Report for devices that will be up-classified under MDR but are currently under MDD EU Medical Device Regulations 6

Similar threads

Top Bottom