I read somewhere that the SAL for High Level Disinfection is 10^-3 verses a SAL of 10^-6 for sterilization. Now I can't find the source and I am wondering if I was dreaming.
Does anyone know if for a FDA class II medical device if a tested SAL level of <10^-3 (but not <10^-6) is sufficient for claiming High Level Disinfectability?
Thanks!
:confused:
Marc
Re: SAL for High Level Disinfection vs Sterilization
For those of us with inquiring minds who are not in this field, what is SAL?
harry
Re: SAL (Sterility Assurance Level) for High Level Disinfection vs Sterilization
Check this: Directory of open access journals - The limits of sterility assurance
Proceeding from this, we would like to make the following suggestions for tiered SAL values, adjusted according to the respective sterilization task: * SAL 10–6 for heat-resistant pharmaceutical preparations (parenterals), suggested term: “Pharmaceutical sterilization”, * SAL 10–4 for heat-resistant medical devices, suggested term: “High-level sterilization”, * SAL 10–3 for heat-sensitive re-usable medical devices, under the precondition of a validated cleaning efficacy of >4 lg increments, suggested term: “Low-level sterilization”.
MIREGMGR
Are you referring to using a germicide with a High-Level-Disinfectant indication that also has a Sterilant indication, under its required Sterilant use-conditions?
Two FDA reference documents:
FDA-Cleared Sterilants and High Level Disinfectants with General Claims for Processing Reusable Medical and Dental Devices
Content and Format of Premarket Notification [510(k)] Submissions for Liquid Chemical Sterilants/High Level Disinfectants
In the latter document, the following statement appears:
"A SAL cannot be inferred for a device sterilization claim based on this recommended testing protocol."
(FDA's emphasis.)
W
WilliamS
The FDA has clear guidelines for validation of "sterility" (10-6) and HLD (10-).
I've worked for 2 medical device companies (not Pharm) who's products fall into one or both of these areas. Both have validated there processes and passed FDA Audits.
There's no need to try and rename these "levels of sterility assurances. It should be remembered that even at 10-6, there is the mathimatical probability that there will be one non-sterile package.
Broccoli
Thank you WilliamS
Could you please check the first line of your response again and tell me what the power of HLD you were trying to enter was?
Also, do you have a reference for this?
Thank you :)!
dendrenn
Hi Broccoli, I realise that this is probably coming a little late for you but as I came across your post while I was looking for this information myself, I thought I would leave my findings for future reference.
I don't have sufficient 'Cove' experience to post you the link directly but if you Google the 'CDC Guideline for Disinfection and Sterilization in Healthcare Facilities', in it you will see that:
The FDA definition of high-level disinfection is a sterilant used for a shorter contact time to achieve a 6-log10 kill of an appropriate Mycobacterium species.
This on the face of it looks like an equivalent requirement to that of sterilisation i.e. a SAL of 10^-6, but I believe it is the fact that a mycobacterium species is used for validation that makes the difference; high-level disinfection is not required to kill anything beyond mycobacteria e.g. bacterial spores:
Unlike sterilization, disinfection is not sporicidal. A few disinfectants will kill spores with prolonged exposure times (3?12 hours); these are called chemical sterilants. At similar concentrations but with shorter exposure periods (e.g., 20 minutes for 2% glutaraldehyde), these same disinfectants will kill all microorganisms except large numbers of bacterial spores; they are called high-level disinfectants.
S Sterility Assurance Level of Liquids (SAL) - Special Culture Medium ISO 13485:2016 - Medical Device Quality Management Systems 3
P Submission of SAL for site in DMF US Food and Drug Administration (FDA) 3
S Aseptic Processing - Maximum SAL using Aseptic Processing for Sterilization Other ISO and International Standards and European Regulations 4
C How to verify the SAL (sterilie acceptable level) is 10-6 for ETO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 6
D Sterility and BioBurden Testing for Plastics ISO 13485:2016 - Medical Device Quality Management Systems 5
C Sterility and Bioburden Testing - LAL passed under 0.5EU/mL Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 1
B Sterility test on gloves ISO 13485:2016 - Medical Device Quality Management Systems 3
M Medical Devices Sterility - Question from A Regulator Other Medical Device Regulations World-Wide 3
O Freeze-Dried Powder Sterility Test using the Membrane Filtration Technique Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
M How many Pieces to Test - Sterility Testing USP EU Medical Device Regulations 2
B IPCD Sterility test failure during MPQ (EO sterilization validation) ISO 13485:2016 - Medical Device Quality Management Systems 2
N Sterility Testing - USP Chapter 71 and ISO 11737-2 ISO 13485:2016 - Medical Device Quality Management Systems 5
H ISO 13485 for Sterility Only? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Shelf Life Determination - Sterility test (Stasis test) ISO 13485:2016 - Medical Device Quality Management Systems 12
M Sterility Testing vs Bioburden Testing - Differences Qualification and Validation (including 21 CFR Part 11) 7
N Consultant Obligations, CFR 803 and Sterility Quality Manager and Management Related Issues 4
A Shelf Life Testing - We would like to extend the sterility expiration date Misc. Quality Assurance and Business Systems Related Topics 1
A Sterility Assurance Level? Special 510k US Food and Drug Administration (FDA) 5
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
M GLP (Good Laboratory Practices) Requirement for Sterility Validations Qualification and Validation (including 21 CFR Part 11) 1
J Container Closure Integrity & Sterility Test - The Dye Immersion Method Quality Assurance and Compliance Software Tools and Solutions 16
S Sterility - Class I or Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Suitable Sterilization Cycle to achieve Sterility Assurance Level of <10-6 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Sterility (bacteriostasis/fungistasis) Testing by Membrane Filtration validation Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
D FDA Guidance on Computer Software Assurance versus 21 CFR Part 11 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Marc Job Opportunity – Quality Assurance Specialist - Suspense 7 April 2021 Job Openings, Consulting and Employment Opportunities 0
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Conformity assessment based on a quality management system or production quality assurance EU Medical Device Regulations 3
D Safety Assurance Case Report Format ISO 14971 - Medical Device Risk Management 4
C Quality assurance in Pharmacovigilance Document Control Systems, Procedures, Forms and Templates 2
F Risk for Quality Assurance Department in a Hospital - Hospital Incident Reporting ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
J Quality Assurance in China - Developing a quality management system for a California company Misc. Quality Assurance and Business Systems Related Topics 9
C Independence of DQA (Design Quality Assurance) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Pancho Definition Quality Assurance Program - Definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
chris1price MDR Conformity Assessment for a Class 2a device - Production Quality Assurance EU Medical Device Regulations 2
L Definition Mini Builds - Quality Assurance Terms and Definitions Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
Marc Adventures in Design Quality Assurance - Two-week-old Pixel 2 XL displays World News 0
P Quality Assurance and Quality Control - Which clauses of ISO 9001 deal with each? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A KPI for Quality Assurance Team in a Service Oriented Industry Service Industry Specific Topics 8
C I'm starting my Career in Quality Assurance Career and Occupation Discussions 8
E Employees in Quality Assurance Jobs Career and Occupation Discussions 2
Y Outsourced aerospace work quality assurance? Supplier Quality Assurance and other Supplier Issues 6
R Quality Assurance Assessment of the F-35 Lightning II Program AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
M Quality Assurance in an Architectural firm Misc. Quality Assurance and Business Systems Related Topics 3
M QAM (Quality Assurance Matrix) for a sterilization process Quality Tools, Improvement and Analysis 1
S Quality Assurance Plan template for CNC Milling Industry AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2