SaMD - Software as a Medical Device - Software change control form

racglobal

Involved In Discussions
#1
Hello everyone,

For a company that makes SaMD, (software as a medical device), does anyone have a design change control form template? I know in the world of hardware devices, there are engineering change controls and the associated forms where you assess the impact. Does the world of software have difference requirements on the change control form? Can somebody provide some insight and a template they would like to point me to ?


Thanks.
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
It's still a medical device so I would suggest you always:
  • [from 13485] assess (and document) the significance of the change to function, performance, usability, safety and applicable regulatory requirements
  • [from 13485 and effectively 820] document what you did regarding review, verification and/or validation, and approval (to make the change)
  • [from 13485] assess (and document) the effect of the changes on constituent parts and product in process or already delivered, inputs or outputs of risk management and product realization processes.
  • [from 62304] assess (and document) the effect on the organization, released MEDICAL DEVICE SOFTWARE, released for intended use, and SYSTEMS with which it interfaces
  • [from 62304] assess (and document) a statement of criticality (for example, the effect on performance, SAFETY, or SECURITY) as well as other information that may aid in the resolution of the problem (for example, devices affected, supported accessories affected)
  • [from 62304]
    1. investigate the problem and if possible identify the causes;
    2. EVALUATE the problem’s relevance to SAFETY using the software RISK MANAGEMENT PROCESS;
    3. document the outcome of the investigation and evaluation; and
    4. create a CHANGE REQUEST(S) for actions needed to correct the problem, or document the rationale for
      taking no action.
  • [from 62304] advise relevant parties of the existence of the problem, as appropriate
You should also consider the impact on usability [62366].

Your form can facilitate driving/capturing all this.
 

Ed Panek

QA RA Small Med Dev Company
Trusted Information Resource
#3
Also evaluate the user experience change from the software iteration. A 510K alteration review is advice from FDA that should be reviewed with all SW changes. Regression testing deliberation should also be documented.
 

Lokus200

Starting to get Involved
#4
Same change process applies to all medical devices, you will have to follow the same process.
I assume as a software developer you are using tools such as JIRA to document defects, you can adjust your tools to include the change process instead of creating additional paperwork. Just make sure you add regulatory evaluation of every change and document why new 510k is not required.
 
Thread starter Similar threads Forum Replies Date
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 4
Q Storing and developing SAMD (Software as a Medical Device) in the Cloud IEC 62304 - Medical Device Software Life Cycle Processes 3
F Software as a Medical Device (SaMD) Technical File Requirements Manufacturing and Related Processes 3
A Software as Medical Device (SaMD) definition and its applicability Other Medical Device and Orthopedic Related Topics 4
M Informational TGA – Submissions received: Regulation of software, including Software as a Medical Device (SaMD) Medical Device and FDA Regulations and Standards News 1
M Informational TGA – Webinar: An update on the consultation for the Regulation of Software, including Software as a Medical Device (SaMD) Medical Device and FDA Regulations and Standards News 0
J Qualification of a Software as a Medical Device (SaMD) guidance under MDR EU Medical Device Regulations 9
M Informational TGA Consultation: Regulation of software, including Software as a Medical Device (SaMD) Medical Device and FDA Regulations and Standards News 0
M Medical Device News Health Canada - Scientific Advisory Panel on Software as a Medical Device (SAP-SaMD) - Record of Proceedings – January 26, 2018 Canada Medical Device Regulations 0
K Registering a Software medical device (SaMD) in China China Medical Device Regulations 5
R SaMD - Placing on the Market before DoA EU Medical Device Regulations 2
D CE SaMD modular approach CE Marking (Conformité Européene) / CB Scheme 5
K Regulatory requirement of SaMD with machine learning component IEC 62304 - Medical Device Software Life Cycle Processes 3
Y Distributor for a SaMD Medical Information Technology, Medical Software and Health Informatics 3
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 7
A Demonstration of Equivalence - Need for comparing biological characteristics for an SamD EU Medical Device Regulations 1
N Servicing for SaMD ISO 13485:2016 - Medical Device Quality Management Systems 6
N Agile SOP for SaMD IEC 62304 - Medical Device Software Life Cycle Processes 3
D Not a SaMD - How to be submitted for marketing authorization in EU and USA CE Marking (Conformité Européene) / CB Scheme 5
T Medical device or not - SaMD Other Medical Device and Orthopedic Related Topics 3
C Supplier Assessment for Hardware supporting SaMD ISO 13485:2016 - Medical Device Quality Management Systems 1
T MEDDEV 2.7/1 r4 vs. IMDRF/SaMD WG/N41 Other Medical Device Related Standards 3
M IMDRF document - Application of QMS to SaMD ISO 13485:2016 - Medical Device Quality Management Systems 2
W IEC 62304 vs. IMDRF SaMD Guideline Risk Class IEC 62304 - Medical Device Software Life Cycle Processes 5
S How to perform verification of the Statistical Analysis Software? Qualification and Validation (including 21 CFR Part 11) 2
I Document Control Software Document Control Systems, Procedures, Forms and Templates 2
E Software maintenance Process Software maintenance Process to IEC 6204? IEC 62304 - Medical Device Software Life Cycle Processes 3
L Micro-Vu InSpec Software Program Qualification and Validation (including 21 CFR Part 11) 6
A For software change - New Channel of interoperability CE Marking (Conformité Européene) / CB Scheme 4
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
C SharePoint Contract Management Software General Information Resources 0
gramps What do you think about automated QA testing For software app industry? Misc. Quality Assurance and Business Systems Related Topics 5
U API Spec Q1 - 5.6.1.2 C (3) - Design software Oil and Gas Industry Standards and Regulations 3
B Complaint Records - Accessing records on Easy Track Software Records and Data - Quality, Legal and Other Evidence 3
GreatNate Master Control QMS software Quality Tools, Improvement and Analysis 0
GreatNate Anyone using the Intellect QMS software? Quality Assurance and Compliance Software Tools and Solutions 1
S DHF/DMR/MDF for a software-only, cloud-based, single-instance device Medical Information Technology, Medical Software and Health Informatics 2
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
J Do Software Subcontractors need to be ISO13485 compliant in the EU? EU Medical Device Regulations 3
D Safety data sheets software REACH and RoHS Conversations 2
N What are the software audit and control steps Reliability Analysis - Predictions, Testing and Standards 2
N Validating Software before getting approved as Class 2 device US Food and Drug Administration (FDA) 5
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
P Missing 1m visual alarm signal in case of software/display failure, mitigation? ISO 14971 - Medical Device Risk Management 3
B Software service provider as critical supplier ISO 13485:2016 - Medical Device Quality Management Systems 5
S Asterisk in DOE minitab software Using Minitab Software 23

Similar threads

Top Bottom