Dear Statisticians and regulatory experts,
Kindly help me to solve this problem for one of the medical device project. I asked many statisticians, but none were able to give convincing answer from the regulatory standpoint as you know FDA is very strict with the evidences and rationals. I found the similar problem on this forum but couldn't find the end answer.
Problem: The manufacturer of medical endoscope device equipment (pretty expensive prototypes) and they are trying to figure out justification for using 5, 7 or 10 samples with sufficient confidence level and reliability to proceed w/ manufacturing.
Testing such as bend cycling and environmental testing etc which they can do in house, for these tests sample sizes need to be determined (for each test). *These sample sizes can be calculated based on confidence and reliability levels, but a precedent should be set for the confidence/reliability that is needed for these devices and testing based on the risk level of the testing and the expense of running these tests.
At least with the examples of similar kind with formula and method!
I have very limited statistical knowledge. Kindly help me to solve this problem with the solutions from the regulatory standpoint.
Thank you!
Harini17
Thank you Paulag.
I would like to know the reliability and confidence interval percentages that we can to use for these tests, so that FDA accepts. I would like to know the FDA requirements for these tests.
Thank you!
I am afraid you are asking for something that cannot be answered to your satisfaction. Sample sizes that will 'satisfy the FDA' are not easy or straightforward. You cannot simply plug numbers into a formula. We can give you the formula (I believe that it is in the attachment in the above post) but YOU must come up with the numbers that go into the formula based on your organizations assessment of the risk (hazard) of each characteristic that you will be testing.
The formula that incorporates Reliability with a specified confidence level is:
n = LN(1-C)/LN(R)
n= sample size
LN is the natural log
1-C is 1- confidence level (traditionally this is 95% or 99%. a higher confidence level is selected if the hazard is high)
R is the 'Reliability'. This is 1-defect rate that you can tolerate. again a severe hazard would require a pretty low defect rate
This formula requires that NO DEFECTS be found in the sample in order to accept the validation results.
This formula is based on using only pass/fail categorical results. you can get smaller sample sizes if you have continuous data where you can make predictions regarding the distribution vs. specification limits. These typically require a statistician or at least someone who is well versed in these types of statistics.
Severe hazards require large sample sizes per this Reliability/Confidence formula - although this typically results in the lowest sample sizes possible. Often companies will balk at high sample sizes and so the FDA provides the ability to provide a rationale based on balancing cost of testing vs. cost of the hazard. you can plug various numbers into the formula and see the effect for yourself.
Beyond this you have the complication of understanding that sample size applies to both the number of instruments/products and the number of USES of the product. The reliability/confidence formula typically applies to the number of instruments and the number of uses is based on the either the specified number of maximum uses or the maximum expected number of uses.
Statistical Steven
Statistician
Harini17
Thank you, Bev D and Statistical Steven...I have read all your threads addressing these questions thank so much. You both are doing good work in solving these regulatory obstacles. Special thanks to Bev D for such detailed explanation!
v9991
:agree1:
while FDA is providing enough direction, there is not even an outline/framework or guidance from agency for same. got certain references on risk from PDA & ISPE ; but certainly not ones which reference cost of testvshazard.!
1. point is, agency has ultimate power (veto to say the system/practice is inadequate, non-compliant) but what it says is ...all theory, no reference/ example, guidance, framework; (I remember an section on risk benefit analysis on clinical trial data evaluation!., but no such thing on validations/sample size.)
2. is this one reason, why (especially) pharma is way behind in adopting these tools and techniques. (its only recently that 'statistical' rationale/analysis gaining releac env
Thanks all for this, which has been incredibly informative for determining sample size.
In Paulag's attachment, the example for continuous data states "Unknown Standard Deviation and Delta, but want to detect a 0.5 standard deviation shift." How would one decide the shift that is necessary to detect?
Thanks Bev D. I'm curious what would determine when to use the "sample size for continuous data" formula vs. the k-value formula.
Assuming the same specification as the example, for 95% confidence and 95% reliability, I would get a sample size 52 using the continuous data formula vs sample size 21 using the k-value formula.
Is it only existing information (mean, standard deviation) that justifies using the lower sample size?
T Sample size calculation Using Minitab Software 3
V Sample Size Calculation for Image Analysis - Microscopy Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
J Minitab Sample Size Calculation: 1 Sample t vs Minimum Sample Size for Means Using Minitab Software 3
P Sample Size Calculation help needed Reliability Analysis - Predictions, Testing and Standards 4
C Sample Size Calculation Formula for Process Capability Statistical Analysis Tools, Techniques and SPC 25
P Sample Size for Capability Calculation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 33
D Validate 99.5% reliability 90% confidence - Sample size calculation Reliability Analysis - Predictions, Testing and Standards 4
U The Importance of Correct Sample Size Calculation - Greek Alphabet Soup Book, Video, Blog and Web Site Reviews and Recommendations 7
T Sample Size Calculation for PPM (Parts Per Million) Prediction Statistical Analysis Tools, Techniques and SPC 6
V Sample Size rationale to reduce size and use special levels EU Medical Device Regulations 6
Q Gage Pin Sets? Calibrate/Verify them all or just a sample size? General Measurement Device and Calibration Topics 22
S How to determine sample size to test when the acceptance criteria is x >= Standard value? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
C What Theory is MIL-STD-105E Sample Size Code Letters Based on? Quality Tools, Improvement and Analysis 3
Y FMEA link to Control Plan frequency and sample size FMEA and Control Plans 19
T Noob question regarding sample size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
S Help needed in choosing the method of calculating the minimum sample size Internal Auditing 12
D Estimate Sample Size for very rare event Using Minitab Software 1
S Determining Sample Size and Method Quality Tools, Improvement and Analysis 12
C Complaint Return Sample Size Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
shimonv Frequency and sample size for lot release of EtO sterilized products EU Medical Device Regulations 2
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
P Sample Size for Distribution Simulation Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
K How can Acceptance Number and Reject Number be larger than Sample Size on Z1.4 Table? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 4
D Design Verification Sample Size vs Repeats Statistical Analysis Tools, Techniques and SPC 9
M Minimum sample size - Guidance and statistical rationale Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 41
J Control chart for huge sample size Statistical Analysis Tools, Techniques and SPC 9
S Is there a standard for sample size during R&D phase Other ISO and International Standards and European Regulations 18
J Sample size definition in an Automotive SMT pilot lot run Misc. Quality Assurance and Business Systems Related Topics 1
E Correct way to certify hydrostatic testing when it is not 100% (and Sample Size) Various Other Specifications, Standards, and related Requirements 6
F Bayes Success run Theorem for sample size during OQ&PQ Qualification and Validation (including 21 CFR Part 11) 4
I Device modifications - Clinical sample size rationale EU Medical Device Regulations 5
J Sample size for creating a data base as a reference to a tested variable Other Medical Device and Orthopedic Related Topics 6
T Sample Size for Design Validation Design and Development of Products and Processes 4
C AS9138 Sample Size Determination AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
V Acceptable maximum RSD (relative standard deviation) for an sample size Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
E Sample size for design verification of variable in single use device Design and Development of Products and Processes 20
R Sample size for clinical validation/investigation EU Medical Device Regulations 4
R Process Validation sample size selection Statistical Analysis Tools, Techniques and SPC 0
J Interesting Discussion Sample Size Determination for Medical Device Other Medical Device and Orthopedic Related Topics 19
P Is it possible to make an educated decision using a very very small sample size? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
F Sample Size for Biocompatibility Tests Other Medical Device Related Standards 4
V Firmware Verification Testing & Sample Size Software Quality Assurance 1
P Cleaning and Disinfection Validation Sample Size of 3? Other Medical Device Related Standards 2